CPAP Ramp Pressure start
- SirGaspAlot
- Posts: 143
- Joined: Mon Jul 16, 2018 7:33 am
- Location: Maryland, United States
CPAP Ramp Pressure start
What is a good start and ramp time? My therapy pressure is supposed to be 6cm-22 cm with PS of 4.
Thanks in advance.
Thanks in advance.
US Navy Retired E6, F-18 Aircraft Inspector/ejection seat inspector/mechanic and all around human being.
"If you always do what youve always done then youll always get what you always got"
"If you always do what youve always done then youll always get what you always got"
- Slartybartfast
- Posts: 1633
- Joined: Wed Sep 01, 2010 12:34 pm
Re: CPAP Ramp Pressure start
At 6 cm, you won't notice the pressure. I'd leave the ramp off and just strap on and start breathing. If you had a very high minimum pressure, like 10 or so, then maybe the ramp would keep you from being startled by the sudden inrush of air. The ramp feature appears to exist to help new users get used to the otherwise sudden onset of pressure. I tried it for a few nights, and found it wasn't necessary. My starting pressure was 8 cm.
Re: CPAP Ramp Pressure start
I've had the opposite experience. It's not an issue of being startled, but rather, being inflated like a balloon and struggling against the onslaught of pressure. I have found that, rather than allowing me to ease into sleep, I have to spend time constantly adjusting the mask to close leaks and the like, often finding my sleepy state of mind now being fully alert and frustrated.
I recognize that some folks think ramp usage is for sissies (at worst) or counterproductive, but I still find it a useful feature that let's me get into the rhythm of taking larger-than-normal breaths and ease off into dreamland while the CPAP ramps up.
Cheers!
I recognize that some folks think ramp usage is for sissies (at worst) or counterproductive, but I still find it a useful feature that let's me get into the rhythm of taking larger-than-normal breaths and ease off into dreamland while the CPAP ramps up.
Cheers!
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: Simplus Full Face CPAP Mask with Headgear |
Humidifier: DreamStation Heated Humidifier |
Just an average joe on his way from here to there...
- SirGaspAlot
- Posts: 143
- Joined: Mon Jul 16, 2018 7:33 am
- Location: Maryland, United States
Re: CPAP Ramp Pressure start
Thanks, Ill bump it to 6 I think, Its at 4 for 30 minutes.
US Navy Retired E6, F-18 Aircraft Inspector/ejection seat inspector/mechanic and all around human being.
"If you always do what youve always done then youll always get what you always got"
"If you always do what youve always done then youll always get what you always got"
Re: CPAP Ramp Pressure start
Try 6 but set it at 20 minutes or less, if possible. 30 minutes sounds more like the main highway than the ramp. 

_________________
Machine: DreamStation Auto CPAP Machine |
Mask: Simplus Full Face CPAP Mask with Headgear |
Humidifier: DreamStation Heated Humidifier |
Just an average joe on his way from here to there...
Re: CPAP Ramp Pressure start
Like Slartybartfast, I found the ramp rather useless, so I dropped it. Mine, however, was set at 4 on auto, and since I didn't get to sleep for hours at first, I don't know how long it was before the machine was tricked into believing I was asleep and started building the pressure, set for 8-18. I now just let it start at 8 without ramp and it works for me. I might even decide to bump it up to 9. I'm sure what is comfortable will vary from one individual to another.
_________________
Machine: DreamStation BiPAP® Auto Machine |
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
Humidifier: DreamStation Heated Humidifier |
- Okie bipap
- Posts: 3566
- Joined: Thu Oct 15, 2015 4:14 pm
- Location: Central Oklahoma
Re: CPAP Ramp Pressure start
I would recommend you try without the ramp feature being used. I presume four setting is minimum EPAP of 6, and maximum IPAP of 23. If this is the case, ramp will start you off at minimum EPAP of 4 and IPAP of 8. This only two cm under your treatment pressure which is a very small difference.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Evora Full Face Mask - Fitpack |
Additional Comments: IPAP 20-25, ps 4, OSCAR software |
Growing old is mandatory, but growing up is optional.
Re: CPAP Ramp Pressure start
You don't need ramp for a 6 min pressure.SirGaspAlot wrote: ↑Mon Jul 23, 2018 12:05 pmWhat is a good start and ramp time? My therapy pressure is supposed to be 6cm-22 cm with PS of 4.
Thanks in advance.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Gator5000e
- Posts: 134
- Joined: Mon May 07, 2018 1:07 pm
Re: CPAP Ramp Pressure start
How do you all feel about the Smart Ramp feature on the Dreamstation? I saw it described as follows:
"The Smart RAMP feature. This neat little feature monitors your breathing, and can tell when you go to sleep. When it detects you fall asleep, the AutoSet model goes into ‘active’ mode and can respond to any respiratory events that it detects. Previous models did not do this and would either ramp up too quickly (making it difficult to fall asleep), or too slowly (meaning sleep apnea events would go both undetected and untreated during this period). An excellent feature, and I recommend leaving it on."
Anyone have any thoughts on this feature? The recommendation came from this review (A couple of years old now but the machine doesn't seem to have changed).
https://cpapreviews.org/respironics-dre ... th-review/
"The Smart RAMP feature. This neat little feature monitors your breathing, and can tell when you go to sleep. When it detects you fall asleep, the AutoSet model goes into ‘active’ mode and can respond to any respiratory events that it detects. Previous models did not do this and would either ramp up too quickly (making it difficult to fall asleep), or too slowly (meaning sleep apnea events would go both undetected and untreated during this period). An excellent feature, and I recommend leaving it on."
Anyone have any thoughts on this feature? The recommendation came from this review (A couple of years old now but the machine doesn't seem to have changed).
https://cpapreviews.org/respironics-dre ... th-review/
Re: CPAP Ramp Pressure start
Auto ramp is a Resmed feature, not respironics.Gator5000e wrote: ↑Mon Jul 23, 2018 3:14 pmHow do you all feel about the Smart Ramp feature on the Dreamstation? I saw it described as follows:
"The Smart RAMP feature. This neat little feature monitors your breathing, and can tell when you go to sleep. When it detects you fall asleep, the AutoSet model goes into ‘active’ mode and can respond to any respiratory events that it detects. Previous models did not do this and would either ramp up too quickly (making it difficult to fall asleep), or too slowly (meaning sleep apnea events would go both undetected and untreated during this period). An excellent feature, and I recommend leaving it on."
Anyone have any thoughts on this feature? The recommendation came from this review (A couple of years old now but the machine doesn't seem to have changed).
https://cpapreviews.org/respironics-dre ... th-review/
The article is wrong. As evidenced by " the AutoSet model" which is a resmed machine.
Last edited by palerider on Mon Jul 23, 2018 8:06 pm, edited 1 time in total.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Gator5000e
- Posts: 134
- Joined: Mon May 07, 2018 1:07 pm
Re: CPAP Ramp Pressure start
Ok. I do have a feature in the clinical settings called SmartRamp on my biPAP Pro machine I just got. Is it something different?
Re: CPAP Ramp Pressure start
Maybe I'm wrong, I'll have to check the manual tonight.Gator5000e wrote: ↑Mon Jul 23, 2018 3:45 pmOk. I do have a feature in the clinical settings called SmartRamp on my biPAP Pro machine I just got. Is it something different?
I'd be extremely surprised to have respironics name a feature on their machine the same as what resmed has had though...
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: CPAP Ramp Pressure start
Why not just try 'no' ramp at all for a night... I think the majority of us don't use it beyond the first few wks, if that, unless at a very high setting - and yours is very low.
-
- Posts: 2744
- Joined: Tue Oct 12, 2010 6:42 pm
Re: CPAP Ramp Pressure start
I don't think it takes long to adjust. While ramp is a comfort feature you don't need a lot of time especially if the pressure differential isn't that much. 10 minutes or maybe even less is all that is needed in most cases.
For example: my minimum pressure is 12. I use ramp because I enjoy being comfortable. My ramp starts at 8 and runs for 10 minutes. If I still happen to be awake at the end of the ramp period I am totally unaware of any change in pressure and am totally comfortable.
For example: my minimum pressure is 12. I use ramp because I enjoy being comfortable. My ramp starts at 8 and runs for 10 minutes. If I still happen to be awake at the end of the ramp period I am totally unaware of any change in pressure and am totally comfortable.
_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier. |
SpO2 96+% and holding...
Re: CPAP Ramp Pressure start
Ok, according to the manual, Respironics "SmartRamp" is:Gator5000e wrote: ↑Mon Jul 23, 2018 3:45 pmOk. I do have a feature in the clinical settings called SmartRamp on my biPAP Pro machine I just got. Is it something different?
Vs Resmed, (which doesn't call their feature "smart ramp"... they call it "auto".)Respironics wrote:When SmartRamp mode is enabled, the therapy device’s ramp function utilizes an Auto titrating algorithm during the ramp period. It allows patients the ability to stay at lower pressures during the ramp period, to improve their acclimation to therapy.
SmartRamp mode functions differently, depending on the therapy mode that the device is using.
*In CPAP or CPAP-Check mode, the SmartRamp applies the Auto-CPAP algorithm during the ramp period. The Ramp Start pressure becomes the Auto Minimum pressure during the ramp period. The Auto Maximum pressure during ramp is the CPAP or CPAP-Check pressure.
*In Auto mode, the SmartRamp applies the Auto-CPAP algorithm during the ramp period. The Ramp Start pressure becomes the Auto Minimum pressure during the ramp period. The Auto Maximum pressure during ramp is the Auto Minimum under normal Auto mode.
*In BiPAP or Auto-BiPAP mode, the SmartRamp applies the a modified version of the Auto-BiPAP algorithm during the ramp period. The Ramp Start pressure becomes the EPAP Minimum pressure, and the Pressure Support Minimum pressured is applied. The IPAP Maximum pressure during ramp is the EPAP or EPAP Minimum under normal BiPAP or Auto-BiPAP mode.
The SmartRamp period will terminate in either of two ways:
1) If SmartRamp pressure reaches the minimum pressure of the therapy mode selected, then SmartRamp ends, and the device continues to deliver therapy under the selected therapy mode, or:
2) If SmartRamp pressure does not reach the minimum pressure of the therapy mode selected by the end of the Ramp Time, then pressure is increased at a rate of approximately 1 cm H 2 O per minute. Once the pressure reaches the minimum pressure of the therapy mode selected, then the device will continue to deliver therapy for that mode.
If SmartRamp mode is not enabled, then the standard, linear pressure ramp mode is active.
So, back to my initial... "That's not a Respironics feature, the article is wrong.Resmed wrote:Ramp Time can be set to Off, 5 to 45 minutes or Auto. When Ramp Time is set to Auto, the device will detect sleep onset and then gradually increase from the start pressure to the minimum treatment pressure at a rate of 1 cm H 2 O per minute. However, if sleep onset is not detected, the device will reach the target pressure within 30 minutes.