FLASHING Snowflake Icon ❄️

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Xrayvision
Posts: 1
Joined: Thu Jul 05, 2018 9:30 am

FLASHING Snowflake Icon ❄️

Post by Xrayvision » Thu Jul 05, 2018 9:42 am

HELP! over the past three days I have noticed a flashing icon resembling a snowflake on my ResMed AirSense10 unit. I have consulted the manual but cannot find out what it is. I’ve also searched online and still no explanation. I do not have access to manufacturer technical support. The unit seems to be functioning properly for now but the flashing icon is worrisome to me. Thank you for any suggestions that you can offer.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: FLASHING Snowflake Icon ❄️

Post by chunkyfrog » Thu Jul 05, 2018 12:29 pm

Deleted, due to bad guess.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her
Last edited by chunkyfrog on Thu Jul 05, 2018 5:40 pm, edited 1 time in total.

User avatar
LSAT
Posts: 13403
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: FLASHING Snowflake Icon ❄️

Post by LSAT » Thu Jul 05, 2018 12:41 pm

Here is the Clinical Manual for your machine.....At the bottom Page 3 it shows the snowflake.
https://sleep.tnet.com/home/files/resme ... -guide.pdf

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: FLASHING Snowflake Icon ❄️

Post by prodigyplace » Thu Jul 05, 2018 1:34 pm

The snowflake just indicates the machine is in cool-down mode after you have shut it off. It fully turns off after 5 minutes, I believe. Your DME should have mentioned that.

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: FLASHING Snowflake Icon ❄️

Post by palerider » Thu Jul 05, 2018 2:17 pm

Xrayvision wrote:
Thu Jul 05, 2018 9:42 am
HELP! over the past three days I have noticed a flashing icon resembling a snowflake on my ResMed AirSense10 unit. I have consulted the manual but cannot find out what it is.
Page 3, at the bottom. "Humidifier cooling".

https://www.respshop.com/manuals/ResMed ... %20her.pdf

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Janknitz
Posts: 8518
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: FLASHING Snowflake Icon ❄️

Post by Janknitz » Thu Jul 05, 2018 5:10 pm

Don't feel bad, it took me a couple of days to figure that out too, and it bugged me until I did. :wink:
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: FLASHING Snowflake Icon ❄️

Post by chunkyfrog » Thu Jul 05, 2018 5:39 pm

chunkyfrog wrote:
Thu Jul 05, 2018 12:29 pm
Maybe a problem with either the heated humidifier or the Climateline hose.
See the dealer.
Oopsy!
Removing bad guess . . .

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her