Effective O2 percentage with added O2??

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Effective O2 percentage with added O2??

Post by JDS74 » Fri May 25, 2018 8:53 am

I'm trying to determine the effective oxygen concentration in the air I breathe with my CPAP given 3 Liters/minutes O2 flow.

Here's what I think is true - please post corrections if you see errors. An explanation of the error would also be appreciated.
1) Normal O2 concentration in room air is 21%
2) Minute Vent value is in liters/minute
3) Unintentional leak rate is in liters/minute
4) Supplemental oxygen is 100% oxygen or close enough to not matter
5) Supplemental oxygen flow rate adds to the overall flow through the CPAP

Here are my calculations:
Total room air flow = Minute Vent + Unintentional Leak Rate or 12.2 + 9.9 lpm or 22.1 lpm (for me last night, data from Encore)
Total room O2 flow = O2% * Total Flow or 22.1 * 21% or 4.7 lpm
Supplemental O2 flow = 3 lpm (measured on my O2 concentrator)

Final O2 % for combined air flow = (total room O2 flow + Supplemental O2 flow)/(Total room air flow + Supplemtal O2 flow)
or (4.7 + 3.0)/(22.1 + 3.0)) or 30%

That seems way too high so I must have some of the assumptions wrong or some of the calculations wrong or both.

Any help from an actual RT (or someone who actually knows about these things) would be greatly appreciated.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
TahoeGal
Posts: 104
Joined: Tue Aug 27, 2013 3:07 pm

Re: Effective O2 percentage with added O2??

Post by TahoeGal » Fri May 25, 2018 3:24 pm

Great question. You’ve piqued my interest! I don’t know anything about your calculations, and look forward to any insight others give.

_________________
Mask: Simplus Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Machine: Resmed S9 VPAP Adapt. 25 max pressure/15 max EPAP/ 6 min EPAP/ 15 max PS /0 min PS. Inogen At Home, 4 lpm

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: Effective O2 percentage with added O2??

Post by HoseCrusher » Fri May 25, 2018 6:31 pm


_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.
SpO2 96+% and holding...

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Effective O2 percentage with added O2??

Post by Goofproof » Fri May 25, 2018 6:37 pm

Only the part of the O2 that is used up in the lungs, is the effective part, the rest is like the gas tank in your auto storage and grade of fuel. It must be used to be effective. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
TASmart
Posts: 1071
Joined: Wed Jan 04, 2017 7:23 pm
Location: Eugene, OR

Re: Effective O2 percentage with added O2??

Post by TASmart » Fri May 25, 2018 6:43 pm

The math is way way more complicated than you think. First, are you using an oxygen concentrator, which has an output way lower than 100% O2, or a pure O2 tank? The net flow into a room is the flow in less the flow out, and you have no way to estimate the flow out, through registers doorway, leaks, etc. The total volume of air in the room will have an effect on the growth rate of the O2 concentration. Then, taking into account those factors it is a differential equation to determine the growth rate of the O2 concentration, and the steady-state value. And don't forget that you are removing some of the O2, so that your exhaled O2 concentration must be taken into account. The way to set this up is a compartmental model, with all the inputs, and all of outflows, then a dynamic balance to arrive at the steady state concentration.
All posts reflect my own opinion based on my experience and reading.
Your mileage may vary
Past performance is no guarantee of future results
Consult with your own physician as people very

D.H.
Posts: 3535
Joined: Sun Sep 04, 2016 7:07 pm

Re: Effective O2 percentage with added O2??

Post by D.H. » Fri May 25, 2018 7:10 pm

What is the reason that you're using an oxygen concentrator? If you have plain Sleep Apnea and no other conditions, you shouldn't need it.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Effective O2 percentage with added O2??

Post by Goofproof » Fri May 25, 2018 8:15 pm

TASmart wrote:
Fri May 25, 2018 6:43 pm
The math is way way more complicated than you think. First, are you using an oxygen concentrator, which has an output way lower than 100% O2, or a pure O2 tank? The net flow into a room is the flow in less the flow out, and you have no way to estimate the flow out, through registers doorway, leaks, etc. The total volume of air in the room will have an effect on the growth rate of the O2 concentration. Then, taking into account those factors it is a differential equation to determine the growth rate of the O2 concentration, and the steady-state value. And don't forget that you are removing some of the O2, so that your exhaled O2 concentration must be taken into account. The way to set this up is a compartmental model, with all the inputs, and all of outflows, then a dynamic balance to arrive at the steady state concentration.
Output should be 90% or the machine is broken, 90% is NOT way less than 100%, at least it wasn't taught that way in my math class. Jim

My O2 unit is in the next room, hose running under the door, so my room builds up O2 levels also. It hasn't started a fire yet!
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Effective O2 percentage with added O2??

Post by palerider » Fri May 25, 2018 8:18 pm

JDS74 wrote:
Fri May 25, 2018 8:53 am
4) Supplemental oxygen is 100% oxygen or close enough to not matter
Actually about 95, if you're lucky (assuming it's coming from a concentrator, since what it actually does is filter out the nitrogen).
JDS74 wrote:
Fri May 25, 2018 8:53 am
5
Here are my calculations:
Total room air flow = Minute Vent + Unintentional Leak Rate or 12.2 + 9.9 lpm or 22.1 lpm (for me last night, data from Encore)
Total room O2 flow = O2% * Total Flow or 22.1 * 21% or 4.7 lpm
Supplemental O2 flow = 3 lpm (measured on my O2 concentrator)

Final O2 % for combined air flow = (total room O2 flow + Supplemental O2 flow)/(Total room air flow + Supplemtal O2 flow)
or (4.7 + 3.0)/(22.1 + 3.0)) or 30%

That seems way too high so I must have some of the assumptions wrong or some of the calculations wrong or both.

Any help from an actual RT (or someone who actually knows about these things) would be greatly appreciated.
Actually, all you need to do is figure the *total leak* on your machine, which fortunately, yours shows, and add the 3lpm O2.

So if you're venting 70lpm, then you've got 70lpm at around 21%, plus 3lpm at around 95%.

your MV doesn't factor in, because the O2 is being added upstream, to the total airflow.

At least, that's how I figure it... I'll leave it up to you to do the maths ;)

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Effective O2 percentage with added O2??

Post by palerider » Fri May 25, 2018 8:24 pm

. wrote:
Fri May 25, 2018 6:43 pm
The math is way way more complicated than you think. First, are you using an oxygen concentrator, which has an output way lower than 100% O2, or a pure O2 tank?[
Not if the concentrator is working properly... and many these days have O2 sensors that alarm, (the one on the concentrator here alarms if the O2 percentage is below 85%, normal operation is 85%-95.6%
. wrote:
Fri May 25, 2018 6:43 pm
The net flow into a room is the flow in less the flow out, and you have no way to estimate the flow out, through registers doorway, leaks, etc. The total volume of air in the room will have an effect on the growth rate of the O2 concentration. Then, taking into account those factors it is a differential equation to determine the growth rate of the O2 concentration, and the steady-state value. And don't forget that you are removing some of the O2, so that your exhaled O2 concentration must be taken into account. The way to set this up is a compartmental model, with all the inputs, and all of outflows, then a dynamic balance to arrive at the steady state concentration.
Perhaps if OP were to hermetically seal himself into a room, you might have a point...