Numbers have gone off the rails!

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
1sleepyhubby
Posts: 22
Joined: Sun Mar 12, 2017 8:10 pm

Numbers have gone off the rails!

Post by 1sleepyhubby » Sun Sep 23, 2018 4:58 pm

Hi everyone! My husband has been using a cpap now for about 3.5 years. He had his AHI pretty under control until about 6 months ago, and then it slowly has started creeping up. At this point, he's averaging about a 5.5 in the couple weeks according to Sleepyhead. We're not sure what's really changed for him to cause this. He uses a dreamstation apap with an amara view mask. Is there anything we could possibly adjust or tweak so he could get some better rest? He's really feeling it as it keeps dragging on night after night. TIA!!!

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Numbers have gone off the rails!

Post by palerider » Sun Sep 23, 2018 5:38 pm

1sleepyhubby wrote:
Sun Sep 23, 2018 4:58 pm
Hi everyone! My husband has been using a cpap now for about 3.5 years. He had his AHI pretty under control until about 6 months ago, and then it slowly has started creeping up. At this point, he's averaging about a 5.5 in the couple weeks according to Sleepyhead. We're not sure what's really changed for him to cause this. He uses a dreamstation apap with an amara view mask. Is there anything we could possibly adjust or tweak so he could get some better rest? He's really feeling it as it keeps dragging on night after night. TIA!!!
Please read the sticky post for new users, we need information to be able to offer useful suggestions... wild guesses won't help him.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

1sleepyhubby
Posts: 22
Joined: Sun Mar 12, 2017 8:10 pm

Re: Numbers have gone off the rails!

Post by 1sleepyhubby » Sun Sep 23, 2018 5:41 pm

Hope I'm doing this right! Tried to copy what I saw in Pugsy's tips post.

0923cpap.jpg
0923cpap.jpg (505.42 KiB) Viewed 605 times

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Numbers have gone off the rails!

Post by Julie » Sun Sep 23, 2018 5:53 pm

Any chance he's (slowly) gained even a little weight? It's amazing what a diff. it can make.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Numbers have gone off the rails!

Post by palerider » Sun Sep 23, 2018 6:38 pm

We don't need the resp rate chart, it just causes the good ones to be squished more.

Try raising the minimum to 12.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

1sleepyhubby
Posts: 22
Joined: Sun Mar 12, 2017 8:10 pm

Re: Numbers have gone off the rails!

Post by 1sleepyhubby » Mon Sep 24, 2018 5:04 pm

Julie wrote:
Sun Sep 23, 2018 5:53 pm
Any chance he's (slowly) gained even a little weight? It's amazing what a diff. it can make.
We've actually been on a diet and trying to lose weight, so he doesn't appear that he's gained any...maybe come down a few pounds even.