Mostly Centrals

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
zzzk
Posts: 42
Joined: Sun Feb 24, 2013 5:30 pm

Re: Mostly Centrals

Post by zzzk » Thu Jul 19, 2018 9:24 pm

Ok, what do you think, then? Try ps0 with the new machine? would that be equivalent to my old machine? If that doesn't work, different machine? Thank you for your help.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Mostly Centrals

Post by palerider » Thu Jul 19, 2018 11:22 pm

zzzk wrote:
Thu Jul 19, 2018 9:24 pm
Ok, what do you think, then? Try ps0 with the new machine? would that be equivalent to my old machine? If that doesn't work, different machine? Thank you for your help.
palerider wrote:
Thu Jul 19, 2018 12:56 pm
zzzk wrote:
Thu Jul 19, 2018 12:28 pm
Hello, here is last night at ps1, does not seem to be improving... Recommendations?

Thank you
The only other thing to try is a PS of zero.

Then the next step is a different machine.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

zzzk
Posts: 42
Joined: Sun Feb 24, 2013 5:30 pm

Re: Mostly Centrals

Post by zzzk » Fri Jul 20, 2018 12:40 pm

Ok, ps 0 attached. It is better, but maybe not enough?
Attachments
2018.07.20a.jpg
2018.07.20a.jpg (311.84 KiB) Viewed 14162 times
2018.07.20b.jpg
2018.07.20b.jpg (76.05 KiB) Viewed 14162 times

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Mostly Centrals

Post by palerider » Fri Jul 20, 2018 1:39 pm

Not nearly enough, you're going to need a different machine to avoid what are probably some significant o2 desaturations...

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

zzzk
Posts: 42
Joined: Sun Feb 24, 2013 5:30 pm

Re: Mostly Centrals

Post by zzzk » Fri Jul 20, 2018 1:45 pm

Ok, any recomendations of machines?

Maybe a dumb question: Is my apnea worse with this machine or better with nothing until I get a new one? Thank you.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Mostly Centrals

Post by palerider » Fri Jul 20, 2018 2:03 pm

zzzk wrote:
Fri Jul 20, 2018 1:45 pm
Ok, any recomendations of machines?

Maybe a dumb question: Is my apnea worse with this machine or better with nothing until I get a new one? Thank you.
Resmed s9 adapt, or aircurve 10 asv.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

zzzk
Posts: 42
Joined: Sun Feb 24, 2013 5:30 pm

Re: Mostly Centrals

Post by zzzk » Fri Jul 20, 2018 4:58 pm

Could you give an example of a prescription for the resmed aircurve 10 asv auto (that's the one you recommended). My dr kept getting it wrong on the 1st one. I know it will only be an example, I will talk to the dr. Here's my prescription for the current one:

auto bipap setup
min eppap = 8
max ipap = 25
min ps 4
max ps 8

dx: osa

What would it look like on the asv?
Or if you could tell me recommended settings, that would be better.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Mostly Centrals

Post by palerider » Fri Jul 20, 2018 5:57 pm

zzzk wrote:
Fri Jul 20, 2018 4:58 pm
Could you give an example of a prescription for the resmed aircurve 10 asv auto (that's the one you recommended). My dr kept getting it wrong on the 1st one. I know it will only be an example, I will talk to the dr. Here's my prescription for the current one:

auto bipap setup
min eppap = 8
max ipap = 25
min ps 4
max ps 8

dx: osa

What would it look like on the asv?
Or if you could tell me recommended settings, that would be better.
Well, putting "Aircurve 10 asv" on the rx would help.

MinEPAP 8, MaxIPAP 25, MinPS 4 MaxPS 17

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

zzzk
Posts: 42
Joined: Sun Feb 24, 2013 5:30 pm

Re: Mostly Centrals

Post by zzzk » Fri Jul 20, 2018 6:40 pm

Thank you very much for all of your help. I've been having a hard time with this for years. Hopefully an asv will solve it.

Just so I understand a little more. When I changed the ps to 0, was it then technically no longer a bipap, but only an auto machine between pressure ranges, the same for inhale and exhale? Or not?

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Mostly Centrals

Post by palerider » Fri Jul 20, 2018 7:26 pm

zzzk wrote:
Fri Jul 20, 2018 6:40 pm
When I changed the ps to 0, was it then technically no longer a bipap,
Well, it's technically still a bilevel machine.
zzzk wrote:
Fri Jul 20, 2018 6:40 pm
but only an auto machine between pressure ranges, the same for inhale and exhale? Or not?
It was a bilevel that was acting the as an auto cpap.

The difference between 'bilevel' and 'cpap' function gets vague with 'exhale relief' on cpaps and low PS on the bilevels.

But, in essence, your understanding is spot on.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

zzzk
Posts: 42
Joined: Sun Feb 24, 2013 5:30 pm

Re: Mostly Centrals

Post by zzzk » Fri Aug 10, 2018 1:22 pm

Well, it's only the first night with my new ASV, and I still only sleep for 1-2 hours, then awake 1/2 hour or so, then sleep 1-2 hours again... But my AHI is down to 0 !!!! Wild swings in pressure...
Conclusions? Attached sleephead...
Attachments
CPAP_ASV01.jpg
CPAP_ASV01.jpg (302.01 KiB) Viewed 14034 times

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Mostly Centrals

Post by palerider » Fri Aug 10, 2018 1:30 pm

zzzk wrote:
Fri Aug 10, 2018 1:22 pm
Well, it's only the first night with my new ASV, and I still only sleep for 1-2 hours, then awake 1/2 hour or so, then sleep 1-2 hours again... But my AHI is down to 0 !!!! Wild swings in pressure...
Conclusions? Attached sleephead...
All those wild swings are the machine preventing centrals.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Midwest_non_sleeper
Posts: 436
Joined: Wed Jun 20, 2018 1:03 pm

Re: Mostly Centrals

Post by Midwest_non_sleeper » Fri Aug 10, 2018 2:06 pm

palerider wrote:
Fri Aug 10, 2018 1:30 pm
zzzk wrote:
Fri Aug 10, 2018 1:22 pm
Well, it's only the first night with my new ASV, and I still only sleep for 1-2 hours, then awake 1/2 hour or so, then sleep 1-2 hours again... But my AHI is down to 0 !!!! Wild swings in pressure...
Conclusions? Attached sleephead...
All those wild swings are the machine preventing centrals.
Holy crap, well done PR, all those centrals just *poofed*.

Glad you got it figured out zzzk!

zzzk
Posts: 42
Joined: Sun Feb 24, 2013 5:30 pm

Re: Mostly Centrals

Post by zzzk » Fri Aug 10, 2018 2:11 pm

palerider wrote:
Fri Aug 10, 2018 1:30 pm
zzzk wrote:
Fri Aug 10, 2018 1:22 pm
Well, it's only the first night with my new ASV, and I still only sleep for 1-2 hours, then awake 1/2 hour or so, then sleep 1-2 hours again... But my AHI is down to 0 !!!! Wild swings in pressure...
Conclusions? Attached sleephead...
All those wild swings are the machine preventing centrals.
Thank you very much! I'm trying not to get too excited too soon, but it looks like it may be fixed with this machine / mask! I've been dealing with this for about 20 years! Has this tech come around only recently?

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Mostly Centrals

Post by palerider » Fri Aug 10, 2018 2:52 pm

zzzk wrote:
Fri Aug 10, 2018 2:11 pm
palerider wrote:
Fri Aug 10, 2018 1:30 pm
zzzk wrote:
Fri Aug 10, 2018 1:22 pm
Well, it's only the first night with my new ASV, and I still only sleep for 1-2 hours, then awake 1/2 hour or so, then sleep 1-2 hours again... But my AHI is down to 0 !!!! Wild swings in pressure...
Conclusions? Attached sleephead...
All those wild swings are the machine preventing centrals.
Thank you very much! I'm trying not to get too excited too soon, but it looks like it may be fixed with this machine / mask! I've been dealing with this for about 20 years! Has this tech come around only recently?
In the overall scheme of things, yes, the asv is one of the most recent "big things", the other is *VAPS. Both respond breath by breath. The first asvs started appearing in the 2000s. The earliest had an extra tube running back to the machine to directly measure mask pressure.

As far as your machine is concerned, you can't have a central, because when your tidal volume decreases, it boosts pressure support to maintain your ventilation. You can see this in action I'd you bring the mask pressure chart up under flow rate, and then zoom in (up arrow) enough to see the individual breaths.