Left Power Supply Behind

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
unityco
Posts: 43
Joined: Mon Nov 30, 2009 9:07 pm
Location: Toronto

Left Power Supply Behind

Post by unityco » Thu Jul 12, 2018 11:12 am

Hi all,

I’m away from home in a somewhat remote setting (no CPAP suppliers or fast shipping options) and I left the power brick for my REMstar at home and can’t get it for 6 weeks. >.<

Can I just use a universal laptop power supply (which I can buy today, assuming it has the appropriate options?) Any other ideas?

Thank you!

User avatar
Pugsy
Posts: 65382
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Left Power Supply Behind

Post by Pugsy » Thu Jul 12, 2018 11:26 am

Which Remstar model machine?

You will need some specific power brick requirements as well as specific adapter plug in thing to attach to your Remstar.
Maybe take your machine with you to make sure it works????

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

unityco
Posts: 43
Joined: Mon Nov 30, 2009 9:07 pm
Location: Toronto

Re: Left Power Supply Behind

Post by unityco » Thu Jul 12, 2018 11:42 am

These are the details (in the photo...)
51792B73-C65F-481E-A34F-EAFF5C58D811.jpeg
51792B73-C65F-481E-A34F-EAFF5C58D811.jpeg (882.39 KiB) Viewed 1636 times
seems it would be within the requirements of many computer power supplies?

User avatar
Pugsy
Posts: 65382
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Left Power Supply Behind

Post by Pugsy » Thu Jul 12, 2018 12:27 pm

It's a PR System One 50 series machine and the plug in for it is different than the same model but in the 60 series model line.

I think it probably is a 60 watt brick (100 to 240 volt & 50 to 60 HZ) but I no longer have a System One 50 series brick but I do know that at the machine itself the connectors differed where the cords actually plugged into the machine. When I went from the 50 series to the 60 series nothing was interchangeable except the cheap little power cord from the brick to the wall outlet.

https://www.cpap.com/productpage/respir ... upply.html
The 50 series (the 551) tells me that you have the first System One model series which was the 50 series...the 60 series came later.

https://www.directhomemedical.com/power ... -cpap.html

The power cord from the brick to the electric source at the wall...those are easy to find. It's the brick itself and the connecting part that plugs into the machine that is the hard part.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: Left Power Supply Behind

Post by prodigyplace » Thu Jul 12, 2018 12:59 pm

Pugsy wrote:
Thu Jul 12, 2018 12:27 pm
It's a PR System One 50 series machine and the plug in for it is different than the same model but in the 60 series model line.

I think it probably is a 60 watt brick (100 to 240 volt & 50 to 60 HZ) but I no longer have a System One 50 series brick but I do know that at the machine itself the connectors differed where the cords actually plugged into the machine. When I went from the 50 series to the 60 series nothing was interchangeable except the cheap little power cord from the brick to the wall outlet.

https://www.cpap.com/productpage/respir ... upply.html
The 50 series (the 551) tells me that you have the first System One model series which was the 50 series...the 60 series came later.

https://www.directhomemedical.com/power ... -cpap.html

The power cord from the brick to the electric source at the wall...those are easy to find. It's the brick itself and the connecting part that plugs into the machine that is the hard part.
Even if the connector fits, you need to be sure the DC polarity is correct. At best, the CPAP would not work. At worst, it could damage the CPAP machine.

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.

User avatar
Pugsy
Posts: 65382
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Left Power Supply Behind

Post by Pugsy » Thu Jul 12, 2018 1:15 pm

prodigyplace wrote:
Thu Jul 12, 2018 12:59 pm
Even if the connector fits, you need to be sure the DC polarity is correct. At best, the CPAP would not work. At worst, it could damage the CPAP machine.
Yep, which is why I said take the machine with them and see if it works at all...and was silently saying "cross your fingers that you don't smoke the machine".
I don't have a laptop that has a brick powerful enough to power up a cpap machine. Maybe they are out there...I haven't looked so I can't say that they don't exist.

Now if the OP was using a 60 series machine...I actually have a spare brick for it but it won't hook up with the 50 machine at the machine level. The little tiny pin things are different sizes.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

TheDuke
Posts: 259
Joined: Tue Oct 27, 2009 10:23 pm
Location: Western Kentucky
Contact:

Re: Left Power Supply Behind

Post by TheDuke » Thu Jul 12, 2018 7:29 pm

The post regarding left behind power brick revives one of my major gripes' There is absolutely no reason for CPAP manufacturers to use an external power brick; it is simply a nuisance. My last two machines have been from DeVilbiss precisely because I wanted the freedom to never bother to be sure to pack a power brick. I had that same experience of leaving an external brick at home quite a few years back when were traveling to a rather remote area in Northern New Mexico for a somewhat extended vacation. And don't bother to say that the machine will overheat if there is no external power supply-my machine stays cool as the proverbial cucumber.

TheDuke

unityco
Posts: 43
Joined: Mon Nov 30, 2009 9:07 pm
Location: Toronto

Re: Left Power Supply Behind

Post by unityco » Thu Jul 12, 2018 8:10 pm

Thanks for the help, all! As it turns out, where I’m at DOES have a CPAP supply place, and they had a used brick for $20! Crisis averted!

Oh, and I agree with you, TheDuke! My previous machine was a ResMed and had a standard 2-prong power cable, no brick. Easy to replace if forgotten (sometimes just by canabalizing stuff we already had in hand!)

User avatar
Pugsy
Posts: 65382
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Left Power Supply Behind

Post by Pugsy » Thu Jul 12, 2018 8:33 pm

$20 bucks...the best 20 dollars you have spent in a long time.

I once made a trip to Las Vegas...holiday weekend...arrived on Friday night and at 11 PM realized I had forgot the long hose (thought I had the spare hose in the travel bag with back up machine but I had forgotten that I had removed it and didn't check the back since normally I keep the travel back up machine ready to go).
So lots of cpap supply places in Las Vegas but none open on a Sat the Fourth of July....finally found one medical supply place that was open and they sold all sorts of cpap stuff along with all sorts of medical supply stuff...and I paid $30 for a cheap generic hose that is $6 online and delivered and it was the best 30 bucks I have ever spent. I had one night without cpap...didn't want any more. I would have paid twice that for it. :lol:

Glad your crisis was averted and you don't have to risk blowing up your machine trying something that wasn't designed for your machine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Left Power Supply Behind

Post by palerider » Thu Jul 12, 2018 8:48 pm

TheDuke wrote:
Thu Jul 12, 2018 7:29 pm
There is absolutely no reason for CPAP manufacturers to use an external power brick;
Valid reasons for an external power brick:
  • Smaller footprint on the bedside table, look how big the power brick is, try and cram that inside an air10.
  • Less heat in the unit itself, possibly longer life for the cpap.
  • Power supply dies, plug in a spare one, verses a dead cpap and having to get a new one of those.
You could make the "absolutely no reason" statement for a laptop computer.. but when was the last time you saw one of those with a built in power supply?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Left Power Supply Behind

Post by Goofproof » Thu Jul 12, 2018 9:21 pm

TheDuke wrote:
Thu Jul 12, 2018 7:29 pm
The post regarding left behind power brick revives one of my major gripes' There is absolutely no reason for CPAP manufacturers to use an external power brick; it is simply a nuisance. My last two machines have been from DeVilbiss precisely because I wanted the freedom to never bother to be sure to pack a power brick. I had that same experience of leaving an external brick at home quite a few years back when were traveling to a rather remote area in Northern New Mexico for a somewhat extended vacation. And don't bother to say that the machine will overheat if there is no external power supply-my machine stays cool as the proverbial cucumber.

TheDuke
If you don't remove the cords and power brick from the machine, it will be in the bag with the machine. Best to have a second machine set up in your GO Bag with parts needed ready to go for traveling. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
zoocrewphoto
Posts: 3732
Joined: Mon Apr 30, 2012 10:34 pm
Location: Seatac, WA

Re: Left Power Supply Behind

Post by zoocrewphoto » Thu Jul 12, 2018 11:32 pm

Before I got a backup/travel machine, the biggest problem with cpap and traveling was getting home and being too tired to set up the machine. The problem is t hat my outlet isn't easy to get to, so plugging in that cord was a problem. Then I realized that the cord from the wall to the transformer was a very standard cord. So, I got an extra one and left that always plugged in. I kept another one in the bag to travel.

Now that I have a second machine, I have a full setup with extra power cord, extra mask, etc. So, it is complete. I leave the whole setup in my luggage so that it is there when it is time to pack.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 autoset pressure range 11-17
Who would have thought it would be this challenging to sleep and breathe at the same time?

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Left Power Supply Behind

Post by palerider » Fri Jul 13, 2018 12:38 am

zoocrewphoto wrote:
Thu Jul 12, 2018 11:32 pm
Then I realized that the cord from the wall to the transformer was a very standard cord. So, I got an extra one and left that always plugged in. I kept another one in the bag to travel.
Clever.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zoocrewphoto
Posts: 3732
Joined: Mon Apr 30, 2012 10:34 pm
Location: Seatac, WA

Re: Left Power Supply Behind

Post by zoocrewphoto » Fri Jul 13, 2018 12:45 am

palerider wrote:
Fri Jul 13, 2018 12:38 am
zoocrewphoto wrote:
Thu Jul 12, 2018 11:32 pm
Then I realized that the cord from the wall to the transformer was a very standard cord. So, I got an extra one and left that always plugged in. I kept another one in the bag to travel.
Clever.
There was also one time, before the backup/travel machine when I forgot that part of the cord. I happen to travel with a canon printer, and it uses the same type of cord. I had to keep a note on my cat carrier to remind me to take the cord back and forth so that I could use the cpap in the hotel and the printer at the cat show.

There have also been times when somebody has been able to borrow a spare power cord. I know I loaned one once to somebody who was in my area for a week. I think I have seen other times where somebody posted their need, and somebody was nearby and able to loan them the needed cord or part. It doesn't hurt to ask.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 autoset pressure range 11-17
Who would have thought it would be this challenging to sleep and breathe at the same time?

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: Left Power Supply Behind

Post by prodigyplace » Fri Jul 13, 2018 3:29 am

palerider wrote:
Thu Jul 12, 2018 8:48 pm
TheDuke wrote:
Thu Jul 12, 2018 7:29 pm
There is absolutely no reason for CPAP manufacturers to use an external power brick;
Valid reasons for an external power brick:
  • Smaller footprint on the bedside table, look how big the power brick is, try and cram that inside an air10.
  • Less heat in the unit itself, possibly longer life for the cpap.
  • Power supply dies, plug in a spare one, verses a dead cpap and having to get a new one of those.
You could make the "absolutely no reason" statement for a laptop computer.. but when was the last time you saw one of those with a built in power supply?
One more:
Easier to adapt to a DC power source. Just use a different brick.

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.