by DankShroud » Wed Jul 25, 2018 6:42 pm

- Screenshot (1).png (116.93 KiB) Viewed 3155 times

- Screenshot (2).png (108.16 KiB) Viewed 3154 times

- Screenshot (3).png (112.37 KiB) Viewed 3154 times
-
DankShroud
- Posts: 86
- Joined: Mon Feb 26, 2018 10:25 am
Post
by DankShroud » Wed Jul 25, 2018 6:44 pm

- Screenshot (4).png (110.26 KiB) Viewed 3152 times

- Screenshot (5).png (109.38 KiB) Viewed 3152 times

- Screenshot (6).png (112.09 KiB) Viewed 3152 times
-
D.H.
- Posts: 3535
- Joined: Sun Sep 04, 2016 7:07 pm
Post
by D.H. » Wed Jul 25, 2018 6:47 pm
No, they are not! "Clear Airway" means no obstruction was detected, and the presumption is that they are central (as in "Central Nervous System") events.
-
DankShroud
- Posts: 86
- Joined: Mon Feb 26, 2018 10:25 am
Post
by DankShroud » Wed Jul 25, 2018 6:55 pm
D.H. wrote: ↑Wed Jul 25, 2018 6:47 pm
No, they are not! "Clear Airway" means no obstruction was detected, and the presumption is that they are central (as in "Central Nervous System") events.
You're looking at the wrong things you goon. I'm talking about the flowrate flatlines that precede those swj centrals. I want to hear from palerider or pugsy
-
Pugsy
- Posts: 65447
- Joined: Thu May 14, 2009 9:31 am
- Location: Missouri, USA
Post
by Pugsy » Wed Jul 25, 2018 7:00 pm
Couple of them maybe...the rest SWJ or maybe post arousal.
They (the couple that maybe are obstructive) aren't long enough or enough of a reduction of air flow to earn the OA or hyponea flag.
This kind of microscopic work I am not particularly good at.
I may have to RISE but I refuse to SHINE.
-
Stom
- Posts: 383
- Joined: Sat Jul 21, 2018 9:44 am
Post
by Stom » Wed Jul 25, 2018 7:03 pm
D.H. wrote: ↑Wed Jul 25, 2018 6:47 pm
No, they are not! "Clear Airway" means no obstruction was detected, and the presumption is that they are central (as in "Central Nervous System") events.
The machine did not flag the events at all. So they aren't flagged as CAs and the machine doesn't state a conclusion for a forced oscillation test for whether the airway was obstructed. They could be unobstruced, but they might not be.
One thing to keep in mind is that neither event near User Flag 1 appears to be 10 seconds or more long, which is the requirement to be an apnea.
Official Foe™ on @Palerider's public list of foes.
Rules to live by: Be wary of anyone boasting their behavior is like that of a fictional sociopath.
-
palerider
- Posts: 32308
- Joined: Wed Dec 16, 2009 5:43 pm
- Location: Dallas(ish).
Post
by palerider » Wed Jul 25, 2018 7:05 pm
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
Pugsy
- Posts: 65447
- Joined: Thu May 14, 2009 9:31 am
- Location: Missouri, USA
Post
by Pugsy » Wed Jul 25, 2018 7:09 pm
I would help you up off the floor but I peed my pants laughing....gotta go change my drawers.
I may have to RISE but I refuse to SHINE.
-
palerider
- Posts: 32308
- Joined: Wed Dec 16, 2009 5:43 pm
- Location: Dallas(ish).
Post
by palerider » Wed Jul 25, 2018 7:16 pm
DankShroud wrote: ↑Wed Jul 25, 2018 6:55 pm
I want to hear from palerider or pugsy
Ok, best guess.. (are they obstructive)
Yes and no on the first chart, then
Yes.
Yes.
Yes.
No.
Not sure.
The only way i can get an idea is if the event is long enough for FOT to kick in and then look at the relative amounts of flow vs mask pressure response from the FOT... which is exactly what the machine does, only better.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.