Will two chargers charge two Freedom CPAP batteries faster than one charger?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Will two chargers charge two Freedom CPAP batteries faster than one charger?

Post by palerider » Fri Aug 10, 2018 3:16 pm

mileena wrote:
Fri Aug 10, 2018 3:13 pm
palerider wrote:
Fri Aug 10, 2018 2:57 pm
mileena wrote:
Fri Aug 10, 2018 1:45 pm
Do I have a case for false advertising for this item I bought:
No. You bought the *respironics* 12 cord. There's nothing wrong with it.
Ok, thank you. I just know the DreamStation requires a proprietary voltage signal to use the full battery charge, which these shielded cords do not have. I think there should be a disclaimer about this product on the advertising.
Of course the respironics brand 12volt cord has everything that is needed. They make the machine, they make the cord, it would be ludicrous to think that their own cord, for their own machine, wouldn't work.

Why should there be a disclaimer on the product listing for the official respironics cord for the respironics dreamstation? What do you want? "Caution, this is the cord you need for a dreamstation"?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Last edited by palerider on Fri Aug 10, 2018 11:34 pm, edited 1 time in total.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

mileena
Posts: 194
Joined: Wed Mar 14, 2018 12:09 am

Re: Will two chargers charge two Freedom CPAP batteries faster than one charger?

Post by mileena » Fri Aug 10, 2018 3:24 pm

palerider wrote:
Fri Aug 10, 2018 3:16 pm
mileena wrote:
Fri Aug 10, 2018 3:13 pm
palerider wrote:
Fri Aug 10, 2018 2:57 pm
mileena wrote:
Fri Aug 10, 2018 1:45 pm
Do I have a case for false advertising for this item I bought:
No. You bought the *respironics* 12 cord. There's nothing wrong with it.
Ok, thank you. I just know the DreamStation requires a proprietary voltage signal to use the full battery charge, which these shielded cords do not have. I think there should be a disclaimer about this product on the advertising.
Why should there be a disclaimer on the product listing for the official respironics cord for the respironics dreamstation? What do you want? "Caution, this is the cord you need for a dreamstation"?
Because it is not an official Philips cord. It is false advertising. It is not fully compatible with DreamStation.
Philips DreamStation Auto CPAP HumDOM
Pressure: 13-15 cm H20
Cflex: 2
Fisher & Paykel Healthcare Simplus full-face mask
Last Sleep Study: 6/2018
Soon to get: ResMed Bilevel VAUTO PAP, ResMed Swift FX nasal pillows

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Will two chargers charge two Freedom CPAP batteries faster than one charger?

Post by palerider » Fri Aug 10, 2018 4:02 pm

mileena wrote:
Fri Aug 10, 2018 3:24 pm
palerider wrote:
Fri Aug 10, 2018 3:16 pm
mileena wrote:
Fri Aug 10, 2018 3:13 pm
palerider wrote:
Fri Aug 10, 2018 2:57 pm
mileena wrote:
Fri Aug 10, 2018 1:45 pm
Do I have a case for false advertising for this item I bought:
No. You bought the *respironics* 12 cord. There's nothing wrong with it.
Ok, thank you. I just know the DreamStation requires a proprietary voltage signal to use the full battery charge, which these shielded cords do not have. I think there should be a disclaimer about this product on the advertising.
Why should there be a disclaimer on the product listing for the official respironics cord for the respironics dreamstation? What do you want? "Caution, this is the cord you need for a dreamstation"?
Because it is not an official Philips cord. It is false advertising. It is not fully compatible with DreamStation.
Why do you say the one you bought isn't "fully compatible" with the dreamstation?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Stom
Posts: 383
Joined: Sat Jul 21, 2018 9:44 am

Re: Will two chargers charge two Freedom CPAP batteries faster than one charger?

Post by Stom » Fri Aug 10, 2018 4:22 pm

palerider wrote:
Fri Aug 10, 2018 4:02 pm
Why do you say the one you bought isn't "fully compatible" with the dreamstation?
Perhaps because of this:
jtravel wrote:
Wed Feb 15, 2017 3:55 pm
From another board:

I should probably point out that Philips Respironics has already joined the ResMed proprietary club with the introduction of the DreamStation line.

They use they same type of plug (same dimensions, but right angled) as the A10 but they use a simple 20k ohm resistor to +12V on the center pin.
some testing with a power supply and a decade box shows the DS needs to see between 18.9K to 26.2K to power up and run without an error screen. The blower will not start with the wrong resistance.
Outside barrel of plug is Ground
Inside Barrel of plug is Positive
Center pin is positive.
Official Foe™ on @Palerider's public list of foes. ;-)

Rules to live by: Be wary of anyone boasting their behavior is like that of a fictional sociopath.

mileena
Posts: 194
Joined: Wed Mar 14, 2018 12:09 am

Re: Will two chargers charge two Freedom CPAP batteries faster than one charger?

Post by mileena » Fri Aug 10, 2018 11:18 pm

palerider wrote:
Fri Aug 10, 2018 4:02 pm
mileena wrote:
Fri Aug 10, 2018 3:24 pm
palerider wrote:
Fri Aug 10, 2018 3:16 pm
mileena wrote:
Fri Aug 10, 2018 3:13 pm
palerider wrote:
Fri Aug 10, 2018 2:57 pm


No. You bought the *respironics* 12 cord. There's nothing wrong with it.
Ok, thank you. I just know the DreamStation requires a proprietary voltage signal to use the full battery charge, which these shielded cords do not have. I think there should be a disclaimer about this product on the advertising.
Why should there be a disclaimer on the product listing for the official respironics cord for the respironics dreamstation? What do you want? "Caution, this is the cord you need for a dreamstation"?
Because it is not an official Philips cord. It is false advertising. It is not fully compatible with DreamStation.
Why do you say the one you bought isn't "fully compatible" with the dreamstation?
Because the manufacturer of the Freedom battery said I need to proprietary DreamStation 12V adapter cord to get full usage of the battery charge. The regular shielded DC DreamStation cords do not work when the battery is on the lower end, so it will only be partially usable.
Philips DreamStation Auto CPAP HumDOM
Pressure: 13-15 cm H20
Cflex: 2
Fisher & Paykel Healthcare Simplus full-face mask
Last Sleep Study: 6/2018
Soon to get: ResMed Bilevel VAUTO PAP, ResMed Swift FX nasal pillows

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Will two chargers charge two Freedom CPAP batteries faster than one charger?

Post by palerider » Fri Aug 10, 2018 11:29 pm

mileena wrote:
Fri Aug 10, 2018 11:18 pm
palerider wrote:
Fri Aug 10, 2018 4:02 pm
mileena wrote:
Fri Aug 10, 2018 3:24 pm
palerider wrote:
Fri Aug 10, 2018 3:16 pm
mileena wrote:
Fri Aug 10, 2018 3:13 pm


Ok, thank you. I just know the DreamStation requires a proprietary voltage signal to use the full battery charge, which these shielded cords do not have. I think there should be a disclaimer about this product on the advertising.
Why should there be a disclaimer on the product listing for the official respironics cord for the respironics dreamstation? What do you want? "Caution, this is the cord you need for a dreamstation"?
Because it is not an official Philips cord. It is false advertising. It is not fully compatible with DreamStation.
Why do you say the one you bought isn't "fully compatible" with the dreamstation?
Because the manufacturer of the Freedom battery said I need to proprietary DreamStation 12V adapter cord to get full usage of the battery charge. The regular shielded DC DreamStation cords do not work when the battery is on the lower end, so it will only be partially usable.
Whether you believe the people trying to sell you the "extra special" cord *they* make or not, is up to you.

The Respironics cord, which you bought, was designed to work with a 12v battery, such as the ones that CapnLoki suggested you buy. It wasn't specially designed to work with your battery.

The real issue is "why is your supposed 'cpap battery' unable to provide power to the cpap without an extra 70$ cord?"

You seem to be terribly confused about what you've got, and what BPS is trying to sell you.

You bought a Philips Respironics Dreamstation. You then bought a Philips Respironcs 12v cord. It's made by PR, for your machine to be used on a 12volt source.

The 'special cord' from BPS is NOT made by PR, it's made by BPS.

You're bent out of shape because the *manufacturer* of the CPAP isn't custom designing their product to cater to some third party after market product.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Will two chargers charge two Freedom CPAP batteries faster than one charger?

Post by palerider » Fri Aug 10, 2018 11:45 pm

mileena wrote:
Fri Aug 10, 2018 1:37 pm
I am angry all these CPAP companies are allowed to sell these DC DreamStation shielded cords and claim that is the solution when they are not capable of using the full battery.
It's not THEIR fault you bought the battery you did.
mileena wrote:
Fri Aug 10, 2018 1:37 pm
The DreamStation actually requires a special cord that matches its proprietary voltage system. Other cords can't read when a battery is low voltage and the machine shuts off.
That's not how it works. The dreamstation, just like many *laptops* requires a signal from the power supply telling it what kind of power supply is being used. That's *all it is*, it has nothing to do with low battery voltage.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
CapnLoki
Posts: 1162
Joined: Tue Jun 25, 2013 6:26 pm
Location: North East

Re: Will two chargers charge two Freedom CPAP batteries faster than one charger?

Post by CapnLoki » Sat Aug 11, 2018 12:39 pm

palerider wrote:
Fri Aug 10, 2018 11:45 pm
mileena wrote:
Fri Aug 10, 2018 1:37 pm
I am angry all these CPAP companies are allowed to sell these DC DreamStation shielded cords and claim that is the solution when they are not capable of using the full battery.
It's not THEIR fault you bought the battery you did.
mileena wrote:
Fri Aug 10, 2018 1:37 pm
The DreamStation actually requires a special cord that matches its proprietary voltage system. Other cords can't read when a battery is low voltage and the machine shuts off.
That's not how it works. The dreamstation, just like many *laptops* requires a signal from the power supply telling it what kind of power supply is being used. That's *all it is*, it has nothing to do with low battery voltage.
If there's any truth to the "up-sell" claim of the BPS power cord, it may be that as the voltage drops on the 12v battery, the voltage may also drop on the signal. When that drops below the trigger level, the pump shuts off. Powering down when the voltage goes below some level is the proper thing on a lead acid battery where you don't want to go below about 20%, but with lithium you can take it lower. BPS may be holding the signal at a higher level to squeeze out an extra 10% or so from the battery. I have way of knowing the truth here - perhaps Mileena can do a test.

I'm also a bit skeptical on the claim that pressure 13 uses no more power than 10. Although PR doesn't publish numbers, the Resmed numbers (and others I've seen) indicate about 20% extra load for that pressure. Although its not a huge difference, its probably more than the up-sell power cord could give.

And the 2 extra hours to recharge from one charger don't quite make sense; if they really put out 6.75 amps there should be plenty of juice. Going from 0 to 50% should only take 2 hours, after that its tapering off so the total amperage available doesn't count. If they said an extra hour I might believe it, but this sounds like another up-sell. Again, I'm hoping mileena can report back.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Quattro™ Air Full Face Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Pressure 9-20, average ~9.5; often use battery power while off-grid
Hark, how hard he fetches breath . . .  Act II, Scene IV, King Henry IV Part I, William Shakespeare
Choosing a Battery thread: http://www.cpaptalk.com/viewtopic/t1140 ... ttery.html

mileena
Posts: 194
Joined: Wed Mar 14, 2018 12:09 am

Re: Will two chargers charge two Freedom CPAP batteries faster than one charger?

Post by mileena » Sat Aug 11, 2018 12:40 pm

palerider wrote:
Fri Aug 10, 2018 11:29 pm
mileena wrote:
Fri Aug 10, 2018 11:18 pm
palerider wrote:
Fri Aug 10, 2018 4:02 pm
mileena wrote:
Fri Aug 10, 2018 3:24 pm
palerider wrote:
Fri Aug 10, 2018 3:16 pm

Why should there be a disclaimer on the product listing for the official respironics cord for the respironics dreamstation? What do you want? "Caution, this is the cord you need for a dreamstation"?
Because it is not an official Philips cord. It is false advertising. It is not fully compatible with DreamStation.
Why do you say the one you bought isn't "fully compatible" with the dreamstation?
Because the manufacturer of the Freedom battery said I need to proprietary DreamStation 12V adapter cord to get full usage of the battery charge. The regular shielded DC DreamStation cords do not work when the battery is on the lower end, so it will only be partially usable.
Whether you believe the people trying to sell you the "extra special" cord *they* make or not, is up to you.

The Respironics cord, which you bought, was designed to work with a 12v battery, such as the ones that CapnLoki suggested you buy. It wasn't specially designed to work with your battery.

The real issue is "why is your supposed 'cpap battery' unable to provide power to the cpap without an extra 70$ cord?"

You seem to be terribly confused about what you've got, and what BPS is trying to sell you.

You bought a Philips Respironics Dreamstation. You then bought a Philips Respironcs 12v cord. It's made by PR, for your machine to be used on a 12volt source.

The 'special cord' from BPS is NOT made by PR, it's made by BPS.

You're bent out of shape because the *manufacturer* of the CPAP isn't custom designing their product to cater to some third party after market product.

With all due respect, you are so wrong.

The DC cord that works with a 12V battery you mentioned above is not fully compatible. This is what BPS says:

http://www.batterypowersolutions.net/pr ... r-adapter
While Respironics DreamStation series PAP devices operate from 12V direct current the DC input port on these devices require a proprietary signal in order to recognize a valid power source; therefore, a 12V power adapter is required to complete the connection between the device and the Freedom™ CPAP Battery.
So it is the DreamStation CPAP device itself which requires that a DC cord to be able to deliver the proprietary signal, regardless of the source, be it a Freedom travel battery, or a 12V AGM battery.

So why are companies advertising a shielded DC cord that cannot utilize the full battery charge for a DreamStation? It only works for a lead acid/AGM battery, but not fully with a lithium Freedom battery, like CapnLoki said in the post above.
Last edited by mileena on Sat Aug 11, 2018 1:23 pm, edited 3 times in total.
Philips DreamStation Auto CPAP HumDOM
Pressure: 13-15 cm H20
Cflex: 2
Fisher & Paykel Healthcare Simplus full-face mask
Last Sleep Study: 6/2018
Soon to get: ResMed Bilevel VAUTO PAP, ResMed Swift FX nasal pillows

mileena
Posts: 194
Joined: Wed Mar 14, 2018 12:09 am

Re: Will two chargers charge two Freedom CPAP batteries faster than one charger?

Post by mileena » Sat Aug 11, 2018 12:44 pm

palerider wrote:
Fri Aug 10, 2018 11:45 pm
mileena wrote:
Fri Aug 10, 2018 1:37 pm
I am angry all these CPAP companies are allowed to sell these DC DreamStation shielded cords and claim that is the solution when they are not capable of using the full battery.
It's not THEIR fault you bought the battery you did.
mileena wrote:
Fri Aug 10, 2018 1:37 pm
The DreamStation actually requires a special cord that matches its proprietary voltage system. Other cords can't read when a battery is low voltage and the machine shuts off.
That's not how it works. The dreamstation, just like many *laptops* requires a signal from the power supply telling it what kind of power supply is being used. That's *all it is*, it has nothing to do with low battery voltage.
It is their fault if they advertised falsely, which they did.

The DreamStation requires a proprietary signal for DC inputs. That is quoted from BPS in my post above. A regular shielded DC cord for DreamStation that all these companies advertise will work, but won't be able to tap the full power from the battery.
Last edited by mileena on Sun Aug 12, 2018 1:40 pm, edited 1 time in total.
Philips DreamStation Auto CPAP HumDOM
Pressure: 13-15 cm H20
Cflex: 2
Fisher & Paykel Healthcare Simplus full-face mask
Last Sleep Study: 6/2018
Soon to get: ResMed Bilevel VAUTO PAP, ResMed Swift FX nasal pillows

mileena
Posts: 194
Joined: Wed Mar 14, 2018 12:09 am

Re: Will two chargers charge two Freedom CPAP batteries faster than one charger?

Post by mileena » Sat Aug 11, 2018 1:00 pm

CapnLoki wrote:
Sat Aug 11, 2018 12:39 pm
palerider wrote:
Fri Aug 10, 2018 11:45 pm
mileena wrote:
Fri Aug 10, 2018 1:37 pm
I am angry all these CPAP companies are allowed to sell these DC DreamStation shielded cords and claim that is the solution when they are not capable of using the full battery.
It's not THEIR fault you bought the battery you did.
mileena wrote:
Fri Aug 10, 2018 1:37 pm
The DreamStation actually requires a special cord that matches its proprietary voltage system. Other cords can't read when a battery is low voltage and the machine shuts off.
That's not how it works. The dreamstation, just like many *laptops* requires a signal from the power supply telling it what kind of power supply is being used. That's *all it is*, it has nothing to do with low battery voltage.
If there's any truth to the "up-sell" claim of the BPS power cord, it may be that as the voltage drops on the 12v battery, the voltage may also drop on the signal. When that drops below the trigger level, the pump shuts off. Powering down when the voltage goes below some level is the proper thing on a lead acid battery where you don't want to go below about 20%, but with lithium you can take it lower. BPS may be holding the signal at a higher level to squeeze out an extra 10% or so from the battery. I have way of knowing the truth here - perhaps Mileena can do a test.

I'm also a bit skeptical on the claim that pressure 13 uses no more power than 10. Although PR doesn't publish numbers, the Resmed numbers (and others I've seen) indicate about 20% extra load for that pressure. Although its not a huge difference, its probably more than the up-sell power cord could give.

And the 2 extra hours to recharge from one charger don't quite make sense; if they really put out 6.75 amps there should be plenty of juice. Going from 0 to 50% should only take 2 hours, after that its tapering off so the total amperage available doesn't count. If they said an extra hour I might believe it, but this sounds like another up-sell. Again, I'm hoping mileena can report back.
Thanks CapnLoki!

You know what? I may just keep the less expensive DC shielded cord for DreamStation, and do a comparison with the more expensive proprietary signal cord from BPS. I was going to retun or sell the former, but it will be wroth it do to a test. A very good idea. I may not be able to do it for a few weeks, but I will do so!

Regarding the claim that a higher pressure doesn't drain the battery too much more, I should have made it more clear that the tech from BPS said they will be more drainage of the battery, but it should matter too much. He said around 18 cm H20, it will drain much more, but lower than that it should not matter too much. But maybe he was just trying to sell me or downplay it? 20% loss like you mentioned above does seem a lot!

And regarding the extra charging time, you probably are right! The tech was absolutely not trying to sell me more (they do not even sell the charger alone for the Freedom battery). He actually said it would not take much longer to charge two batteries with just one charger and a pigtail splitter, only "two hours". Maybe for him that was not a big time difference, but for me, when I am in a location where I don't have too much time to access a wall AC electrical outlet or am not driving, it does make a huge difference! So in the end, I spent another $312 on the special 12V DreamStation proprietary signal cord and another Freedom battery starter kit, which had the battery, another charger, pigtail splitter (which I also need if I am using two batteries as a backup at home), and various other adapter cords/pins.

I have spent a total of $585.65 on travel batteries and cords for my CPAP. Way more than I wanted to, but these batteries will last for 2,000 nights, or almost 5.5 years!!!! And are very portable, so I can go to a library or Starbucks or wherever when travelling to charge these quickly. Two batteries will last up to 4 nights. And the batteries have up to 500 charge cycles. so 4 days times 500 is 2,000 days!!
Philips DreamStation Auto CPAP HumDOM
Pressure: 13-15 cm H20
Cflex: 2
Fisher & Paykel Healthcare Simplus full-face mask
Last Sleep Study: 6/2018
Soon to get: ResMed Bilevel VAUTO PAP, ResMed Swift FX nasal pillows

mileena
Posts: 194
Joined: Wed Mar 14, 2018 12:09 am

Re: Will two chargers charge two Freedom CPAP batteries faster than one charger?

Post by mileena » Sat Aug 11, 2018 1:30 pm

Here is my mini-review of what I ordered and what I spent:

$249.95 - Freedom CPAP battery starter kit from SleepDirect.com: https://www.sleepdirect.com/parts-and-a ... attery-kit [you have to add the item to the cart to reveal this lower price]
$249.95 - Second Freedom CPAP battery starter kit from SleepDirect.com (they do sell just a second Freedom battery alone for $199.95: https://www.sleepdirect.com/parts-and-a ... le-battery [you have to add the item to the cart to reveal this lower price], but I needed a second battery charger and a second pigtail splitter, so I bought another kit, which comes to less money than buying individually)
$62.00 - DreamStation 12V Power Adapter from SleepDirect.com: https://www.sleepdirect.com/parts-and-a ... er-adapter
$23.75 - Shielded DC Power Cord for DreamStation from DirectHomeMedical.com: https://www.directhomemedical.com/11207 ... -cpap.html [ I used a coupon promotional code available online if you searched google so I could get the 5% discount ] [Note: this part was unnecessary, and I bought before I realized it wouldn't drain the entire battery, as the DreamStation requires a DC cord that can deliver a proprietary signal, which the cord above from SleepDirect.com does]

TOTAL: $585.65, though it would be only $561.90 if you didn't by the unnecessary cord like I did. All the above items had free shipping, although if you bought the proprietary cord from SleepDirect.com above alone, there would be a shipping charge. But if you combine it wiht the battery kit, which does offer free shipping, shipping then becomes free on the cord too.

I am not trying to advertise any of the above companies, but I found them to be the cheapest, with free shipping and promo codes that worked. They had excellent reviews, and shipped right away. Please be careful, as some online providers have horrible reviews, including with the BBB. Some also charge for shipping and may get you in with a low item price.

Also, you don't need a second battery, but I find it would be useful for me on the road, since without a power source, I cannot use my CPAP or sleep. I need to have the second battery for peace of mind, knowing things can go wrong. Also, I can use the two battery set up, as stated in the Battery Power Solutions manual, to run the CPAP longer without heated humidification or a heated tube (2-4 nights instead of only 1-2), and also use a heated humidifier and heated hose if I want for 1-2 days. Plus, since I have a higher pressure than the officially tested 10 cm H20, and my mask does leak a little and I use a Cflex as well, one battery may not even get me through a full night. I am going to have to do tests to find out for sure.

I also do realize AGM batteries and chargers, with all the adapter cords, as stated in a great writeup thread here by CapnLoki: viewtopic/t114012/Choosing-a-Battery.html, are much less expensive (less than $200), and might last longer over the years. However, for me, I need to have something ultra portable and concealable to charge. I can't go bringing in a AGM battery and charger to Starbucks, for example, along with my laptop and camera to charge. I also can't bring it into a library, along with the portable laser printer and scanner I have, as it would simply be too much to wheel in. I look like a bag lady enough as it is going into public places with a huge rolling suitcase and bags carrying all these electronics! :D
Last edited by mileena on Sat Aug 11, 2018 1:50 pm, edited 2 times in total.
Philips DreamStation Auto CPAP HumDOM
Pressure: 13-15 cm H20
Cflex: 2
Fisher & Paykel Healthcare Simplus full-face mask
Last Sleep Study: 6/2018
Soon to get: ResMed Bilevel VAUTO PAP, ResMed Swift FX nasal pillows

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Will two chargers charge two Freedom CPAP batteries faster than one charger?

Post by palerider » Sat Aug 11, 2018 1:40 pm

mileena wrote:
Sat Aug 11, 2018 12:44 pm
It is their fault if they advertised falsely, which they did.

The DreamStation requires a proprietary signal for DC inputs. That is quioted from BPS in my post above. A regular shielded DC cord for DreamStation that all these companies advertise will work, but won't be able to tap the full power from the battery.
Nobody advertised anything falsely. You need to get that notion out of your head.

It would only be "false advertising" if the AD claimed that the cable "works perfectly with the BPS lithium battery packs".

And it *doesn't*.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Last edited by palerider on Sat Aug 11, 2018 1:49 pm, edited 1 time in total.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Will two chargers charge two Freedom CPAP batteries faster than one charger?

Post by palerider » Sat Aug 11, 2018 1:44 pm

mileena wrote:
Sat Aug 11, 2018 1:00 pm
and do a comparison with the more expensive proprietary signal cord from BPS.
*ALL* cords that work on the dreamstation have the 'proprietary signal', otherwise the cord would not work. You're getting confused by the marketing people at BPS... they like that, because they can sell you more overpriced things.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

mileena
Posts: 194
Joined: Wed Mar 14, 2018 12:09 am

Re: Will two chargers charge two Freedom CPAP batteries faster than one charger?

Post by mileena » Sat Aug 11, 2018 1:46 pm

palerider wrote:
Sat Aug 11, 2018 1:40 pm
mileena wrote:
Sat Aug 11, 2018 12:44 pm
It is their fault if they advertised falsely, which they did.

The DreamStation requires a proprietary signal for DC inputs. That is quioted from BPS in my post above. A regular shielded DC cord for DreamStation that all these companies advertise will work, but won't be able to tap the full power from the battery.
Nobody advertised anything falsely. You need to get that notion out of your head.
I guess we will have to agree that we disagree then. I respect your opinion. I just wish the companies would have stated their product does not have the need proprietary signal needed to get full usage out of a lithium battery, like Freedom by BPS. It is good for lead acid batteries, but not other types. Now I have a useless adapter.