AHI calculcation in SH vs CPAP and MyAir

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
BlueDragon
Posts: 547
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

AHI calculcation in SH vs CPAP and MyAir

Post by BlueDragon » Tue Aug 14, 2018 10:58 am

I've noticed occasional discrepancies between the AHI displayed on the CPAP machine (and also reported in MyAir) vs Sleepyhead. Last night, SH reported AHI of 0.32 and the machine and MyAir reported 2.8. Looking at the numbers, it would appear that the machine includes the number of large leak events in its AHI calculation, while SH does not. (I had 25 large leak events last night because of a mask issue.)

Is this correct?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
See OSCAR for the latest release.
OSCAR Team
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AHI calculcation in SH vs CPAP and MyAir

Post by palerider » Tue Aug 14, 2018 12:57 pm

BlueDragon wrote:
Tue Aug 14, 2018 10:58 am
I've noticed occasional discrepancies between the AHI displayed on the CPAP machine (and also reported in MyAir) vs Sleepyhead. Last night, SH reported AHI of 0.32 and the machine and MyAir reported 2.8. Looking at the numbers, it would appear that the machine includes the number of large leak events in its AHI calculation, while SH does not. (I had 25 large leak events last night because of a mask issue.)

Is this correct?
That is not correct.

AHI is Apnea Hypopnea Index. Apneas + Hypopneas / hours of use.

You'll have to dig deeper to see where the discrepancy is. It may be the reporting period.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Uff Da
Posts: 125
Joined: Mon May 07, 2018 5:11 pm

Re: AHI calculcation in SH vs CPAP and MyAir

Post by Uff Da » Tue Aug 14, 2018 1:41 pm

I had the same problem with discrepancies between Sleepyhead and ResMed on August 1. Some of the regulars on this forum gave me some ideas of what might be the cause. I still haven't figured it out and I had more problems a couple of times in the last week. I haven't added the details of those issues to my thread yet, but will try to find time later today or tomorrow to do so.

Anyway, you might want to check out the input people had on my similar issues in this thread:
viewtopic/t172484/Major-discrepancy-bet ... d-AHI.html

User avatar
BlueDragon
Posts: 547
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

Re: AHI calculcation in SH vs CPAP and MyAir

Post by BlueDragon » Tue Aug 14, 2018 2:28 pm

OK, so LL isn't the issue. Still, something very odd is going on with SH and ResScan/MyAir/CPAP not agreeing.

I compared ResScan and SH. Both have the same reporting period of 9:26 hours. There are two sessions. Both programs agree about 3 events in the first session. ResScan reports a large number of events in the second session, while SH reports none.

fwiw, I use FlashPap to transfer data from the SD card to my computer, and then import it into SH. I tried importing the data directly from the card to my existing SH user, with no change in results. However, when I created a new user in SH and imported all data again, SH then agreed with ResScan. That would seem to narrow the possible culprits to SH and FlashPap.

Is there a way in SH to delete a single day's data? If so, I'll delete 8/13 and try importing again from both FlashPap and the card to see if that shows anything different.

Suggestions welcome!
20180814-SH vs RS.png
2018-08-13 Resscan.png

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
See OSCAR for the latest release.
OSCAR Team
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.

User avatar
Pugsy
Posts: 65130
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AHI calculcation in SH vs CPAP and MyAir

Post by Pugsy » Tue Aug 14, 2018 2:41 pm

Try creating a new user profile in SleepyHead and importing again.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
BlueDragon
Posts: 547
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

Re: AHI calculcation in SH vs CPAP and MyAir

Post by BlueDragon » Tue Aug 14, 2018 4:30 pm

Pugsy wrote:
Tue Aug 14, 2018 2:41 pm
Try creating a new user profile in SleepyHead and importing again.
Yes, I did try that, importing both from the SD card and from FlashPap. With a new user, both imports worked (i.e., agreed with ResScan).

So, what's wrong with SH that it intermittently doesn't get all the data (or whatever it is that is going wrong)?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
See OSCAR for the latest release.
OSCAR Team
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.

User avatar
Pugsy
Posts: 65130
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AHI calculcation in SH vs CPAP and MyAir

Post by Pugsy » Tue Aug 14, 2018 5:23 pm

BlueDragon wrote:
Tue Aug 14, 2018 4:30 pm
So, what's wrong with SH that it intermittently doesn't get all the data (or whatever it is that is going wrong)?
It's Beta software so buggy...though this is a rather new bug.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
BlueDragon
Posts: 547
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

Re: AHI calculcation in SH vs CPAP and MyAir

Post by BlueDragon » Tue Aug 14, 2018 5:49 pm

Pugsy wrote:
Tue Aug 14, 2018 5:23 pm
It's Beta software so buggy...though this is a rather new bug.
Is there someone I should pass this on to? Where?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
See OSCAR for the latest release.
OSCAR Team
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.

User avatar
Pugsy
Posts: 65130
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AHI calculcation in SH vs CPAP and MyAir

Post by Pugsy » Tue Aug 14, 2018 6:01 pm

There is but Mark has supposedly been working on another new version of SleepyHead which has yet to be released so I don't know that he will care about bugs in the older version.

memberlist.php?mode=viewprofile&u=54879

I don't know how often he checks PMs here.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AHI calculcation in SH vs CPAP and MyAir

Post by palerider » Tue Aug 14, 2018 6:11 pm

Pugsy wrote:
Tue Aug 14, 2018 6:01 pm
There is but Mark has supposedly been working on another new version of SleepyHead which has yet to be released so I don't know that he will care about bugs in the older version.

memberlist.php?mode=viewprofile&u=54879

I don't know how often he checks PMs here.
It might be a bug that carries over.

You'd probably be better able to get in contact via his sleepyhead facebook page: https://www.facebook.com/groups/sleepyheadcpap/about/

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 65130
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AHI calculcation in SH vs CPAP and MyAir

Post by Pugsy » Tue Aug 14, 2018 6:51 pm

palerider wrote:
Tue Aug 14, 2018 6:11 pm
It might be a bug that carries over.
Yeah, but it's odd that it's just now happening and to just a couple of people maybe.

I haven't had it happen once and I have had my AirSense right at a year now.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
BlueDragon
Posts: 547
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

Re: AHI calculcation in SH vs CPAP and MyAir

Post by BlueDragon » Tue Aug 14, 2018 7:07 pm

According to notes on the Discord board, Mark is focused intently on a new version now, so bug reports may not be handled immediately. I'll wait for an update. SH is still the tool I like best for looking at what's happening.

All software has bugs.

Recently, I had two days with 11 hours sleep each (in multiple, short sessions as I was ill). One of the sleep sessions went from about 11 am to 2 pm. The CPAP nicely broke that into two sessions at noon sharp. However, SH managed to put 18 hours in one "day" and 4 hours in another, and didn't get the AHI computed reasonably. ResScan, on the other hand, put 11 hours in one day, and about 2 hours in the next and apparently discarded the other sessions. Only MyAir got it right, with two days with 11 hours of usage each.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
See OSCAR for the latest release.
OSCAR Team
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AHI calculcation in SH vs CPAP and MyAir

Post by palerider » Tue Aug 14, 2018 7:12 pm

Pugsy wrote:
Tue Aug 14, 2018 6:51 pm
palerider wrote:
Tue Aug 14, 2018 6:11 pm
It might be a bug that carries over.
Yeah, but it's odd that it's just now happening and to just a couple of people maybe.

I haven't had it happen once and I have had my AirSense right at a year now.
Yeah, it is weird.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: AHI calculcation in SH vs CPAP and MyAir

Post by Jas_williams » Wed Aug 15, 2018 2:24 am

Some times I have seen sleepyhead fail to import a day correctly. My answer find the day, open data => advanced => delete current day

Then reimport normally fixes the issue.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask

User avatar
BlueDragon
Posts: 547
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

Re: AHI calculcation in SH vs CPAP and MyAir

Post by BlueDragon » Wed Aug 15, 2018 8:11 am

Jas_williams wrote:
Wed Aug 15, 2018 2:24 am
Some times I have seen sleepyhead fail to import a day correctly. My answer find the day, open data => advanced => delete current day

Then reimport normally fixes the issue.
Thanks for pointing me to the data delete function. Tried that, and unfortunately it didn't make any difference. Still had the incorrect import.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
See OSCAR for the latest release.
OSCAR Team
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.