Success Finally But Why Don't I Feel Better?
-
lwieland11
- Posts: 249
- Joined: Sat Dec 10, 2016 5:57 am
Success Finally But Why Don't I Feel Better?
I've been doing well with the BPAP...low AHI(around 5), long sleep time, little to no leaks, no aerophagia. For example, my MyAir score last night was 95/100. AHI was 5.5. Sleep with mask on for over 8 hours. This has the typical scenario lately. I'm encouraged by the numbers. So why don't I feel amazing? I know how it feels when CPAP/BPAP works. That's not how I'm feeling. I don't look or feel good. No clarity of thought. Brain fog. No burst in energy. Could this be diet related? Thoughts? Thanks.
Re: Success Finally But Why Don't I Feel Better?
Wish you had attached this to the last notes you wrote so we could see what's already gone on... so not having them to compare to what I see is possibly longish sleep hrs (as you might with not-well treated apnea) and what we consider to be a high-ish AHI for Cpap use (5 is an arbitrary no. docs use as some kind of cutoff, but has been shown to be inadequate for well treated apnea.
Don't remember your settings, so can only suggest raising the low one a bit... and not depending on My Air to tell you much - I'm sure Sleepyhead must have been suggested long ago and will show you much more detail. I would not call what you have a 'success' esp. if you only go by the numbers... not if you still don't feel better.
Don't remember your settings, so can only suggest raising the low one a bit... and not depending on My Air to tell you much - I'm sure Sleepyhead must have been suggested long ago and will show you much more detail. I would not call what you have a 'success' esp. if you only go by the numbers... not if you still don't feel better.
-
lwieland11
- Posts: 249
- Joined: Sat Dec 10, 2016 5:57 am
Re: Success Finally But Why Don't I Feel Better?
I have tried several times unsuccessfully to post Sleepyhead results because I have a Mac.
Re: Success Finally But Why Don't I Feel Better?
So do I ... and SH was written on a Mac. You shouldn't have a problem... look at the instructions. There are people here who can help with that (I have DL'd the program but don't use it as I don't feel the need - very consistent sleep and results - my machine gives enough info using Cpap.com's website prod. pg for it where you enter screen data, that I don't bother).
-
freetimecreations
- Posts: 145
- Joined: Thu May 24, 2012 9:02 pm
- Location: Utah, USA
Re: Success Finally But Why Don't I Feel Better?
What step are you stuck on when posting your Sleepyhead results? Or what steps have you completed successfully before you you get stuck?lwieland11 wrote: ↑Thu Sep 20, 2018 10:22 amI have tried several times unsuccessfully to post Sleepyhead results because I have a Mac.
_________________
| Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
| Additional Comments: Pressure 8.2cm-14.6cm, EPR 1, Humidity Auto, Contec CMS50F |
Re: Success Finally But Why Don't I Feel Better?
That's not a low AHI. that's a bad AHI. AHI under 2 is a good AHI
Unfortunately, meaningless.
I'd feel like roadkill with an AHI of 5.5.
See above.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Success Finally But Why Don't I Feel Better?
Your definition of "success" falls short of a positive outcome.
Some numbers are useful--but not always the ones we are told.
PaleRider knows.
Some numbers are useful--but not always the ones we are told.
PaleRider knows.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
- zoocrewphoto
- Posts: 3732
- Joined: Mon Apr 30, 2012 10:34 pm
- Location: Seatac, WA
Re: Success Finally But Why Don't I Feel Better?
Doctors and insurance companies use ahi of 5 as a criterion for whether you have sleep apnea or not. But itdoesn't really represent what is great and when you will feel better. '
Consider this. An ahi of 5 means that your sleep is disturbed an average of 5 times an hour. If you sleep 8 hours, that is 40 times a night. That's really not that great. Better than untreated, but most people can get much better results if they use the data to dial in on their ideal settings.
Many people have found that they notice more of a difference when the ahi is below 2. Many is usually under 1, and extremely rare to reach 2.
Other reasons to have good numbers and not feel improved:
waking up frequently - which could be caused by leaks, pressure changes, something uncomfortable, something unrelated to cpap
not enough time asleep
pain
medications
mattress, pillow, noises, temperature, etc
Consider this. An ahi of 5 means that your sleep is disturbed an average of 5 times an hour. If you sleep 8 hours, that is 40 times a night. That's really not that great. Better than untreated, but most people can get much better results if they use the data to dial in on their ideal settings.
Many people have found that they notice more of a difference when the ahi is below 2. Many is usually under 1, and extremely rare to reach 2.
Other reasons to have good numbers and not feel improved:
waking up frequently - which could be caused by leaks, pressure changes, something uncomfortable, something unrelated to cpap
not enough time asleep
pain
medications
mattress, pillow, noises, temperature, etc
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Resmed S9 autoset pressure range 11-17 |
Who would have thought it would be this challenging to sleep and breathe at the same time?
- Midwest_non_sleeper
- Posts: 436
- Joined: Wed Jun 20, 2018 1:03 pm
Re: Success Finally But Why Don't I Feel Better?
How long have you been on therapy?
I can tell you right now that if you asked me how I felt the day after my first night, I would tell you that I felt like I had been catapulted into a solid stone wall by a giant medieval trébuchet and then run over by a bus. A modern double-decker bus.
But, if you asked me right now, after being on therapy for four months, I would tell you that I feel friggin' fantastic. The difference between day one me and four month me is astounding. If you haven't been on it for very long, give your body some time to adjust to being able to breath properly, although an AHI of 5.5 is...high, for me at least. I rarely hit over 2, but when I do I feel like garbage, or like I described previously. Some lucky people feel fantastic after the first night, but the majority of people tend to see very small daily upward swings.
Don't give up!
I can tell you right now that if you asked me how I felt the day after my first night, I would tell you that I felt like I had been catapulted into a solid stone wall by a giant medieval trébuchet and then run over by a bus. A modern double-decker bus.
But, if you asked me right now, after being on therapy for four months, I would tell you that I feel friggin' fantastic. The difference between day one me and four month me is astounding. If you haven't been on it for very long, give your body some time to adjust to being able to breath properly, although an AHI of 5.5 is...high, for me at least. I rarely hit over 2, but when I do I feel like garbage, or like I described previously. Some lucky people feel fantastic after the first night, but the majority of people tend to see very small daily upward swings.
Don't give up!
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: DreamWear Gel Nasal Pillow CPAP Mask with Headgear - Fit Pack (All Nasal Pillows with Medium Frame) |
| Additional Comments: 9.2cmH20 / 20cmH20 pressure |
-
lwieland11
- Posts: 249
- Joined: Sat Dec 10, 2016 5:57 am
Re: Success Finally But Why Don't I Feel Better?
Thanks, Midwest. I was first diagnosed with severe sleep apnea in 2003! I've tried many different types of machines, masks, etc. SEVEN sleep studies to date and they want me to do another. Ridiculous. I'm currently on a BPAP now w/backup rate. Pressure of 17. For a girl whose AHI was 29+, AHI of 5 and sometimes under seems like a miracle that I thought I'd never see. I have found a way to eliminate the aerophagia I was dealing with for so long. I sleep only on my side with mouth shut and this seems to be working. Still tweaking and hoping for more because I don't look or feel rested. There may be other reasons for that. Just wondering if I can improve my BPAP therapy any more or is this the best it's going to be.
Re: Success Finally But Why Don't I Feel Better?
Are you suggesting that 29 was some kind of extreme (on testing)? It's not, compared to so many who test much higher, but in any case the 'severity' of your tested AHI has no bearing on how high of a pressure(s) your Cpap is set at... that's done after gauging how narrow your airway is when titrated - e.g. I may have had a really low AHI (14) on testing, but a very narrow airway so would need high pressures, whereas others might have tested high (e.g. 75) but only need a relatively low min. pressure of 6 to keep their airway more open. Apples and oranges!
- Jas_williams
- Posts: 1120
- Joined: Tue Dec 19, 2017 2:12 pm
- Location: Somerset UK
Re: Success Finally But Why Don't I Feel Better?
post some daily graphs to this thread and we can see if we can help you further, an AHI ave of 5.5 still has room for improvement.
_________________
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask |
-
freetimecreations
- Posts: 145
- Joined: Thu May 24, 2012 9:02 pm
- Location: Utah, USA
Re: Success Finally But Why Don't I Feel Better?
Jas_williams wrote: ↑Sat Sep 22, 2018 5:26 ampost some daily graphs to this thread and we can see if we can help you further, an AHI ave of 5.5 still has room for improvement.
What step are you stuck on when posting your Sleepyhead results? Or what steps have you completed successfully before you you get stuck?lwieland11 wrote: ↑Thu Sep 20, 2018 10:22 amI have tried several times unsuccessfully to post Sleepyhead results because I have a Mac.
_________________
| Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
| Additional Comments: Pressure 8.2cm-14.6cm, EPR 1, Humidity Auto, Contec CMS50F |
Re: Success Finally But Why Don't I Feel Better?
Considering the fact that SH is written on a Mac, ...freetimecreations wrote: ↑Sat Sep 22, 2018 2:32 pmJas_williams wrote: ↑Sat Sep 22, 2018 5:26 ampost some daily graphs to this thread and we can see if we can help you further, an AHI ave of 5.5 still has room for improvement.What step are you stuck on when posting your Sleepyhead results? Or what steps have you completed successfully before you you get stuck?lwieland11 wrote: ↑Thu Sep 20, 2018 10:22 amI have tried several times unsuccessfully to post Sleepyhead results because I have a Mac.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- ChicagoGranny
- Posts: 15533
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Success Finally But Why Don't I Feel Better?
That's a high pressure. Wearing a firm foam cervical collar has helped some lower their pressure and improve their AHI. The collar holds the neck in the best position for airway patency. You can get some idea about the different collars here - viewtopic/t172782/Success-with-Dr-Dakot ... evice.html