S/H Pressure vs. Mask Pressure?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
MurrayNevada
Posts: 235
Joined: Sun May 20, 2018 8:25 am

S/H Pressure vs. Mask Pressure?

Post by MurrayNevada » Tue Aug 21, 2018 7:59 am

What is the difference between SleepyHead "Pressure" and "Mask Pressure?" They are always different. Which one most accurately reflects my pressures throughout the night?

_________________
Mask: ResMed AirTouch™ N30i Frame System - SW/STD
Additional Comments: Machine: Lowenstein Prisma 20a - Software: SleepHQ & OSCAR

D.H.
Posts: 3534
Joined: Sun Sep 04, 2016 7:07 pm

Re: S/H Pressure vs. Mask Pressure?

Post by D.H. » Tue Aug 21, 2018 8:01 am

I think that you need to give us more detail. For instance, actual numbers for both pressures.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

User avatar
Pugsy
Posts: 65234
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: S/H Pressure vs. Mask Pressure?

Post by Pugsy » Tue Aug 21, 2018 8:08 am

Examples please.
I think I know what you are seeing but need to see examples to be sure. Are you using EPR?

Mask pressure shows the pressure at the mask end and will reflect inhale and exhale in one graph

Regular pressure line shows whatever pressure you have selected to show and for me it was EPAP until I changed it to show IPAP.
Hover your mouse over the word pressure on the graph line...right click mouse and then click on plots. Mine wants to continually default to EPAP so I see the one line of EPAP pressure which reflects my drop during exhale because of EPR. Worried me until I realized what it was showing. Now I just mentally adjust in my head. Or go in and change the plots to show both IPAP and EPAP.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
MurrayNevada
Posts: 235
Joined: Sun May 20, 2018 8:25 am

Re: S/H Pressure vs. Mask Pressure?

Post by MurrayNevada » Tue Aug 21, 2018 8:45 am

These graphs hopefully show Pressure vs Mask Pressure:

_________________
Mask: ResMed AirTouch™ N30i Frame System - SW/STD
Additional Comments: Machine: Lowenstein Prisma 20a - Software: SleepHQ & OSCAR
Attachments
Pressure Plot.png
Pressure Plot.png (204.54 KiB) Viewed 515 times

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: S/H Pressure vs. Mask Pressure?

Post by palerider » Tue Aug 21, 2018 2:01 pm

MurrayNevada wrote:
Tue Aug 21, 2018 7:59 am
What is the difference between SleepyHead "Pressure" and "Mask Pressure?" They are always different. Which one most accurately reflects my pressures throughout the night?
"Pressure" is the programmed 'I want the pressure to be this' in the machine, either what's set, or what the auto has decided the pressure should be.

"Mask Pressure" is the *measured* pressure at the machine outlet. You can see it vary up and down if you zoom in, you can see the vibration of the Forced Oscillation Technique" when an auto tries to determine if an apnea is central or obstructive.

Also, if you look at the examples you've shown, they're not different. The y axis scales on the graphs are different, but the traces are the same, except for brief glitches.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.