AHI numbers getting higher.

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Muddpa
Posts: 48
Joined: Mon Jul 31, 2006 5:41 pm

AHI numbers getting higher.

Post by Muddpa » Wed Sep 12, 2018 10:55 am

Can't figure it out, I been on CPAP for over twenty years. Started out with a S7, then to the Phillips line, now on the Dream Station auto Bi-pap. my start numbers are 8 and 5. With use they are 13 and 10. My ahi for most nights are any where from a high of 9.0 to a low of 1.0. I use to never get over a 2.4. Yes an older now still getting 7+ hours sleep, but late afternoon I start to get sleepy. Anybody got an idea about this. My doctor is not worry about this, I am. :?

_________________
MachineMask
Additional Comments: Looking for the right mask still
Some were there is a mask for me :D

User avatar
Pugsy
Posts: 65195
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AHI numbers getting higher.

Post by Pugsy » Wed Sep 12, 2018 7:25 pm

Are you using software to see what that AHI is made up of?

Can you post an example of a night with high AHI and nice low AHI?
So we can see if anything stands out that might explain the difference in AHI and how you feel.

Don't know how to post an image...see this thread.
viewtopic/t158560/How-to-post-images-for-review.html

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: AHI numbers getting higher.

Post by Julie » Wed Sep 12, 2018 8:22 pm

Also, don't take this wrong, but are you on any (relatively) new meds? Have you (maybe slowly over time) gained any weight? Even a little can make a difference.

User avatar
Muddpa
Posts: 48
Joined: Mon Jul 31, 2006 5:41 pm

Re: AHI numbers getting higher.

Post by Muddpa » Mon Sep 24, 2018 6:27 am

To all thanks for the help. Sorry for the slow reply. No new meds, my weigh 235-250, I'm on a yo-yo diet. Still trying to upload images.

_________________
MachineMask
Additional Comments: Looking for the right mask still
Some were there is a mask for me :D

User avatar
Muddpa
Posts: 48
Joined: Mon Jul 31, 2006 5:41 pm

Re: AHI numbers getting higher.

Post by Muddpa » Mon Sep 24, 2018 6:42 am

2018-09-24.png
2018-09-24.png (174.61 KiB) Viewed 1088 times
I hope this works.

_________________
MachineMask
Additional Comments: Looking for the right mask still
Some were there is a mask for me :D

User avatar
Muddpa
Posts: 48
Joined: Mon Jul 31, 2006 5:41 pm

Re: AHI numbers getting higher.

Post by Muddpa » Mon Sep 24, 2018 6:46 am

2018-09-24 (1).png
2018-09-24 (1).png (193.06 KiB) Viewed 1088 times
Bad night

_________________
MachineMask
Additional Comments: Looking for the right mask still
Some were there is a mask for me :D

User avatar
Pugsy
Posts: 65195
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AHI numbers getting higher.

Post by Pugsy » Mon Sep 24, 2018 7:36 am

For some reason or other your pressure needs are higher during some parts of the night and not others.
When this happens we normally look at 2 of the usual suspects....sleeping on your back or REM stage sleep both of which it is common to see our OSA worsen and/or pressure needs increase.

Your minimum EPAP isn't doing a good job holding the airway open and the machine can't respond quickly enough to get to where it needs to be when those whatever (supine sleeping or REM sleep) changes are happening.

You need more minimum EPAP. 7 or 8 would be what I would look at.
Even your "good" night isn't a great night IMHO.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Muddpa
Posts: 48
Joined: Mon Jul 31, 2006 5:41 pm

Re: AHI numbers getting higher.

Post by Muddpa » Wed Sep 26, 2018 7:13 am

Thanks for the help, will change the min should I also change the other? Right now on auto is 5 & 8. I would like to stay on auto if possible.

_________________
MachineMask
Additional Comments: Looking for the right mask still
Some were there is a mask for me :D

User avatar
Pugsy
Posts: 65195
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AHI numbers getting higher.

Post by Pugsy » Wed Sep 26, 2018 7:32 am

I would only change that minimum EPAP.
When you change the minimum EPAP it automatically will change the IPAP minimum and I see no reason to change the max IPAP at this time.
Don't blame you for wanting to stay with auto mode. That's my preference also.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Muddpa
Posts: 48
Joined: Mon Jul 31, 2006 5:41 pm

Re: AHI numbers getting higher.

Post by Muddpa » Fri Sep 28, 2018 7:46 am

Hey Pugsy, thanks for the help and all you do. I checked with my better half and found out I was sleeping more on my back. Now back on my side and my numbers are dropping. I will still raise my number a little bit. :D :) :)

_________________
MachineMask
Additional Comments: Looking for the right mask still
Some were there is a mask for me :D