Tor wrote:Thanks.
But I don't understand why the minimum pressure matters so much. I mean, isn't the minimum pressure just the startup pressure, and when my APAP figure out it needs more pressure its just rise to the level i need?
Is it something I have missed here?
Yep......but if your starting pressure is TOO low, it's got too far to go to reach the "optimal" starting pressure. If your pressure is too low when you go to sleep, it's like not having any pressure at all and you're going to get lots of events......UNTIL the pressure gets higher.
Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
-
NightHawkeye
- Posts: 2431
- Joined: Thu Dec 29, 2005 11:55 am
- Location: Iowa - The Hawkeye State
Post
by NightHawkeye » Mon Feb 19, 2007 10:44 am
Tor wrote:But I don't understand why the minimum pressure matters so much. I mean, isn't the minimum pressure just the startup pressure, and when my APAP figure out it needs more pressure its just rise to the level i need?
The pressure rise sometimes only occurs after the machine senses an apnea, but by then it's too late to stop that particular apnea from occurring. The machine's pressure rise tends to stop subsequent apneas, but can't really do anything to fix the first few apneas until the pressure rises to a therapeutic value.
Raising the minimum pressure simply minimizes the number of these residual apneas. Hence, people find that their AHI is a little lower when the minimum pressure is close to titration pressure.