Resmed Airsense 10 Autoset stops sending data to Resmed Cloud?
Resmed Airsense 10 Autoset stops sending data to Resmed Cloud?
So starting on October 5th my Resmed machine just mysteriously stopped sending data to ResMed. According to the unit it has full cell signal but I can't get any data from MyAir and my sleep clinic can't get data from AirView. The last data we had was from October 4th.
Anyone here have the same issues or has seen modems mysteriously fail?
Could it be that I didn't write protect my card (I am using Macintosh) when I was using sleepyhead and that may have messed up the cellular data upload?
Anyone here have the same issues or has seen modems mysteriously fail?
Could it be that I didn't write protect my card (I am using Macintosh) when I was using sleepyhead and that may have messed up the cellular data upload?
- greatunclebill
- Posts: 1503
- Joined: Mon Feb 20, 2012 7:48 pm
- Location: L.A. (lower alabama)
Re: Resmed Airsense 10 Autoset stops sending data to Resmed Cloud?
the sd card should have nothing to do with the cellular signals.
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Additional Comments: myAir, OSCAR. cms-50D+. airsense 10 auto & (2009) remstar plus m series backups |
First diagnosed 1990
please don't ask me to try nasal. i'm a full face person.
the avatar is Rocco, my Lhasa Apso. Number one "Bama fan. 18 championships and counting.
Life member VFW Post 4328 Alabama
MSgt USAF (E-7) medic Retired 1968-1990
please don't ask me to try nasal. i'm a full face person.
the avatar is Rocco, my Lhasa Apso. Number one "Bama fan. 18 championships and counting.
Life member VFW Post 4328 Alabama
MSgt USAF (E-7) medic Retired 1968-1990
Re: Resmed Airsense 10 Autoset stops sending data to Resmed Cloud?
The A10 Autoset has internal memory for compliance data, so even if you remove the SD card it can still record and send basic compliance data via the cellular modem. It is conceivable that there is a bug that interacts with locked SD cards, but that seems a bit unlikely. It's also possible that there is an error with ResMed's cloud system. Just because you have cell signal doesn't mean that ResMed's accounts are working properly.
This is not an easy issue to trouble shoot without sending it to a qualified repair clinic because we have no easy way to isolate the problem and test different aspects. We can't swap out the modem, We don't have data from the cell company telling us how much data has been transmitted. We don't have access to ResMed's cloud system to see if their is an error there some how.
You can try the usual brute force methods, including resetting the data on the A10. You can also try moving it to a location where you know another A10 with an identical modem is successfully transmitting to rule out issues with the cell tower. And your DME can, perhaps, put in a trouble ticket to the ResMed cloud to see if there are any issues there.
In the meantime, have you checked the cellular modem info in the clinical menu? Not sure if it will tell you anything pertinent, but it is one step forward in arming yourself with data, such as whether you have an outdated modem or not (though the cell bars should be programed to show actual connectivity).
You can bring data to your DME to be read off the card. Easiest thing to do is for your DME to swap out your A10 and see if that works in your location. If it doesn't, then the cell tower connectivity could be an issue.
This is not an easy issue to trouble shoot without sending it to a qualified repair clinic because we have no easy way to isolate the problem and test different aspects. We can't swap out the modem, We don't have data from the cell company telling us how much data has been transmitted. We don't have access to ResMed's cloud system to see if their is an error there some how.
You can try the usual brute force methods, including resetting the data on the A10. You can also try moving it to a location where you know another A10 with an identical modem is successfully transmitting to rule out issues with the cell tower. And your DME can, perhaps, put in a trouble ticket to the ResMed cloud to see if there are any issues there.
In the meantime, have you checked the cellular modem info in the clinical menu? Not sure if it will tell you anything pertinent, but it is one step forward in arming yourself with data, such as whether you have an outdated modem or not (though the cell bars should be programed to show actual connectivity).
You can bring data to your DME to be read off the card. Easiest thing to do is for your DME to swap out your A10 and see if that works in your location. If it doesn't, then the cell tower connectivity could be an issue.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: DreamWear Nasal CPAP Mask with Headgear - Fit Pack (All Cushions Included with Medium Frame) |
Additional Comments: APAP 6-12. EPR 2. Sleepyhead. |
Official Foe™ on @Palerider's public list of foes. 
Rules to live by: Be wary of anyone boasting their behavior is like that of a fictional sociopath.

Rules to live by: Be wary of anyone boasting their behavior is like that of a fictional sociopath.
Re: Resmed Airsense 10 Autoset stops sending data to Resmed Cloud?
Correct, the SD card has nothing to do with the transmission.greatunclebill wrote: ↑Fri Oct 19, 2018 9:01 pmthe sd card should have nothing to do with the cellular signals.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Resmed Airsense 10 Autoset stops sending data to Resmed Cloud?
If it’s an older unit it is probably still on 2G, which has turned turned off in lost of Oz.
I had mine upgraded to the 3G network at no charge thanks to Resmed.
I had mine upgraded to the 3G network at no charge thanks to Resmed.
- Okie bipap
- Posts: 3567
- Joined: Thu Oct 15, 2015 4:14 pm
- Location: Central Oklahoma
Re: Resmed Airsense 10 Autoset stops sending data to Resmed Cloud?
The phone module in the machine may have failed. My first machine would not transmit data. They gave me a new machine in order to correct the problem.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Evora Full Face Mask - Fitpack |
Additional Comments: IPAP 20-25, ps 4, OSCAR software |
Growing old is mandatory, but growing up is optional.
Re: Resmed Airsense 10 Autoset stops sending data to Resmed Cloud?
I am 99% certain my modem failed. The cell network in my area is perfect with full signal and our other CPAP is sending satay to the cloud with zero issues. My fathers CPAP just up the road from me is also transmitting data to the cloud as well.
It is remarkable at what Resmed Canada is getting my DME to do to troubleshoot and yet they don’t want to speak directly to the customer (me) to do whatever they need to figure out if they will allow my DME to issue me a new unit.
Silly to make the DME the middleman when it is obvious that the modem is likely pooched.
Oh well it isn’t life or death but just annoying that I have to jump through Resmed’s ridiculous policy.
My DME was nice enough to declare to Resmed that if there is no closure after two weeks they are going to give me a new unit if Resmed drags this out for too long.
It is remarkable at what Resmed Canada is getting my DME to do to troubleshoot and yet they don’t want to speak directly to the customer (me) to do whatever they need to figure out if they will allow my DME to issue me a new unit.
Silly to make the DME the middleman when it is obvious that the modem is likely pooched.
Oh well it isn’t life or death but just annoying that I have to jump through Resmed’s ridiculous policy.
My DME was nice enough to declare to Resmed that if there is no closure after two weeks they are going to give me a new unit if Resmed drags this out for too long.
Re: Resmed Airsense 10 Autoset stops sending data to Resmed Cloud?
Definitely not an older unit. It is a new AirSense 10 Autoset manufactured in 2018.
Re: Resmed Airsense 10 Autoset stops sending data to Resmed Cloud?
I didn’t think so either but it’s good that the folks here confirm that SD cards have nothing to do with cloud data uploads.palerider wrote: ↑Fri Oct 19, 2018 10:24 pmCorrect, the SD card has nothing to do with the transmission.greatunclebill wrote: ↑Fri Oct 19, 2018 9:01 pmthe sd card should have nothing to do with the cellular signals.
Re: Resmed Airsense 10 Autoset stops sending data to Resmed Cloud?
"the cell network" is made up of many things.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Sunnyworld
- Posts: 4
- Joined: Thu Mar 22, 2018 7:07 am
- Location: Amsterdam / Netherlands
Re: Resmed Airsense 10 Autoset stops sending data to Resmed Cloud?
lock the memorycard before you put it in the pc
-
- Posts: 1795
- Joined: Mon Mar 27, 2017 5:45 pm
- Location: Central Virginia
Re: Resmed Airsense 10 Autoset stops sending data to Resmed Cloud?
That was needed for S9 but not S10
_________________
Machine: AirSense 11 Autoset |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software |
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.
Re: Resmed Airsense 10 Autoset stops sending data to Resmed Cloud?
Unnecessary, *unless* you have a S9, and a mac or windows 8.1+
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Resmed Airsense 10 Autoset stops sending data to Resmed Cloud?
Ah only s9s are affected by hidden index files. Gotcha. Good to know!
It turned out that the cellular modem module in my machine e bit the dust. Managed to get a brand new Resmed AirSense 10 Autoset for her
It turned out that the cellular modem module in my machine e bit the dust. Managed to get a brand new Resmed AirSense 10 Autoset for her

- Okie bipap
- Posts: 3567
- Joined: Thu Oct 15, 2015 4:14 pm
- Location: Central Oklahoma
Re: Resmed Airsense 10 Autoset stops sending data to Resmed Cloud?
Glad to hear you got it straightened out.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Evora Full Face Mask - Fitpack |
Additional Comments: IPAP 20-25, ps 4, OSCAR software |
Growing old is mandatory, but growing up is optional.