High AHI, Flat Pressure

Post by aherman18 » Fri Sep 28, 2018 8:44 am

AHI seems high, yet machine is not increasing pressure. Min set for 9, Max 14.4. However it never increases much above 11.

Any thoughts would be appreciated.
screenshot-20180928-103123.png
screenshot-20180928-103123.png (105.93 KiB) Viewed 809 times

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: High AHI, Flat Pressure

Post by palerider » Fri Sep 28, 2018 11:13 am

It's maxing at 13.3.

Turn your min to 11, max 20. see what happens.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

longwood
Posts: 14
Joined: Wed Sep 12, 2018 3:27 pm

Re: High AHI, Flat Pressure

Post by longwood » Fri Sep 28, 2018 6:08 pm

Very strange to me, I am wondering if your CPAP is broken.?

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: High AHI, Flat Pressure

Post by Wulfman... » Fri Sep 28, 2018 6:12 pm

aherman18 wrote:
Fri Sep 28, 2018 8:44 am
AHI seems high, yet machine is not increasing pressure. Min set for 9, Max 14.4. However it never increases much above 11.

Any thoughts would be appreciated.

Which specific model of machine do you have?
Do you have the "For Her" model?


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

aherman18
Posts: 21
Joined: Thu Mar 22, 2018 8:17 am

Re: High AHI, Flat Pressure

Post by aherman18 » Fri Sep 28, 2018 6:20 pm

I have a regular Resmed 10. I'll try upping the lower and upper pressure.

We will see.

Thanks for the help.

Andrew

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: High AHI, Flat Pressure

Post by Julie » Sat Sep 29, 2018 2:29 am

Many leave the max setting quite high, even leaving it at 20, because it's the min. setting that does the work and capping the max one so close to it (as you do) can be an unnecessary problem.