It has been awhile since I checked my data. ResScan has two options for downloading the data. You have to pick the one that selects all the detailed data.Uff Da wrote: ↑Sat Aug 18, 2018 11:47 amHere we go again. I did download Res Scan, but I'm not sure I know what I'm doing in it yet. Couldn't handle reading all the instructions, so just jumped in.
Last night my Sleepyhead showed an AHI OF 5.55. Res Scan and my machine show 10.3. But for the first time Sleepyhead shows me a detail page that is totally blank. No graphs shown. No events listed other than session start and end times.
ResScan shows the summary graphs page, but on the detailed graphs page just says, "No detailed data available." That's not just for last night, but for the entire time since I started the SD disk in June. What information should I be getting on this detailed graphs page?
Looking back over the previous dates that I had discrepancies, ResScan agrees with the figure shown on the display on my machine, not with the Sleepyhead totals for AHI.
So what's going on? If it is a Sleepyhead problem, why just me?
Major discrepancy between ResMed and Sleepyhead AHI
- zoocrewphoto
- Posts: 3732
- Joined: Mon Apr 30, 2012 10:34 pm
- Location: Seatac, WA
Re: Major discrepancy between ResMed and Sleepyhead AHI
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Resmed S9 autoset pressure range 11-17 |
Who would have thought it would be this challenging to sleep and breathe at the same time?
Re: Major discrepancy between ResMed and Sleepyhead AHI
Well, I just checked the download screen in ResScan again and this is what it shows. I selected the button on the device line. Shouldn't that have given me the complete download? But when I look at the "notes" section, it just lists "received summary data for" all the various dates.zoocrewphoto wrote: ↑Sat Aug 18, 2018 6:55 pmIt has been awhile since I checked my data. ResScan has two options for downloading the data. You have to pick the one that selects all the detailed data.
Okay, I just tried to download the SD to ResScan again and overwrite. But when I finished, I got a screen that said Summary data (download complete), Detailed data (not marked for download). So there must be some place else I have to select "detailed," but I can't find it. Grrrrrrr, I hate computers.
_________________
| Machine: DreamStation BiPAP® Auto Machine |
| Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
- Attachments
-
- ResScan download screen.png (40.4 KiB) Viewed 501 times
- BlueDragon
- Posts: 547
- Joined: Sun May 27, 2018 10:20 am
- Location: Sonoran Desert
Re: Major discrepancy between ResMed and Sleepyhead AHI
You see where it says "Data Summary data only"? Click on the Select button to the right, then specify that you want detailed data. ResScan will remember that for the future. Then re-import your card.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020. |
See OSCAR for the latest release.
OSCAR Team
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
OSCAR Team
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
Re: Major discrepancy between ResMed and Sleepyhead AHI
Make it look like this:BlueDragon wrote: ↑Sat Aug 18, 2018 9:13 pmYou see where it says "Data Summary data only"? Click on the Select button to the right, then specify that you want detailed data. ResScan will remember that for the future. Then re-import your card.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: Backup machine: AirSense 10 AutoSet with all the fixins |
- Attachments
-
- resscan download.PNG (57.49 KiB) Viewed 496 times
Re: Major discrepancy between ResMed and Sleepyhead AHI
So to get detailed information downloaded one needs to click on data summary data only. What bright individual came up with those instructions, I wonder. 
_________________
| Machine: DreamStation BiPAP® Auto Machine |
| Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
Re: Major discrepancy between ResMed and Sleepyhead AHI
Following those brilliant
instructions on ResScan, I was able to download some of my detailed data. Not much, though. I got a "download not responding" message after a bit four times and gave up. I think the detailed information got downloaded from the 10th of August until last night, so about a week.
I sure don't like how ResScan presents their graphs as well as Sleepyhead, but if SH continues to give me troubles, at least I may have a little more information than on the machine display alone.
I sure don't like how ResScan presents their graphs as well as Sleepyhead, but if SH continues to give me troubles, at least I may have a little more information than on the machine display alone.
_________________
| Machine: DreamStation BiPAP® Auto Machine |
| Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
Re: Major discrepancy between ResMed and Sleepyhead AHI
Just to repeat the PM I sent you... the data on your SD card is somewhat corrupted, in a way that SH can't compensate for, For instance, the Mask Pressure trace occasionally goes completely berserk, showing recorded pressures of 600+cm/h2o, of course, the machine can't do that, it'd have popped you like a balloon if it could... (assuming the mask kept a seal, which, I'm sure it wouldn'tUff Da wrote: ↑Sat Aug 18, 2018 4:34 pmWhat a generous offer! Thanks so much, PR. I don't know how to do the zip file thing, but I'll have my BF show me later. Later today, if he has the time.palerider wrote: ↑Sat Aug 18, 2018 2:39 pmYou're welcome to zip up your *entire* sd card and upload it to https://www.dropbox.com/request/0LalWvEsijRIgX8sdM7H
and I'll take a look at it, see if it matches what you have.
There's a lot of event flags that SH isn't picking up, but the apneas are clearly there when examining the waveform. Counting those up, matches pretty closely with the dates you gave me numbers from the machine.
Also, the places where SH says your Respiration rate (don't know if you looked at that) is 50.... it's actually around 16, but cardiac effects are causing your breath to hover back and forth over the 0 line, and SH is counting each heartbeat as a breath.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Major discrepancy between ResMed and Sleepyhead AHI
I'll reformat the disk and start all over. After all the hassles, it will feel refreshing to start over.palerider wrote: ↑Sat Aug 18, 2018 10:54 pmJust to repeat the PM I sent you... the data on your SD card is somewhat corrupted, in a way that SH can't compensate for, For instance, the Mask Pressure trace occasionally goes completely berserk, showing recorded pressures of 600+cm/h2o, of course, the machine can't do that, it'd have popped you like a balloon if it could... (assuming the mask kept a seal, which, I'm sure it wouldn't)
I never saw any pressures of 600 on my reports, did I do see that on my Sleepyhead statistics it showed max pressure at 24.52, and I know my machine doesn't go that high, so I knew something was amiss there.
When I finally was able to get some of the detailed disk information into ResScan, it showed the totals AHI the same as the machine said. It could apparently read that part of the disk whereas Sleepyhead couldn't. But the more detailed information was hit or miss. One day it read it and another it didn't. And last night was crazy. It uploaded to Sleepyhead just fine and the AHI matched the machine dial. But last night's session wouldn't upload to ResScan at all. Yep, reformat and start over.
Yes, I did look at that respiration chart and noted several times over the weeks that it couldn't possibly be right. I considered asking about it on the board, but since I've had more important things about this CPAP to figure out, decided it was low priority at that point. Interesting that the cardiac effects are influencing the way SH interprets it.
Thanks for all your help, PR. I'd be totally lost without the help of the good people on this board.
_________________
| Machine: DreamStation BiPAP® Auto Machine |
| Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
- zoocrewphoto
- Posts: 3732
- Joined: Mon Apr 30, 2012 10:34 pm
- Location: Seatac, WA
Re: Major discrepancy between ResMed and Sleepyhead AHI
Have you tried a different card? Maybe the card has defective spots on it, so it can't write the data correctly.
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Resmed S9 autoset pressure range 11-17 |
Who would have thought it would be this challenging to sleep and breathe at the same time?
Re: Major discrepancy between ResMed and Sleepyhead AHI
You'll have to zoom in, but look at the number on the mask pressure graph.Uff Da wrote: ↑Sun Aug 19, 2018 12:13 pmI'll reformat the disk and start all over. After all the hassles, it will feel refreshing to start over.palerider wrote: ↑Sat Aug 18, 2018 10:54 pmJust to repeat the PM I sent you... the data on your SD card is somewhat corrupted, in a way that SH can't compensate for, For instance, the Mask Pressure trace occasionally goes completely berserk, showing recorded pressures of 600+cm/h2o, of course, the machine can't do that, it'd have popped you like a balloon if it could... (assuming the mask kept a seal, which, I'm sure it wouldn't)
I never saw any pressures of 600 on my reports, did I do see that on my Sleepyhead statistics it showed max pressure at 24.52, and I know my machine doesn't go that high, so I knew something was amiss there.

You're quite welcome.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Major discrepancy between ResMed and Sleepyhead AHI
Wild! No, I never zoomed in and saw all those variations in the level. I just saw the line that went to the top of my graph and assumed it was like a previous line up and down a graph that I'd asked about which you said was probably just a glitch in the Sleepyhead software, so I didn't explore it any further.
So much to see in those Sleepyhead graphs! I'm having a hard time interpreting the wave forms on the flow rate charts, but feel I need to try to do a bit more studying of it all before I bother you good folks with more questions about it. But you'll likely be hearing from me about it eventually.
_________________
| Machine: DreamStation BiPAP® Auto Machine |
| Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
Re: Major discrepancy between ResMed and Sleepyhead AHI
I'd say that trying to get a sense of what's going on by looking at the breath shapes is one of the more involved things, much more detail than looking at the pressure curves, or leaks
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Major discrepancy between ResMed and Sleepyhead AHI
Saw this in one of Pugsy's posts and immediately saved it to my computer. It's helpful. There's much more to interpreting a chart, but this is handy.Uff Da wrote: ↑Sun Aug 19, 2018 12:13 pmSo much to see in those Sleepyhead graphs! I'm having a hard time interpreting the wave forms on the flow rate charts, but feel I need to try to do a bit more studying of it all before I bother you good folks with more questions about it. But you'll likely be hearing from me about it eventually.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: Backup machine: AirSense 10 AutoSet with all the fixins |
- Attachments
-
- flow rate chart.png (680.52 KiB) Viewed 435 times
Re: Major discrepancy between ResMed and Sleepyhead AHI
Thanks for that chart, djams. I'd seen the chart before, but the words identifying the difference between the lines on the chart weren't clear enough to read, so it wasn't of much help. I can read this one. Interesting in comparing what I see here to some of my breathing patterns last night.
_________________
| Machine: DreamStation BiPAP® Auto Machine |
| Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |