Gotta Be A Typo
84 arousals in 6.5 minutes would be an arousal index of 775.4, that's getting up there.
By definition, the shortest arousal is 3 seconds, and the shortest period of intervening sleep must be 10 seconds, so theoretically, the most REM arousals you could have in that period is 30.
SAG
By definition, the shortest arousal is 3 seconds, and the shortest period of intervening sleep must be 10 seconds, so theoretically, the most REM arousals you could have in that period is 30.
SAG

Aromatherapy may help CPAP compliance. Lavender, Mandarin, Chamomile, and Sweet Marjoram aid in relaxation and sleep. Nature's Gift has these and a blend of all four called SleepEase.
-
SleepySandy
- Posts: 330
- Joined: Sat Jan 13, 2007 5:25 pm
- Location: Seattle, WA
Not really a typo - more like a big giant error
A typo is an understatement.
I reversed all of the REM and NREM numbers.
I had 84 NREM arousals / 8 REM arousals.
I hope you all will forgive my great big mistake and still help me with this.
I have another post nearly ready - it lists all of my data from both studies and should be correct.
I reversed all of the REM and NREM numbers.
I had 84 NREM arousals / 8 REM arousals.
I hope you all will forgive my great big mistake and still help me with this.
I have another post nearly ready - it lists all of my data from both studies and should be correct.
-
SleepySandy
- Posts: 330
- Joined: Sat Jan 13, 2007 5:25 pm
- Location: Seattle, WA
I apologize again for the big mistake I made in my initial post. I'm hoping I didn't do myself in on this forum because I do still need help figuring this out.
My fatigue is completely debilitating. I'm much more tired than I was before my first sleep study. At this point, I am so tired I can't focus on much of anything or, apparently, accurately type sleep study results. I work part-time from home writing software requirement documents. I cannot focus enough to get anything done. This is completely unlike me. I've always been an overachiever and I've always been able to push through my fatigue. Now, I stare at my computer and wish I could go to sleep. It's not just my work that's affected. I don't make plans because I don't know if I'll be too tired to do them. My husband taking care of most of the housework. I'm extremely frustrated.
I'm would really appreciate any additional thoughts on why I am still so completely exhausted.
Kathy suggested I look at my study data side by side, but I'm not sure how similar / different certain things should be. I thought I would post the side by side information here (this includes more information from my titration study than my first post did and I just double-checked the REM / NREM numbers).
That's all I can think of.
Any additional insights are appreciated.
Sandy
My fatigue is completely debilitating. I'm much more tired than I was before my first sleep study. At this point, I am so tired I can't focus on much of anything or, apparently, accurately type sleep study results. I work part-time from home writing software requirement documents. I cannot focus enough to get anything done. This is completely unlike me. I've always been an overachiever and I've always been able to push through my fatigue. Now, I stare at my computer and wish I could go to sleep. It's not just my work that's affected. I don't make plans because I don't know if I'll be too tired to do them. My husband taking care of most of the housework. I'm extremely frustrated.
I'm would really appreciate any additional thoughts on why I am still so completely exhausted.
Kathy suggested I look at my study data side by side, but I'm not sure how similar / different certain things should be. I thought I would post the side by side information here (this includes more information from my titration study than my first post did and I just double-checked the REM / NREM numbers).
Code: Select all
Initial Titration
Sleep Summary
Lights Out 21:30:34 22:17:34
Lights On 05:59:04 05:04:04
Total Study Time 552.1 415.5
Time in Bed 508.5 406.5
Sleep Period Time 494 301
Total Sleep Time 430 251
Total Wake Time 78.5 155
Sleep Efficiency 84.6% 61.7%
Sleep maint. Efficiency 87.0% 83.4%
Sleep Latency 14 52
Latency to REM 215.5 289.5
Sleep Stage Summary
WASO Duration (min) 65 50
WASO % SPT 13.2 16.6
WASO Latency (min) 2 75.5
Stage 1 Duration (min) 85 46
Stage 1 % SPT 17.2 15.3
Stage 1 Latency (min) 7.5 52
Stage 2 Duration (min) 153.5 134
Stage 2 % SPT 31.1 44.5
Stage 2 Latency (min) 23 58.5
Delta Duration (min) 108.5 64.5
Delta % SPT 22 21.4
Delta Latency (min) 35.5 24
Total NREM Duration (min) 347 244.5
Total NREM % SPT 70.2 81.2
REM Duration (min) 83 6.5
REM % SPT 16.8 2.2
REM Latency (min) 215.5 289.5
Respiratory Summary
AHI 14.1 0.7
Total # Apneas 34 3
Total # Hypopneas 67 0
Total # Respiratory Events 101 3
Mean SaO2 % 97 97
# Desats 77 0
Lowest SaO2 desat % 91 0
Limb Movement Summary
Total # LMs 108 29
LM Index 15.1 6.9
Total # PLMs 196 91
PLM Index 27.3 21.8
Total # PLMS w/arousal 22 6
PLMs w/arousal Index 3.8 1.5
Total # RRLMs 24 0
RRLMs Index 3.3 0
Sleep Continuity: NREM Arousals
Spontaneous 6 15
Spontaneous Index 1.0 3.7
Resp Events 58 3
Resp Events Index 10.0 0.7
RERAs / Snoring 272 46
RERAs / Snoring Index 47.0 11.3
Desaturation 0 0
Desaturation Index 0.0 0.0
Limb Movement 16 14
Limb Movement Index 2.8 3.4
PLMs 22 6
PLMs Index 3.8 1.5
RRLM 0 0
RRLM Index 0.0 0.0
Total Arousals 374 84
Total Arousal Index 64.7 20.6
Sleep Continuity: REM Arousals
Spontaneous 0 3
Spontaneous Index 0.0 27.7
Resp Events 43 0
Resp Events Index 31.1 0.0
RERAs / Snoring 40 4
RERAs / Snoring Index 28.9 36.9
Desaturation 0 0
Desaturation Index 0.0 0.0
Limb Movement 3 1
Limb Movement Index 2.2 9.2
PLMs 2 0
PLMs Index 1.4 0.0
RRLM 0 0
RRLM Index 0.0 0.0
Total Arousals 88 8
Total Arousal Index 63.6 73.8
Sleep Continuity: Total Arousals
Spontaneous 6 18
Spontaneous Index 0.8 4.3
Resp Events 101 3
Resp Events Index 14.1 0.7
RERAs / Snoring 312 50
RERAs / Snoring Index 43.5 12.0
Desaturation 0 0
Desaturation Index 0.0 0.0
Limb Movement 19 15
Limb Movement Index 2.7 3.6
PLMs 24 6
PLMs Index 3.3 1.4
RRLM 0 0
RRLM Index 0.0 0.0
Total Arousals 462 92
Total Arousal Index 64.5 22.0- Interpretation for Initial Study:
Sleep architecture was severely fragmented with a total of 462 arousals recorded. REM sleep was slightly reduced. Snoring was continuous and moderate in severity. Apnea-hypopnea index was moderately elevated at 14 events / h (normal < 5) yet what was most impressive was the frequency of relative airflow limitation that was terminated by an arousal/snore as is seen with upper airway resistance syndrome (UARS). Periodic limb movements were frequent as well totaling 196 or 27 per hour yet are likely secondary to the patient's sleep disordered breathing.
Interpretation for Titration Study
1. Baseline moderate OSA with AHI 14
2. Average AHI on CPAP 0.7
3. Improved sleep architecture from baseline with reduction in total arousals from 462 to 92.
4. Optimal CPAP at 10 cwp
That's all I can think of.
Any additional insights are appreciated.
Sandy
Hey Sandy,
The inability to concentrate properly is one of the symtoms of OSA, and you'll notice lots of mistakes in posts here. You are not writing a thesis, just asking for help, and we all make mistakes. This is a pretty understanding community.
Good luck as the knowledgeable guys chime in to help you out.
Peter
The inability to concentrate properly is one of the symtoms of OSA, and you'll notice lots of mistakes in posts here. You are not writing a thesis, just asking for help, and we all make mistakes. This is a pretty understanding community.
Good luck as the knowledgeable guys chime in to help you out.
Peter

_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear |
| Additional Comments: SleepHQ |
- StillAnotherGuest
- Posts: 1005
- Joined: Sun Sep 24, 2006 6:43 pm
Well, To Start...
I would also suggest looking at the medications you're on, the REM distribution certainly suggests a medication effect. The graphs might be helpful if you can get them.
SAG
SAG

Aromatherapy may help CPAP compliance. Lavender, Mandarin, Chamomile, and Sweet Marjoram aid in relaxation and sleep. Nature's Gift has these and a blend of all four called SleepEase.
one thing that caught my eye was the line about 77 destats
& the lowest desat being 91% ?
My 1st reaction is that if the Spo2 is over 90, surely that isn't a desat. This leads me to ask the experts who know, just what constitutes a desat ?
DSM
& the lowest desat being 91% ?
My 1st reaction is that if the Spo2 is over 90, surely that isn't a desat. This leads me to ask the experts who know, just what constitutes a desat ?
DSM
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)
-
SleepySandy
- Posts: 330
- Joined: Sat Jan 13, 2007 5:25 pm
- Location: Seattle, WA
Re: Question for the techs out there
Obviously the PLM's with arousals are the important leg movements to take into consideration. It does happen at times where a leg movement and airflow reduction happen close to the same time, followed immediately by an arousal (and if any other techs on here read this I'm sure they'll agree), it really depends on the tech thats scoring that study. One thing to look at is if the Leg kick happened slightly before the arousal, in such a case the tech may score it as a PLM w/ arousal. Though more often than not it would be scored as a repiratory event w/ arousal.kteague wrote:Been dealing with this stuff for ten years as a patient, but still so much to learn. Can you help me understand how the tests interpret a situation where an arousal is recorded when an obstructive event is happening simultaneously to a PLMD movement? I'm speaking of times when either would have happened independently in the other's absence. How does the data reveal which caused the arousal? Are these by default counted as respiratory arousals?
Kathy
There are really so many things to take into consideration, if the patient has alot of PLM's they will pop out, so scoring a resp. event over a PLM wouldn't generally sway it too much. Also you have to consider if the Patient is on their diagnostic study or on their CPAP study. If they are getting titrated, those resp. events will eventually disappear (or atleast thats the goal), leaving those pesky leg movements w/ arousals sticking out like a sore thumb. Though the tech may not realize the severity of the leg kicks in time to medicate the patient that night (due to resp. events covering them up), the proof will be there for the doctor to take leg medication into consideration.
So I guess to answer your question, its truly up to how that perticular tech wants to score that. More often than not though, the respiratory event is treated first, and if the patient has PLMD it will stand out in the end, hopefully in time for the tech to treat it that night.
Chris
_________________
CPAPopedia Keywords Contained In This Post (Click For Definition): Arousal, CPAP
Sleep Tech
Virginia
Virginia
A scorable desaturation is a 4% drop in saturation. So really we could score a desat in a drop from 98% to 94%. Various labs may use a different time line, we generally will score a desat if it falls within 30 seconds.dsm wrote:one thing that caught my eye was the line about 77 destats
& the lowest desat being 91% ?
My 1st reaction is that if the Spo2 is over 90, surely that isn't a desat. This leads me to ask the experts who know, just what constitutes a desat ?
DSM
The reason we track the low saturation is for the purpose of possibly adding O2 for the patient. There are qualifications for adding O2 depending on the situation.
But to answer your question, just a 4% drop in sats qualifies a scorable desaturation.
Sleep Tech
Virginia
Virginia
I need to learn how y'all quote several times in one reply.....lol
Sandy,
In taking a quick look (I dont have a ton of time at the moment). BTW, nice catch on the REM to NREM arousals SAG....that shoulda stuck out like a sore thumb.....ah well.
I noticed your spontaneous arousals grew considerably in most every category. That could be attributed to over titrating at a certain time throughout the night, if that is that case hopefully the final pressure you've been given eliminates all or most of those spontaneous arousals. Another thing to consider is you may be getting spontaneous arousals (especially in REM) from mouth leaks. I noticed your using an internasal mask, and if your getting even a slight mouth leak while your sleeping that could cause an arousal. I really suggest using a chin strap....I know the dreaded chin strap, but if mouth leaks are the issue it can help. Though I will say a chin strap wont always cure the leaks.
Just a thought. As I said before, a study can look alot worse than it really is, because the tech is literally playing around with different pressures trying to find what works in every position and well as every stage of sleep. The unfortunate thing is when someone is REM dominant or has a positional component, because then you'll see someone that is fine at 5cm of pressure when lateral in Non-REM but will need 17cm of pressure when supine in REM.....its happened.
As far as SAG's suggestion on looking deeper into your medication, it could be a thought, but I'll leave him to give you advice in that category.
OH...also somone else (cant remember) brought up a valid point about caffeine. This is very true, sleep hygiene is very important. Make a nice quiet, dark room for you to sleep in.......well quiet outside of the machine anyways ;-P
Try to avoid anything caffeinated.....even chocolate within a few hours of sleep (4 hours if possible). In fact just avoid eating within a few hours of sleeping, cause if your body is trying to digest food while your sleeping....guess whats happening to your quality of sleep.
Chris
_________________
CPAPopedia Keywords Contained In This Post (Click For Definition): Arousal
Sandy,
In taking a quick look (I dont have a ton of time at the moment). BTW, nice catch on the REM to NREM arousals SAG....that shoulda stuck out like a sore thumb.....ah well.
I noticed your spontaneous arousals grew considerably in most every category. That could be attributed to over titrating at a certain time throughout the night, if that is that case hopefully the final pressure you've been given eliminates all or most of those spontaneous arousals. Another thing to consider is you may be getting spontaneous arousals (especially in REM) from mouth leaks. I noticed your using an internasal mask, and if your getting even a slight mouth leak while your sleeping that could cause an arousal. I really suggest using a chin strap....I know the dreaded chin strap, but if mouth leaks are the issue it can help. Though I will say a chin strap wont always cure the leaks.
Just a thought. As I said before, a study can look alot worse than it really is, because the tech is literally playing around with different pressures trying to find what works in every position and well as every stage of sleep. The unfortunate thing is when someone is REM dominant or has a positional component, because then you'll see someone that is fine at 5cm of pressure when lateral in Non-REM but will need 17cm of pressure when supine in REM.....its happened.
As far as SAG's suggestion on looking deeper into your medication, it could be a thought, but I'll leave him to give you advice in that category.
OH...also somone else (cant remember) brought up a valid point about caffeine. This is very true, sleep hygiene is very important. Make a nice quiet, dark room for you to sleep in.......well quiet outside of the machine anyways ;-P
Try to avoid anything caffeinated.....even chocolate within a few hours of sleep (4 hours if possible). In fact just avoid eating within a few hours of sleeping, cause if your body is trying to digest food while your sleeping....guess whats happening to your quality of sleep.
Chris
_________________
CPAPopedia Keywords Contained In This Post (Click For Definition): Arousal
Sleep Tech
Virginia
Virginia
-
SleepySandy
- Posts: 330
- Joined: Sat Jan 13, 2007 5:25 pm
- Location: Seattle, WA
Thanks again, everyone. So much information...
The practical tips are excellent suggestions, particularly the medication one. I have several medications I take each morning, but they're things I really won't be able to stop. BUT...I do have a medication I take at bedtime. I had never thought about it possibly affecting my sleep. Now that I've thought about it, I can definitely see how it might be. It's a divided dose (1/2 morning and 1/2 night), so I can't move the night dose to the morning. However, I think I can move it to late afternoon / evening. I'm going to give that a try.
I think I'm good about caffeine. I can't stand coffee or tea, so my only source of caffeine is soda. I do drink 2 sodas a day and always before 4pm. Since I don't drink it late in the day, I'm hoping it's not contributing to my problems. I don't really want to give it up completely yet. I need my Pepsi One!
The food thing is something else I'm going to change. I tend to not eat enough for dinner, which leaves me hungry closer to bedtime. I can't fall asleep if I'm hungry, so I have something before bed. It's always something like a banana or peanut butter toast, but I can see how that just complicates things for me.
I'm glad to have some things I can do while waiting for the blood test results (which won't show anything anyway) and my next sleep doc appointment.
Sleep is amazingly complicated.
The practical tips are excellent suggestions, particularly the medication one. I have several medications I take each morning, but they're things I really won't be able to stop. BUT...I do have a medication I take at bedtime. I had never thought about it possibly affecting my sleep. Now that I've thought about it, I can definitely see how it might be. It's a divided dose (1/2 morning and 1/2 night), so I can't move the night dose to the morning. However, I think I can move it to late afternoon / evening. I'm going to give that a try.
I think I'm good about caffeine. I can't stand coffee or tea, so my only source of caffeine is soda. I do drink 2 sodas a day and always before 4pm. Since I don't drink it late in the day, I'm hoping it's not contributing to my problems. I don't really want to give it up completely yet. I need my Pepsi One!
The food thing is something else I'm going to change. I tend to not eat enough for dinner, which leaves me hungry closer to bedtime. I can't fall asleep if I'm hungry, so I have something before bed. It's always something like a banana or peanut butter toast, but I can see how that just complicates things for me.
I'm glad to have some things I can do while waiting for the blood test results (which won't show anything anyway) and my next sleep doc appointment.
Sleep is amazingly complicated.