try to order the ClimateLine Heated Tubing to avoid rain out

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
longwood
Posts: 14
Joined: Wed Sep 12, 2018 3:27 pm

try to order the ClimateLine Heated Tubing to avoid rain out

Post by longwood » Tue Oct 23, 2018 5:17 pm

I am trying to order the ClimateLine Heated Tubing to avoid rain out, but my supplier said it won't fit the Resmed autoset sense 10. And also said I need a new prescription for the doctor even it fit. So I must visit my doctor again, let doctor give me a new prescription for a themal tubing?

https://www.cpap.com/productpage/resmed ... ircurve-10

User avatar
Pugsy
Posts: 65144
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: try to order the ClimateLine Heated Tubing to avoid rain out

Post by Pugsy » Tue Oct 23, 2018 5:22 pm

longwood wrote:
Tue Oct 23, 2018 5:17 pm
I am trying to order the ClimateLine Heated Tubing to avoid rain out, but my supplier said it won't fit the Resmed autoset sense 10.
Hogwash.
longwood wrote:
Tue Oct 23, 2018 5:17 pm
And also said I need a new prescription for the doctor even it fit.
More hogwash.
longwood wrote:
Tue Oct 23, 2018 5:17 pm
So I must visit my doctor again, let doctor give me a new prescription for a themal tubing?
I suppose if you want insurance to pay for it..you have to jump through their hoops (they are free to come up with their own hoops) but do the math.
How much is the doctor's visit and how much is your co pay if the insurance pays for it?

Personally...I would just buy it from cpap.com or wherever because it does NOT need a RX to purchase a heated hose (or any hose for that matter) any where on the internet except maybe the yahoos that call themselves your DME.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

longwood
Posts: 14
Joined: Wed Sep 12, 2018 3:27 pm

Re: try to order the ClimateLine Heated Tubing to avoid rain out

Post by longwood » Tue Oct 23, 2018 5:38 pm

sorry not familiar with the web, I just reported and can't see my post ....................

So it fit? it seems they are just trying to fool me around? I could feel the water is moving now during night and afraid it could be worse when winter comes. really dont like the feeling whenever wake up.

yes, I want the insurance pay for it and it's only $10 more than the other. The copay is only 20 dollars and seems not too much, but it's very time comsuming for visting the doctor for this. I would order it from CPAP.com then, since I want to order a mask from them.

Do we have some coupon now for mask or parts?

User avatar
Pugsy
Posts: 65144
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: try to order the ClimateLine Heated Tubing to avoid rain out

Post by Pugsy » Tue Oct 23, 2018 5:42 pm

They are free to come up with their own stupid in house rules.
Doesn't make it right though.

Just be careful that you get the correct heated hose for your model machine.
There's the Climate line for the S9 model
and the Climate line Air for the AirSense/AirCurve models.
S9 heated hose won't fit the AirSense/AirCurve 10 machines.

I have owned both...S9 and AirSense 10 heated hoses...and they do indeed fit and no RX is required...wasn't even required from my DME and no...I haven't seen the doctor in over 9 years.

No problem about the reporting by mistake. I can fix that easily. I sent you a PM essentially saying what I said here.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

old dude
Posts: 643
Joined: Sat Jul 27, 2013 4:08 pm

Re: try to order the ClimateLine Heated Tubing to avoid rain out

Post by old dude » Tue Oct 23, 2018 5:54 pm

As far as fitment goes, I will say this. I have two fairly close serial numbered Air Sense 10 Autoset machines and two Climateline Air hoses. One of the hoses has the little orange stripe on the machine end fitting as shown in the link and the other does not. Although it can be persuaded gently to fit, the hose with the orange stripe is a much tougher fit in both machines than is the plain gray hose, which slides in effortlessly.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: try to order the ClimateLine Heated Tubing to avoid rain out

Post by palerider » Tue Oct 23, 2018 5:57 pm

longwood wrote:
Tue Oct 23, 2018 5:17 pm
I am trying to order the ClimateLine Heated Tubing to avoid rain out, but my supplier said it won't fit the Resmed autoset sense 10.
Get a different "supplier"

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

jimc2
Posts: 30
Joined: Fri Sep 14, 2018 9:13 am
Location: New York

Re: try to order the ClimateLine Heated Tubing to avoid rain out

Post by jimc2 » Tue Oct 23, 2018 6:05 pm

longwood wrote:
Tue Oct 23, 2018 5:38 pm
sorry not familiar with the web, I just reported and can't see my post ....................

So it fit? it seems they are just trying to fool me around? I could feel the water is moving now during night and afraid it could be worse when winter comes. really dont like the feeling whenever wake up.

yes, I want the insurance pay for it and it's only $10 more than the other. The copay is only 20 dollars and seems not too much, but it's very time comsuming for visting the doctor for this. I would order it from CPAP.com then, since I want to order a mask from them.

Do we have some coupon now for mask or parts?
fall18 15% off.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: backup machine:PR SystemOne REMstar Auto 560 Auto-CPAP Machine with A-Flex cmH20 10-20

longwood
Posts: 14
Joined: Wed Sep 12, 2018 3:27 pm

Re: try to order the ClimateLine Heated Tubing to avoid rain out

Post by longwood » Tue Oct 23, 2018 6:12 pm

jimc2 wrote:
Tue Oct 23, 2018 6:05 pm
longwood wrote:
Tue Oct 23, 2018 5:38 pm
sorry not familiar with the web, I just reported and can't see my post ....................

So it fit? it seems they are just trying to fool me around? I could feel the water is moving now during night and afraid it could be worse when winter comes. really dont like the feeling whenever wake up.

yes, I want the insurance pay for it and it's only $10 more than the other. The copay is only 20 dollars and seems not too much, but it's very time comsuming for visting the doctor for this. I would order it from CPAP.com then, since I want to order a mask from them.

Do we have some coupon now for mask or parts?
fall18 15% off.
great and thanks a lot

User avatar
Elphaba
Posts: 8
Joined: Sat Sep 15, 2018 1:05 pm
Location: Colorado..6K'+ Elevation

Re: try to order the ClimateLine Heated Tubing to avoid rain out

Post by Elphaba » Tue Oct 23, 2018 9:15 pm

I had a rainout issue when the weather turned cold. I sent my doc an email and he sent the order to the supplier. Ordered, billed insurance and had it in a couple of days. Might be worth trying—don’t lose but a few minutes if your doc says no.
Christine

I only live for today...but I'm one day behind~~All The Love by The Outfield

longwood
Posts: 14
Joined: Wed Sep 12, 2018 3:27 pm

Re: try to order the ClimateLine Heated Tubing to avoid rain out

Post by longwood » Tue Oct 23, 2018 10:13 pm

Elphaba wrote:
Tue Oct 23, 2018 9:15 pm
I had a rainout issue when the weather turned cold. I sent my doc an email and he sent the order to the supplier. Ordered, billed insurance and had it in a couple of days. Might be worth trying—don’t lose but a few minutes if your doc says no.

Super nice , thanks a lot! I will give this a try tomorrow.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: try to order the ClimateLine Heated Tubing to avoid rain out

Post by chunkyfrog » Tue Oct 23, 2018 10:50 pm

I am not sure the AirSTART machines support the heated hose. It might not.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Pugsy
Posts: 65144
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: try to order the ClimateLine Heated Tubing to avoid rain out

Post by Pugsy » Wed Oct 24, 2018 2:15 am

Oh crap...do you have the AirStart machine and not the AirSense machine?
I was basing my comments on this
longwood wrote:
Tue Oct 23, 2018 5:17 pm
Resmed autoset sense 10
The AirStart machines are NOT heated hose capable...sorry if that is what you have.

Exactly which model do you have ?
the AirSense 10 Autoset
or the AirStart 10 Autoset?

The AirStart is the el cheapo model from ResMed...minimal data (if any) and doesn't have the same climate control options as the AirSense.

Might be worth it to go to the doctor if the new RX is for a machine that is heated hose capable. :lol:

Otherwise there are some options...hose cozy or stand alone heated hose.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

D.H.
Posts: 3532
Joined: Sun Sep 04, 2016 7:07 pm

Re: try to order the ClimateLine Heated Tubing to avoid rain out

Post by D.H. » Wed Oct 24, 2018 7:04 pm

The tube that you need for your machine (the one in your profile) is the "SlimLine Tubing."

BTW, tube and hose are synonyms in the CPAP world.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

longwood
Posts: 14
Joined: Wed Sep 12, 2018 3:27 pm

Re: try to order the ClimateLine Heated Tubing to avoid rain out

Post by longwood » Wed Oct 24, 2018 10:57 pm

I am not sure about it now. will check with the guy give to me tomorrow.

The option on my machine for the tubling is "SlimLine Tubing" and "standard", should there be something else for this option? something like "themal tubling"?

User avatar
Pugsy
Posts: 65144
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: try to order the ClimateLine Heated Tubing to avoid rain out

Post by Pugsy » Wed Oct 24, 2018 11:03 pm

No...there isn't any special option for the heated hose.
Machines that are heated hose capable automatically make the necessary choice once the heated hose is attached.
Those choices pertain to diameter of the hose only.

Look on your machine....do you see AirStart or AirSense?
AirStart (which you have in your profile) is NOT a heated hose capable machine...
AirSense (which you mentioned in your first post here) is a heated hose capable machine.

Which do you have?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.