Reduce noise on Airmini and use any mask

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
digitaldog888
Posts: 22
Joined: Sun Jul 22, 2018 3:56 pm

Reduce noise on Airmini and use any mask

Post by digitaldog888 » Wed Aug 15, 2018 1:46 pm

I'm using an AirSense 10 at home and wanted a travel CPAP machine. Without reading the forums, I decided to purchase the Airmini based on its size and portability. I was disappointed when I discovered it to be noisier than the AirSense 10. I found this forum and read that the noise was coming from the humidX housing. I designed a connector that by-passed the humidX housing but unfortunately, it didn't solve the issue.

After some research, I found the following:

https://goo.gl/5sA9tE

As the Airmini uses a proprietary connector, I designed a connector that could be used with any standard 22mm tubing:
AM-connector.jpg
AM-connector.jpg (242.45 KiB) Viewed 48530 times
AM-connector-with-hose.jpg
AM-connector-with-hose.jpg (295.21 KiB) Viewed 48530 times
Here it is with the Qtube muffler attached:
Qtube-muffler.jpg
Qtube-muffler.jpg (348.98 KiB) Viewed 48530 times
From my personal observations, the Qtube muffler greatly reduced the noise with the N20 mask I'm using. It's still not as quiet as my AirSense 10 but it's acceptable, at least for me.

For anyone interested in my connector, it could be found here:

https://www.thingiverse.com/thing:3049870

It's connected to the Airmini using an interference fit. I may add the "tabs" at a later date and will update the files accordingly.

jagstyles
Posts: 26
Joined: Tue Jun 05, 2018 8:16 pm

Re: Reduce noise on Airmini and use any mask

Post by jagstyles » Thu Sep 06, 2018 7:00 pm

How do I go about ordering this connector? Also, is the QTube you attached the same one that is used for the Z1 machine? I believe that's the only one available? I just need to know which parts I need to get and where to purchase them from as the noise coming from the humidex fitting is driving me nuts and I was really hoping to use the Airmini for travel.

Thanks

_________________
MachineMask

jagstyles
Posts: 26
Joined: Tue Jun 05, 2018 8:16 pm

Re: Reduce noise on Airmini and use any mask

Post by jagstyles » Mon Sep 10, 2018 10:15 am

So I've gone ahead and ordered it using one of the 3D print services available through Thingiverse. I've also ordered QTube and hope this all works out as I really like the travel size of this machine.

Thanks

_________________
MachineMask

digitaldog888
Posts: 22
Joined: Sun Jul 22, 2018 3:56 pm

Re: Reduce noise on Airmini and use any mask

Post by digitaldog888 » Mon Sep 10, 2018 9:05 pm

jagstyles wrote: ↑
Thu Sep 06, 2018 7:00 pm
How do I go about ordering this connector? Also, is the QTube you attached the same one that is used for the Z1 machine? I believe that's the only one available? I just need to know which parts I need to get and where to purchase them from as the noise coming from the humidex fitting is driving me nuts and I was really hoping to use the Airmini for travel.

Thanks
The QTube attached is the same one used for the Z1. It comes with the 8" extension tube that connects to one end of the QTube and the other to the custom connector for the Airmini on Thingiverse. If you don't have access to a 3D printer, you could order it through one of the services on Thingiverse.

digitaldog888
Posts: 22
Joined: Sun Jul 22, 2018 3:56 pm

Re: Reduce noise on Airmini and use any mask

Post by digitaldog888 » Mon Sep 10, 2018 9:09 pm

jagstyles wrote: ↑
Mon Sep 10, 2018 10:15 am
So I've gone ahead and ordered it using one of the 3D print services available through Thingiverse. I've also ordered QTube and hope this all works out as I really like the travel size of this machine.

Thanks
It does reduce the noise to an acceptable level that I could fall asleep. It's still not as quiet as my AirSense 10. Without the QTube, the noise was so bad that I could not fall asleep and had to stop using it.

jagstyles
Posts: 26
Joined: Tue Jun 05, 2018 8:16 pm

Re: Reduce noise on Airmini and use any mask

Post by jagstyles » Tue Sep 11, 2018 9:11 am

digitaldog888 wrote: ↑
Mon Sep 10, 2018 9:09 pm
jagstyles wrote: ↑
Mon Sep 10, 2018 10:15 am
So I've gone ahead and ordered it using one of the 3D print services available through Thingiverse. I've also ordered QTube and hope this all works out as I really like the travel size of this machine.

Thanks
It does reduce the noise to an acceptable level that I could fall asleep. It's still not as quiet as my AirSense 10. Without the QTube, the noise was so bad that I could not fall asleep and had to stop using it.
As long as it reduces the "darth vadar" sound enough that I can fall asleep, I'm good with that. Anything is better than the current sound and I'm quite shocked that resmed thought this level of noise from the humidex chamber was acceptable.

_________________
MachineMask

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Reduce noise on Airmini and use any mask

Post by chunkyfrog » Tue Sep 11, 2018 9:22 am

Maybe Resmed's intended market for the Mini is the HOH/deaf. :shock:

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

jagstyles
Posts: 26
Joined: Tue Jun 05, 2018 8:16 pm

Re: Reduce noise on Airmini and use any mask

Post by jagstyles » Mon Sep 17, 2018 8:16 pm

digitaldog888 wrote: ↑
Mon Sep 10, 2018 9:09 pm
jagstyles wrote: ↑
Mon Sep 10, 2018 10:15 am
So I've gone ahead and ordered it using one of the 3D print services available through Thingiverse. I've also ordered QTube and hope this all works out as I really like the travel size of this machine.

Thanks
It does reduce the noise to an acceptable level that I could fall asleep. It's still not as quiet as my AirSense 10. Without the QTube, the noise was so bad that I could not fall asleep and had to stop using it.
Got the 3D printed item today and definitely improves the sound. Still a bit louder than my airsense 10 machine but works well. Question, I received two pieces. One has the latches where you insert it into the airmini, not sure what the other piece is for though?

_________________
MachineMask

digitaldog888
Posts: 22
Joined: Sun Jul 22, 2018 3:56 pm

Re: Reduce noise on Airmini and use any mask

Post by digitaldog888 » Mon Sep 17, 2018 9:23 pm

jagstyles wrote: ↑
Mon Sep 17, 2018 8:16 pm
digitaldog888 wrote: ↑
Mon Sep 10, 2018 9:09 pm
jagstyles wrote: ↑
Mon Sep 10, 2018 10:15 am
So I've gone ahead and ordered it using one of the 3D print services available through Thingiverse. I've also ordered QTube and hope this all works out as I really like the travel size of this machine.

Thanks
It does reduce the noise to an acceptable level that I could fall asleep. It's still not as quiet as my AirSense 10. Without the QTube, the noise was so bad that I could not fall asleep and had to stop using it.
Got the 3D printed item today and definitely improves the sound. Still a bit louder than my airsense 10 machine but works well. Question, I received two pieces. One has the latches where you insert it into the airmini, not sure what the other piece is for though?
I designed two versions of the connector. One with the "tabs" to lock it in place with the Airmini. The other version without the "tabs" uses an interference fit and may slip out. Depending on the print settings, the one with the "tabs" might break if too much pressure is applied.

digitaldog888
Posts: 22
Joined: Sun Jul 22, 2018 3:56 pm

Re: Reduce noise on Airmini and use any mask

Post by digitaldog888 » Mon Sep 17, 2018 9:30 pm

jagstyles wrote: ↑
Mon Sep 17, 2018 8:16 pm
digitaldog888 wrote: ↑
Mon Sep 10, 2018 9:09 pm
jagstyles wrote: ↑
Mon Sep 10, 2018 10:15 am
So I've gone ahead and ordered it using one of the 3D print services available through Thingiverse. I've also ordered QTube and hope this all works out as I really like the travel size of this machine.

Thanks
It does reduce the noise to an acceptable level that I could fall asleep. It's still not as quiet as my AirSense 10. Without the QTube, the noise was so bad that I could not fall asleep and had to stop using it.
Got the 3D printed item today and definitely improves the sound. Still a bit louder than my airsense 10 machine but works well. Question, I received two pieces. One has the latches where you insert it into the airmini, not sure what the other piece is for though?
Another item you might want to consider is the Z1 HME for humidification:

https://www.amazon.com/Bundle-Moisture- ... humidifier

Essentially, with the custom connector, Qtube and the Z1 HME, you have a Z1 travel CPAP solution but only with an Airmini :)

lnichols
Posts: 2
Joined: Sat Feb 02, 2019 5:45 pm

Re: Reduce noise on Airmini and use any mask

Post by lnichols » Sat Feb 02, 2019 5:54 pm

I see how to connect the Thingiverse connector to the AirMini and to the Qtube, and how to connect a Z1 HME. But how do I connect the AirMini air tube + mask to the Z1 HME (or directly to the Qtube)? I have the two Thingiverse connectors shown above, but neither connects to the AirMini air tube. I think I must be missing something. Help?

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: Reduce noise on Airmini and use any mask

Post by raisedfist » Sat Feb 02, 2019 7:13 pm

ResMed really screwed up with the air mini...gtfoh with all this proprietary stuff

_________________
Mask
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Reduce noise on Airmini and use any mask

Post by chunkyfrog » Sat Feb 02, 2019 9:40 pm

No reason for all this proprietary nonsense.
Just greed and a very TWISTED sense of humor.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: Reduce noise on Airmini and use any mask

Post by Jas_williams » Sun Feb 03, 2019 1:33 am

lnichols wrote: ↑
Sat Feb 02, 2019 5:54 pm
I see how to connect the Thingiverse connector to the AirMini and to the Qtube, and how to connect a Z1 HME. But how do I connect the AirMini air tube + mask to the Z1 HME (or directly to the Qtube)? I have the two Thingiverse connectors shown above, but neither connects to the AirMini air tube. I think I must be missing something. Help?
You are this was designed to connect a normal cpap tube not the airmini tube

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask

lnichols
Posts: 2
Joined: Sat Feb 02, 2019 5:45 pm

Re: Reduce noise on Airmini and use any mask

Post by lnichols » Mon Feb 04, 2019 12:10 am

Sorry, I am rather new to CPAP. The heading mentions AirMini. What is a “normal CPAP tube?” And how do I connect the mask that came with my AirMini to this Qtube? Thanks!