Aerophagia disturbing my sleep

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
SarahB
Posts: 98
Joined: Thu Jan 18, 2018 1:09 pm
Location: Scotland

Aerophagia disturbing my sleep

Post by SarahB » Fri Sep 28, 2018 1:15 am

The last couple of nights I've been aware throughout the night that I'm constantly trying to burp up lots of air and in the morning I'm still quite full of air. I'm even trying to get rid of excess air in my dreams! I feel that I'm fighting to get the extra air back out faster than it's coming in.

I'm using an auto CPAP (ResMed Airsense 10) and a nasal mask. I tape my mouth shut and there are no problems with leaks. My pressures haven't changed (max is around 13), my AHI is fine (0.00 last night, generally below 0.5) and my sleep position is the same. I've been using CPAP for seven months. I can't work out why this should now start to be a problem.
Resmed AirSense 10 Autoset. Mask: Resmed Airfit N20 nasal mask

User avatar
Julie
Posts: 20041
Joined: Tue Feb 28, 2006 12:58 pm

Re: Aerophagia disturbing my sleep

Post by Julie » Fri Sep 28, 2018 4:33 am

Can you try lowering the min. pressure by .5 for a couple of nights each to see if it helps (often does) and you can 'afford' it with your very low AHI. Also, you could raise the head end of your bed on some kind of 4-5" blocks which also apparently helps.

nicholasjh1
Posts: 517
Joined: Wed Jun 14, 2017 9:07 am

Re: Aerophagia disturbing my sleep

Post by nicholasjh1 » Fri Sep 28, 2018 2:44 pm

I get issues with aerophagia unless I use my chin strap and the strap has to be tight enough. Not only do I get aerophagia if my mouth opens too much, but I get a lot of mouth leak (so chip munk cheeks since I tape). So if everything else fails you might want to try that.
Instead of Sleep apnea it should be called "Sleep deprivation, starving of oxygen, being poisoned by high CO2 levels, damaging the body and brain while it's supposed to be healing so that you constantly get worse and can never get healthy Apnea"

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Aerophagia disturbing my sleep

Post by palerider » Fri Sep 28, 2018 3:02 pm

nicholasjh1 wrote:
Fri Sep 28, 2018 2:44 pm
I get issues with aerophagia unless I use my chin strap and the strap has to be tight enough. Not only do I get aerophagia if my mouth opens too much, but I get a lot of mouth leak (so chip munk cheeks since I tape). So if everything else fails you might want to try that.
I don't understand how you can have air forced into your stomach at the same time it's venting out your mouth...

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Aerophagia disturbing my sleep

Post by chunkyfrog » Fri Sep 28, 2018 3:27 pm

When I get gas, it would be easy to blame it on cpap.
But I know better.
It is what I EAT.
It is as simple as that.
"Aerophagia" isn't always what we think it is.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

SarahB
Posts: 98
Joined: Thu Jan 18, 2018 1:09 pm
Location: Scotland

Re: Aerophagia disturbing my sleep

Post by SarahB » Sat Sep 29, 2018 11:35 am

Julie wrote:
Fri Sep 28, 2018 4:33 am
Can you try lowering the min. pressure by .5 for a couple of nights each to see if it helps (often does) and you can 'afford' it with your very low AHI. Also, you could raise the head end of your bed on some kind of 4-5" blocks which also apparently helps.
Minimum pressure is 4 but I set it to 6 because I feel a bit short of air at 4. I could try playing around with it. Interesting idea about raising the head of the bed, thanks.
Resmed AirSense 10 Autoset. Mask: Resmed Airfit N20 nasal mask

SarahB
Posts: 98
Joined: Thu Jan 18, 2018 1:09 pm
Location: Scotland

Re: Aerophagia disturbing my sleep

Post by SarahB » Sat Sep 29, 2018 11:37 am

chunkyfrog wrote:
Fri Sep 28, 2018 3:27 pm
When I get gas, it would be easy to blame it on cpap.
But I know better.
It is what I EAT.
It is as simple as that.
"Aerophagia" isn't always what we think it is.
:o
Resmed AirSense 10 Autoset. Mask: Resmed Airfit N20 nasal mask

SarahB
Posts: 98
Joined: Thu Jan 18, 2018 1:09 pm
Location: Scotland

Re: Aerophagia disturbing my sleep

Post by SarahB » Sat Sep 29, 2018 11:38 am

nicholasjh1 wrote:
Fri Sep 28, 2018 2:44 pm
I get issues with aerophagia unless I use my chin strap and the strap has to be tight enough. Not only do I get aerophagia if my mouth opens too much, but I get a lot of mouth leak (so chip munk cheeks since I tape). So if everything else fails you might want to try that.
I already tape so it's not getting in that way.
Resmed AirSense 10 Autoset. Mask: Resmed Airfit N20 nasal mask

User avatar
Julie
Posts: 20041
Joined: Tue Feb 28, 2006 12:58 pm

Re: Aerophagia disturbing my sleep

Post by Julie » Sat Sep 29, 2018 4:17 pm

Had no idea you were at such a low pressure - do you have no effort inhaling (as so many do at that setting)?

Better to look at your diet as has been suggested, and maybe try to raise the bed.

SarahB
Posts: 98
Joined: Thu Jan 18, 2018 1:09 pm
Location: Scotland

Re: Aerophagia disturbing my sleep

Post by SarahB » Sun Sep 30, 2018 1:28 am

Julie wrote:
Sat Sep 29, 2018 4:17 pm
Had no idea you were at such a low pressure - do you have no effort inhaling (as so many do at that setting)?

Better to look at your diet as has been suggested, and maybe try to raise the bed.
The pressure range on the machine is 4-20 by default but I got into the clinician's settings and altered that to 6-20 as 4 was uncomfortably low. The moment I fall asleep the pressure shoots up and my average is around 9 or 10, going up to a maximum of about 13. Occasionally it goes higher but anything over about 17 wakes me up.
Resmed AirSense 10 Autoset. Mask: Resmed Airfit N20 nasal mask