Newby, Review please

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Newby, Review please

Post by palerider » Sat Sep 22, 2018 2:08 pm

STRONGSLTN wrote:
Sat Sep 22, 2018 8:49 am
I first had two full face masks the comfort gel and then the Amara. I am now on the Amara view, That large leak was caused be having the chin portion slip into my mouth while asleep.
Maybe this graph of last night is better.
Drag these charts into place:
Flags
Flow Rate
Pressure
Leaks
Snore
Flow Limitations.

None other, resize them till those six fill the screen.

Don't cover part of the charts with your 'on screen keyboard'

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

STRONGSLTN
Posts: 29
Joined: Tue Aug 07, 2018 8:27 am
Location: Jekyll Island, Ga

Re: Newby, Review please

Post by STRONGSLTN » Sun Sep 23, 2018 8:55 am

Well, I was just told to pull snore, but I put it back in for you. I am hitting machine set pressure rates. I get a great seal, but lose it when chin cover starts slipping up to mouth. I don't yet know about the different apnea types. Sleep is much better with this mask.
Graphic_9_23_2018 10_44_41 AM.jpg
Graphic_9_23_2018 10_44_41 AM.jpg (426.99 KiB) Viewed 5290 times
I wou;d like to thank everyone for their concern and help. I'm trying.

_________________
MachineMask
Life is good

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Newby, Review please

Post by palerider » Sun Sep 23, 2018 1:46 pm

STRONGSLTN wrote:
Sun Sep 23, 2018 8:55 am
Well, I was just told to pull snore, but I put it back in for you.
Check my message again. raise your minimum to 12, max to 20.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

STRONGSLTN
Posts: 29
Joined: Tue Aug 07, 2018 8:27 am
Location: Jekyll Island, Ga

Re: Newby, Review please

Post by STRONGSLTN » Tue Sep 25, 2018 9:05 am

Only 5 on machine, but resmed gave me 20/20 for mask seal

_________________
MachineMask
Attachments
Graphic_9_25_2018 9_55_44 AM.jpg
Graphic_9_25_2018 9_55_44 AM.jpg (419.4 KiB) Viewed 5257 times
Life is good

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Newby, Review please

Post by palerider » Tue Sep 25, 2018 11:26 am

STRONGSLTN wrote:
Tue Sep 25, 2018 9:05 am
Only 5 on machine, but resmed gave me 20/20 for mask seal
Who knows, you don't have the leaks strip, as requested, multiple times.

Flags
Flow
Pressure
Leaks
flow limitations
snore.

That's six.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

STRONGSLTN
Posts: 29
Joined: Tue Aug 07, 2018 8:27 am
Location: Jekyll Island, Ga

Re: Newby, Review please

Post by STRONGSLTN » Tue Sep 25, 2018 11:43 am

It switched back, I missed it. Sorry

_________________
MachineMask
Attachments
Graphic_9_25_2018 1_36_45 PM.jpg
Graphic_9_25_2018 1_36_45 PM.jpg (428.28 KiB) Viewed 5250 times
Life is good

STRONGSLTN
Posts: 29
Joined: Tue Aug 07, 2018 8:27 am
Location: Jekyll Island, Ga

Re: Newby, Review please

Post by STRONGSLTN » Wed Sep 26, 2018 6:47 am

Next day

_________________
MachineMask
Attachments
Graphic_9_26_2018 8_42_21 AM.jpg
Graphic_9_26_2018 8_42_21 AM.jpg (453.2 KiB) Viewed 5233 times
Life is good

User avatar
Pugsy
Posts: 65121
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Newby, Review please

Post by Pugsy » Wed Sep 26, 2018 7:05 am

Leaks are much better managed. I could live with that small segment of large leak.

Your pressure settings are sub optimal. The minimum is too low and the maximum is too low. The machine wants to go higher but it can't and that is allowing the airway to collapse and you get those OAs and a higher AHI than we would like to see.
Probably going to need at least 12 or 13 minimum because once you go to sleep the machine seems to go up to around 12 ish and stay there anyway. For sure the maximum needs to be probably 20. Are you comfortable with changing your own pressure settings?
What position are you usually sleeping in?
You may need pressures in the upper teens and if that is the case you may want to consider adding in some EPR if there is a problem exhaling with the higher pressures.

How come the fragmented sleep? I see at least 3 times where you turned the machine off and back on again. Any idea why the wake ups?
Were there more than the 3 but you just didn't turn the machine off.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

STRONGSLTN
Posts: 29
Joined: Tue Aug 07, 2018 8:27 am
Location: Jekyll Island, Ga

Re: Newby, Review please

Post by STRONGSLTN » Thu Sep 27, 2018 8:40 am

I have adjusted machine before and will follow up. After review method. The breaks are bathroom most likely., length w/o need has improved.I mostly like to sleep on my side, but have found on bcak to help with easy breathing. I was told not to sleep on back. Reason? Wakeups may well be dry mouth as well, don't know how to solve this.Higher humidity did not seem to work,and I like cooler air.
Graphic_9_27_2018 10_35_26 AM.jpg
Graphic_9_27_2018 10_35_26 AM.jpg (443.74 KiB) Viewed 5215 times
That was last night. Ill adjust pressures today.

_________________
MachineMask
Life is good

PaulKTF
Posts: 190
Joined: Fri Jun 09, 2017 8:57 pm
Location: Wisconsin

Re: Newby, Review please

Post by PaulKTF » Thu Sep 27, 2018 12:41 pm

Try a F&P Simplus FFM- I find it's easy to make a good seal and it's lightweight and comfortable.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: Pressure is set to 10 cmH2O.
-Paul.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Newby, Review please

Post by palerider » Thu Sep 27, 2018 1:36 pm

I'd suggest trying min 14 max 20 for your pressures... that'll help.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

STRONGSLTN
Posts: 29
Joined: Tue Aug 07, 2018 8:27 am
Location: Jekyll Island, Ga

Re: Newby, Review please

Post by STRONGSLTN » Fri Sep 28, 2018 6:27 am

Went to the 12/20 for last night.
Graphic_9_28_2018 8_22_22 AM.jpg
Graphic_9_28_2018 8_22_22 AM.jpg (429.99 KiB) Viewed 5188 times

_________________
MachineMask
Life is good

STRONGSLTN
Posts: 29
Joined: Tue Aug 07, 2018 8:27 am
Location: Jekyll Island, Ga

Re: Newby, Review please

Post by STRONGSLTN » Fri Sep 28, 2018 6:29 am

I was told not to sleep on back. Reason? Wakeups may well be dry mouth as well, don't know how to solve this.

_________________
MachineMask
Life is good

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Newby, Review please

Post by Julie » Fri Sep 28, 2018 6:42 am

Back sleeping provokes more apneas... but if all other parts of the process work well and leaks are down, etc. it may be possible to do it if your AHI stays low and you feel ok.

STRONGSLTN
Posts: 29
Joined: Tue Aug 07, 2018 8:27 am
Location: Jekyll Island, Ga

Re: Newby, Review please

Post by STRONGSLTN » Fri Sep 28, 2018 6:55 am

Yup, three LL's last night but my resmed score was best ever.
Graphic_9_28_2018 8_46_40 AM.jpg
Graphic_9_28_2018 8_46_40 AM.jpg (86.87 KiB) Viewed 5185 times
Any suggestions on dry mouth problem. I think this is what wakes me up. Bathroom down to minimum now. I can consciously relube my mouth without demasking, but I would hope to prevent it from occuring.

_________________
MachineMask
Life is good