Question for ASV users

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
SirGaspAlot
Posts: 146
Joined: Mon Jul 16, 2018 7:33 am
Location: Maryland, United States

Re: Question for ASV users

Post by SirGaspAlot » Sun Oct 14, 2018 9:19 am

What would you recommend? Same as before with my Bipap- no ramp time and set pressure to 6? Or is more involved?
Pugsy wrote: ↑
Sun Oct 14, 2018 8:17 am
My first suspect would be PS max limit might be a suspect.
There's not much range in the PS and if it will only go to 6 and max EPAP is 10 the the most IPAP you can get is 12 (when PS is 4) or up to 18 (when PS is 12).
Second suspect would be max EPAP tying the machine's little hands so it can deal with the obstructives well.

See this document
http://www.isetonline.org/yahoo_site_ad ... 190318.pdf

From your report...with the snores...I tend to lean towards obstructive for those hyponeas because we don't usually snore with centrals.

Did they put you on this because of centrals? PS of 6 isn't going to do much for centrals is why I asked.
US Navy Retired E6, F-18 Aircraft Inspector/ejection seat inspector/mechanic and all around human being.

"If you always do what youve always done then youll always get what you always got"

User avatar
Pugsy
Posts: 65444
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Question for ASV users

Post by Pugsy » Sun Oct 14, 2018 9:32 am

If you were getting events on the BiPap at EPAP min of 6 that were obstructive in nature...you are going to need more EPAP on the ASV most likely. While it will respond a little faster for the obstructive stuff than the regular bipap could...it can't start at 4 and make up the difference fast enough especially with restrictions in place.

Don't rely on what SH is saying the settings are ....get the settings off the machine itself. There's a bug in SH when it comes to settings sometimes.
Settings:
Max Pressure: 25 cmH20
MIN EPAP: 4 cmH20
MAX EPAP: 10 cmH20
MIN PRESS SUPPORT: 4 cmH20, MAX: 6 cmH20
BREATH RATE: AUTO
I would change minimum EPAP to 6 (more in tune with your prior machine)
Max EPAP...15....the machine will only go there if it feels the need.
Minimum and maximum pressure support...4 min and 15 max...again it won't go where it doesn't feel the need to go but you need more than 6 for central treating.

Leave it on Auto Breath rate.

Essentially give the machine more room to work and see how it handles things. You might still need more EPAP minimum but right now with the way the restrictive settings are working it's really hard to know for sure.
I may have to RISE but I refuse to SHINE.

User avatar
SirGaspAlot
Posts: 146
Joined: Mon Jul 16, 2018 7:33 am
Location: Maryland, United States

Re: Question for ASV users

Post by SirGaspAlot » Sun Oct 14, 2018 9:40 am

This is what the Sleepyhead says my machine settings are. Not so sure Encore is accurate about those settings.
Attachments
Machine settings report from sleepyhead.jpg
Machine settings report from sleepyhead.jpg (556.12 KiB) Viewed 1198 times
US Navy Retired E6, F-18 Aircraft Inspector/ejection seat inspector/mechanic and all around human being.

"If you always do what youve always done then youll always get what you always got"

User avatar
SirGaspAlot
Posts: 146
Joined: Mon Jul 16, 2018 7:33 am
Location: Maryland, United States

Re: Question for ASV users

Post by SirGaspAlot » Sun Oct 14, 2018 9:52 am

I think its the Sleepyhead reading settings wrong
US Navy Retired E6, F-18 Aircraft Inspector/ejection seat inspector/mechanic and all around human being.

"If you always do what youve always done then youll always get what you always got"

User avatar
Pugsy
Posts: 65444
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Question for ASV users

Post by Pugsy » Sun Oct 14, 2018 9:56 am

Encore Settings I would believe.
SleepyHead...not so much especially with this DreamStation model. Mark just didn't have enough DreamStation ASV data files to sort through IMHO to get it right and the version you are using wasn't even supposed to be bug free. It was a testing version anyway.

I would still prefer to get them straight from the machine itself though.
I may have to RISE but I refuse to SHINE.

User avatar
SirGaspAlot
Posts: 146
Joined: Mon Jul 16, 2018 7:33 am
Location: Maryland, United States

Re: Question for ASV users

Post by SirGaspAlot » Sun Oct 14, 2018 10:05 am

Gotcha. Stepped right past that thought. Ok, all set. I deleted the ramp time and set ramp start to 6 like before. Will see how it goes. Thank you as always.
US Navy Retired E6, F-18 Aircraft Inspector/ejection seat inspector/mechanic and all around human being.

"If you always do what youve always done then youll always get what you always got"

User avatar
SirGaspAlot
Posts: 146
Joined: Mon Jul 16, 2018 7:33 am
Location: Maryland, United States

Re: Question for ASV users

Post by SirGaspAlot » Sun Oct 14, 2018 10:07 am

Turns out Encore was the more accurate one.
US Navy Retired E6, F-18 Aircraft Inspector/ejection seat inspector/mechanic and all around human being.

"If you always do what youve always done then youll always get what you always got"

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: Question for ASV users

Post by Jas_williams » Sun Oct 14, 2018 11:04 am

The graphs for the ASV on sleepyhead I would believe but not the settings, Mark dropped off grid before they were properly decoded. Your graphs show the settings you stated earlier in the thread nothing like the settings shown in sleepyhead.

User avatar
SirGaspAlot
Posts: 146
Joined: Mon Jul 16, 2018 7:33 am
Location: Maryland, United States

Re: Question for ASV users

Post by SirGaspAlot » Sun Oct 14, 2018 12:18 pm

I see this. Always an education to be had. I do vaguely recall the nurse telling me that its their policy to stay with a ramp up from 4 cmH20 accross the board, perhaps a safety buffer. Theres a lot to risk in that profession I can imagine. However, no complaint was raised about my modifying the settings on the Bipap machine as I expect will be the same with ASV machine as well.
Jas_williams wrote: ↑
Sun Oct 14, 2018 11:04 am
The graphs for the ASV on sleepyhead I would believe but not the settings, Mark dropped off grid before they were properly decoded. Your graphs show the settings you stated earlier in the thread nothing like the settings shown in sleepyhead.
US Navy Retired E6, F-18 Aircraft Inspector/ejection seat inspector/mechanic and all around human being.

"If you always do what youve always done then youll always get what you always got"

User avatar
Pugsy
Posts: 65444
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Question for ASV users

Post by Pugsy » Sun Oct 14, 2018 1:25 pm

Ramp is strictly optional starting point to start at a lower minimum pressure and take a little time to work up to the real minimum pressure.
Ramp is only engaged IF you push the ramp button. Doesn't matter what ramp is set to if you don't push the ramp button.
Default ramp minimum is 4 cm.

I think she meant starting minimum EPAP when she said "ramp" since that is what they set the minimum pressure at which is essentially also the factory default for minimum EPAP. Don't understand that the minimum of 4 doesn't always give the machine a good enough head start when the airway starts to collapse. Works the same on ASV as with regular apap minimum pressure settings.

The actual ramp function itself was useless at the original settings even if you even did push the button to use ramp because the machine already was starting at the lowest it could go (that 4 cm minimum EPAP) anyway.

Ramp refers to the special time frame where the machine starts low and moves up to the minimum EPAP setting taking however long a person sets ramp time to. Once the minimum EPAP setting is reached then the ramp term isn't used and instead therapy pressure increase term is used. People will often say "ramping up" but it's not actually "ramping up" but instead simply responding to whatever the machine senses needs to be responded to.
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Question for ASV users

Post by palerider » Sun Oct 14, 2018 4:02 pm

SirGaspAlot wrote: ↑
Sun Oct 14, 2018 12:18 pm
their policy to stay with a ramp up from 4 cmH20 accross the board, perhaps a safety buffer.
Absolutely nothing 'safety' related about ramp, or a setting of 4. Just ignorance or rigidity on their part.

Many people feel suffocated at a pressure of 4, others are fine with it, and a very few, that's all they need.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
SirGaspAlot
Posts: 146
Joined: Mon Jul 16, 2018 7:33 am
Location: Maryland, United States

Re: Question for ASV users

Post by SirGaspAlot » Sun Oct 14, 2018 5:00 pm

I'm not certain which. However, the changes have been made, will update soon.
palerider wrote: ↑
Sun Oct 14, 2018 4:02 pm
SirGaspAlot wrote: ↑
Sun Oct 14, 2018 12:18 pm
their policy to stay with a ramp up from 4 cmH20 accross the board, perhaps a safety buffer.
Absolutely nothing 'safety' related about ramp, or a setting of 4. Just ignorance or rigidity on their part.

Many people feel suffocated at a pressure of 4, others are fine with it, and a very few, that's all they need.
US Navy Retired E6, F-18 Aircraft Inspector/ejection seat inspector/mechanic and all around human being.

"If you always do what youve always done then youll always get what you always got"