New Dreamstation and Wisp mask and AHIs jumped significantly

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
CpapWife
Posts: 74
Joined: Wed Sep 24, 2008 6:34 pm
Location: Arizona

New Dreamstation and Wisp mask and AHIs jumped significantly

Post by CpapWife » Mon Jul 09, 2018 2:26 am

DH got a new apap. He stayed with the Dreamstation as he has had the Phillips Respironics machines since he was diagnosed. Last one was the System One Remstar.
He did change masks also to a wisp mask. He has used many over the years (about 10 years now), using a nasal with the pillow over his nose and also a full face mask when his nose is stuffed up.

Since starting apap therapy, he went from about 91 ahi to about 1 (typically between .5 and 1.3).

With his new dreamstation, while a couple days the first week the AHIs were about 1.5 or 2.5, most days he is between 4-11 AHI. There doesn't seem to be a pattern but it might be positional or REM sleep (original diagnosis thought he had more during REM sleep). He is primarily hypoapnea with occasional central/clear airway. I do notice more snoring with the new setup but it is pretty sparse and there was some with the old setup.

Attached is:
Encore report for 3/10/18 of his older setup - 1.4 AHI (this was one we had handy).
lower ahi.png
lower ahi.png (200.67 KiB) Viewed 701 times
Sleepytime report for 7/4/18 of new setup - 10.27 AHI (a high AHI day).
high ahi.png
high ahi.png (267.91 KiB) Viewed 701 times
Sleepytime report for 7/6/18 of new setup - 3.40 AHI (a low AHI day).
old ahi.png
old ahi.png (230.47 KiB) Viewed 701 times
Hypoapneas have always been higher than other types.
DH does shift work so sleeps early morning to early afternoon typically.
Other than apap and mask no other real differences.

Thanks for any insights.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack
Humidifier: DreamStation Heated Humidifier
Additional Comments: Invocare O2 concentrator - 3 LPM. Wife to new cpap user Sept 2008
Last edited by CpapWife on Mon Jul 09, 2018 3:27 pm, edited 1 time in total.

User avatar
zonker
Posts: 11358
Joined: Fri Jun 19, 2015 4:36 pm

Re: New Dreamstation and Wisp mask and AHIs jumped significantly

Post by zonker » Mon Jul 09, 2018 11:26 am

sorry, but no reports came through. you may want to try again.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

CpapWife
Posts: 74
Joined: Wed Sep 24, 2008 6:34 pm
Location: Arizona

Re: New Dreamstation and Wisp mask and AHIs jumped significantly

Post by CpapWife » Mon Jul 09, 2018 3:46 pm

Sorry, attachments now there. This is my first time doing attachments though I read the thread on it, so anything else needed just let me know.

Doing one for last night - I noticed more reports including a flow report.
More charts.png
More charts.png (197.36 KiB) Viewed 700 times
Also, I noticed last night that DH was laying on his side and there was a quiet hiss from the mask at 8:43 am and also 11:40 am. These seemed to be high AHI times. We will do more tests with him laying in different positions and pretending to sleep. He uses a buckwheat pillow - one many cpap users use.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack
Humidifier: DreamStation Heated Humidifier
Additional Comments: Invocare O2 concentrator - 3 LPM. Wife to new cpap user Sept 2008

CpapWife
Posts: 74
Joined: Wed Sep 24, 2008 6:34 pm
Location: Arizona

Re: New Dreamstation and Wisp mask and AHIs jumped significantly

Post by CpapWife » Tue Jul 10, 2018 3:56 pm

OK, a big part of the issue is solved.
DH used an old mask, not his new wisp mask, with the dreamstation and got a low AHI. So next step checking why that mask is not working.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack
Humidifier: DreamStation Heated Humidifier
Additional Comments: Invocare O2 concentrator - 3 LPM. Wife to new cpap user Sept 2008

D.H.
Posts: 3532
Joined: Sun Sep 04, 2016 7:07 pm

Re: New Dreamstation and Wisp mask and AHIs jumped significantly

Post by D.H. » Tue Jul 10, 2018 4:07 pm

CpapWife wrote:
Tue Jul 10, 2018 3:56 pm
OK, a big part of the issue is solved.
DH used an old mask, not his new wisp mask, with the dreamstation and got a low AHI. So next step checking why that mask is not working.
Yes, to the extent you can help it, never change more than one thing at a time. Otherwise, if you have an issue, you won't know what caused it.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max