Something's gone wrong.

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Emmet
Posts: 16
Joined: Sun Jan 21, 2018 11:45 am
Location: Near Smackass Gap, North Carolina

Something's gone wrong.

Post by Emmet » Tue May 08, 2018 10:51 pm

I’ve been using CPAP since January, averaging 8.5 hours per night with no issues; quickly becoming so inured to the thing that at times I’ve forgotten that I have it on. I started feeling better almost immediately, and have gotten progressively better over time; my AHI's finally decreased below 5, and I look forward to sleeping with it.
In the past few days I’ve been awakening with air hunger and deep labored respiration, compelling me to breathe through my mouth, panting until I’ve caught my breath. The past couple of nights I’ve been able to resume therapy once I’ve relaxed and my breathing has returned to normal, but tonight I awoke after only an hour and a half, and just can’t breathe through the thing; after a few minutes my respirations become progressively labored, long and deep, as if I’m trying to breathe through a coffee stirrer, and I feel like I’m suffocating; yet if I pull off the mask, the thing's blowing a gale.
Everything’s clean, new filter 9 days ago, 10 cm H2O as always; nothing appears to be wrong with the set.
Suggestions would be greatly appreciated.

KLAussie
Posts: 9
Joined: Wed Feb 07, 2018 11:46 pm

Re: Something's gone wrong.

Post by KLAussie » Wed May 09, 2018 3:37 am

Emmet,
Have you made sure that the vents on the P10 are clear? You should be able to feel air coming out of the back of the mask (on either side of the hose connection) if you put your hand up near it when in use.
The vents are there to allow carbon dioxide out of the mask (exhaled from our breathing). If they were blocked it may cause what you are experiencing (just my non-medically educated guess!)

The vents on the P10 are a very fine mesh, so it may not take a lot to block up if not cleaned regularly.
I tackle mine with a toothbrush weekly.

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: Something's gone wrong.

Post by prodigyplace » Wed May 09, 2018 4:28 am

Cleaning the vents regularly (weekly is usually fine) as well as washing & soaking the headgear on the P10 is a must. The headgear stretches and soaking helps it contract to seal better. I find sometimes replacing the nasal pillow with a new one helps leaks there too.

The vents definitely tend to get plugged up but the small mesh is what makes it a quiet mask.

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.

User avatar
Emmet
Posts: 16
Joined: Sun Jan 21, 2018 11:45 am
Location: Near Smackass Gap, North Carolina

Re: Something's gone wrong.

Post by Emmet » Wed May 09, 2018 5:04 am

Thanks; I use a heated hose and hose cover, and just soaked the mask in vinegar and scrubbed the exhaust ports with a toothbrush yesterday.

KLAussie
Posts: 9
Joined: Wed Feb 07, 2018 11:46 pm

Re: Something's gone wrong.

Post by KLAussie » Wed May 09, 2018 5:49 am

Maybe try the next larger pillow size?
I found that although the medium fit my nose it left me feeling air starved.
I use the large now and have no issues.

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: Something's gone wrong.

Post by prodigyplace » Wed May 09, 2018 5:54 am

soap & water and a water soak is recommended for the elastic headgear IIRC. A am not sure if vinegar would damage the removable elastic headgear.

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.

User avatar
RicaLynn
Posts: 526
Joined: Fri Aug 15, 2014 4:23 pm
Location: Western MT

Re: Something's gone wrong.

Post by RicaLynn » Wed May 09, 2018 4:36 pm

prodigyplace wrote:
Wed May 09, 2018 5:54 am
soap & water and a water soak is recommended for the elastic headgear IIRC. A am not sure if vinegar would damage the removable elastic headgear.
No more than it would damage any other fabric that was soaked in vinegar. Hot water wash for headgear to shrink it is recommended, I know plenty of folks who simply pace it in a lingerie bag and toss it in with their regular laundry, just hang it to dry.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Backup/travel unit is an identical S9 AutoSet for Her w/Eson nasal mask

Mike A
Posts: 9
Joined: Thu Apr 12, 2018 8:36 am

Re: Something's gone wrong.

Post by Mike A » Wed May 09, 2018 4:44 pm

Maybe try another mask if you have one

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: Something's gone wrong.

Post by prodigyplace » Wed May 09, 2018 5:21 pm

RicaLynn wrote:
Wed May 09, 2018 4:36 pm
prodigyplace wrote:
Wed May 09, 2018 5:54 am
soap & water and a water soak is recommended for the elastic headgear IIRC. A am not sure if vinegar would damage the removable elastic headgear.
No more than it would damage any other fabric that was soaked in vinegar. Hot water wash for headgear to shrink it is recommended, I know plenty of folks who simply pace it in a lingerie bag and toss it in with their regular laundry, just hang it to dry.
I'm not a lingerie kind of guy ;)

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.

User avatar
zonker
Posts: 11485
Joined: Fri Jun 19, 2015 4:36 pm

Re: Something's gone wrong.

Post by zonker » Wed May 09, 2018 5:30 pm

prodigyplace wrote:
Wed May 09, 2018 5:21 pm
RicaLynn wrote:
Wed May 09, 2018 4:36 pm
prodigyplace wrote:
Wed May 09, 2018 5:54 am
soap & water and a water soak is recommended for the elastic headgear IIRC. A am not sure if vinegar would damage the removable elastic headgear.
No more than it would damage any other fabric that was soaked in vinegar. Hot water wash for headgear to shrink it is recommended, I know plenty of folks who simply pace it in a lingerie bag and toss it in with their regular laundry, just hang it to dry.
I'm not a lingerie kind of guy ;)
we won't hold that against you!
:lol:
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Something's gone wrong.

Post by palerider » Wed May 09, 2018 7:56 pm

Emmet wrote:
Wed May 09, 2018 5:04 am
Thanks; I use a heated hose and hose cover, and just soaked the mask in vinegar and scrubbed the exhaust ports with a toothbrush yesterday.
Forget the vinegar, use dish soap...(not hand soap)

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Emmet
Posts: 16
Joined: Sun Jan 21, 2018 11:45 am
Location: Near Smackass Gap, North Carolina

Re: Something's gone wrong.

Post by Emmet » Wed May 16, 2018 6:07 am

Thanks for all of the suggestions. Although invisible (at least with bifocals), some sort of accretion apparently had gradually clogged the diffusers of the exhaust vents. A good scrub with a toothbrush and Ivory Liquid followed by a thorough hot water rinse did the trick, and is now part of my weekly maintenance schedule; good as new. Thanks all.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Something's gone wrong.

Post by palerider » Wed May 16, 2018 6:12 am

Emmet wrote:
Wed May 16, 2018 6:07 am
Thanks for all of the suggestions. Although invisible (at least with bifocals), some sort of accretion apparently had gradually clogged the diffusers of the exhaust vents. A good scrub with a toothbrush and Ivory Liquid followed by a thorough hot water rinse did the trick, and is now part of my weekly maintenance schedule; good as new. Thanks all.
Excellent!

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: Something's gone wrong.

Post by prodigyplace » Wed May 16, 2018 6:42 am

Emmet wrote:
Wed May 16, 2018 6:07 am
Thanks for all of the suggestions. Although invisible (at least with bifocals), some sort of accretion apparently had gradually clogged the diffusers of the exhaust vents. A good scrub with a toothbrush and Ivory Liquid followed by a thorough hot water rinse did the trick, and is now part of my weekly maintenance schedule; good as new. Thanks all.
That is basically what I (try) to do along with cleaning/soaking the headgear but sometimes it is 2 weeks.
I have 2 frames & headgear and alternate between them so the headgear has time to fully dry out. I just transfer the nasal pillow.

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.