High flow oxygen therapy

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Oltremare
Posts: 187
Joined: Tue Dec 29, 2015 7:01 pm
Location: On the other side of the sea - Italy

High flow oxygen therapy

Post by Oltremare » Tue Apr 10, 2018 9:39 am

My mother-in-law is seriously ill (COPD) and she has a high flow oxygen therapy (10-12 l / m) with humification at 37 degrees Celsius.
Use an AIRVO2 from Fisher & Paykel.
I am worried because the doctor said to use physiological solution for the humidifier rather than distilled water.
Your advice will be very welcome.

Thank you
Oltremare

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Quattro FX FF mask > CMS50D+ pulse oximeter > Climate line > Settings: auto-CPAP pressure Min 11.00 Max 18.0 EPR 3
I hope you will forgive my language errors. To write in English I use a translator online. Thank you

rcobourn
Posts: 207
Joined: Mon Apr 03, 2017 8:25 am

Re: High flow oxygen therapy

Post by rcobourn » Tue Apr 10, 2018 10:15 am

Oltremare wrote:My mother-in-law is seriously ill (COPD) and she has a high flow oxygen therapy (10-12 l / m) with humification at 37 degrees Celsius.
Use an AIRVO2 from Fisher & Paykel.
I am worried because the doctor said to use physiological solution for the humidifier rather than distilled water.
Your advice will be very welcome.

Thank you
Oltremare
It's a mild salt water solution. It won't harm the chamber, and even if it did, you can replace it. I'd say follow doctor's orders here.

User avatar
Oltremare
Posts: 187
Joined: Tue Dec 29, 2015 7:01 pm
Location: On the other side of the sea - Italy

Re: High flow oxygen therapy

Post by Oltremare » Tue Apr 10, 2018 1:12 pm

Thanks rcobourn.
The situation is as follows:
in hospital for 15 days they used water for injectable solutions. When the patient was discharged from the hospital, the doctor said to use the same solution, but it isn't found on the market like that, the doctor said that the physiological solution is fine too.
I am worried about the health of my elderly mother-in-law, she is a very sweet, delicate and fragile woman. I hope the therapy is good too.

I hope my English is acceptable ...

Thank you
Oltremare

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Quattro FX FF mask > CMS50D+ pulse oximeter > Climate line > Settings: auto-CPAP pressure Min 11.00 Max 18.0 EPR 3
I hope you will forgive my language errors. To write in English I use a translator online. Thank you

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: High flow oxygen therapy

Post by palerider » Tue Apr 10, 2018 1:37 pm

Oltremare wrote:
Tue Apr 10, 2018 1:12 pm
Thanks rcobourn.
The situation is as follows:
in hospital for 15 days they used water for injectable solutions. When the patient was discharged from the hospital, the doctor said to use the same solution, but it isn't found on the market like that, the doctor said that the physiological solution is fine too.
I am worried about the health of my elderly mother-in-law, she is a very sweet, delicate and fragile woman. I hope the therapy is good too.

I hope my English is acceptable ...

Thank you
Oltremare
Saline for injection is 0.9% salt, by weight, in water. You could make a good substitute with kosher salt and distilled water.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Julie
Posts: 20056
Joined: Tue Feb 28, 2006 12:58 pm

Re: High flow oxygen therapy

Post by Julie » Tue Apr 10, 2018 3:09 pm

The reason for distilled water is to help keep the humidifier clean... not the patient. What the doctor said to use cannot affect your mother in law in any meaningful way.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: High flow oxygen therapy

Post by palerider » Tue Apr 10, 2018 4:27 pm

Julie wrote:
Tue Apr 10, 2018 3:09 pm
The reason for distilled water is to help keep the humidifier clean... not the patient. What the doctor said to use cannot affect your mother in law in any meaningful way.
OP is talking about an oxygen humidifier, not a cpap humidifier,

Oxygen humidifiers don't operate by evaporation, they bubble the O2 through the water, and that can, and will, pick up stuff from the water, I imagine that's why they're suggesting saline. (but, that's just a guess)

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Julie
Posts: 20056
Joined: Tue Feb 28, 2006 12:58 pm

Re: High flow oxygen therapy

Post by Julie » Tue Apr 10, 2018 5:00 pm

Didn't realize there was such a thing... not fun.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: High flow oxygen therapy

Post by palerider » Tue Apr 10, 2018 6:07 pm

Julie wrote:
Tue Apr 10, 2018 5:00 pm
Didn't realize there was such a thing... not fun.
They're common with oxygen delivery... here's one strapped on the front of a concentrator:
Image

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
klm49
Posts: 442
Joined: Sat May 28, 2016 4:24 pm
Location: Texas

Re: High flow oxygen therapy

Post by klm49 » Tue Apr 10, 2018 6:17 pm

Oltremare wrote:
Tue Apr 10, 2018 9:39 am
My mother-in-law is seriously ill (COPD) and she has a high flow oxygen therapy (10-12 l / m) with humification at 37 degrees Celsius.
Use an AIRVO2 from Fisher & Paykel.
I am worried because the doctor said to use physiological solution for the humidifier rather than distilled water.
Your advice will be very welcome.

Thank you
Oltremare
Personally I would ask the question on a more appropriate board for people with COPD such as:
copdforum.portalone.us

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Amara View Full Face CPAP Mask with Headgear
Additional Comments: Setting: PS 4.0 over 14.0-25.0; Humidifer 4 & Heated Hose
Resmed AirCurve10 Vauto w/humidifier
Amara View mask
O2 - 2-4 lpm

salinas212
Posts: 123
Joined: Sat Jan 13, 2018 10:22 am
Location: Denver, CO

Re: High flow oxygen therapy

Post by salinas212 » Tue Apr 10, 2018 6:47 pm

As a COPDer/CPAPer who uses the exact machine pictured above I'll go out on a limb and guess that saline is being recommend for an acutely ill patient for a myriad of reasons. First, saline has been used for eons as an antibacterial for wound cleansing and treatment. Secondly its cheaper than sterile, deionized or distilled water and can be easily DIY with tap or even well water. Given the bacteria, fungi and bacillus that can be cultured and thrive in moist environs it makes sense that the humidifier container an ideal candidate for a little salinity given its daily use ad infinitum without sterilization. None of this should be interpreted as fact but just my personal gut level instinct.

10-12 L/min is a T-O-N of O2! my machine and the one pictured above is only rated for 5L/min and I doubt its very efficient at much above 4L/min. Additionally a machine like above costs between $600/$700 new. A 10-12L/min machine runs nearer $4k

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: ClimateLineAir Oxy Heated Tube & Perfecto2 O2 Concentrator during sleep only at 2.0 L/min w/bubbler
If you're gonna be a dog, why intentionally choose Chihuahua!?!

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: High flow oxygen therapy

Post by palerider » Tue Apr 10, 2018 6:54 pm

Oltremare wrote:
Tue Apr 10, 2018 1:12 pm
I hope my English is acceptable ...
Your english is excellent.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: High flow oxygen therapy

Post by raisedfist » Tue Apr 10, 2018 7:05 pm

Do you have a reddit account? I would ask on reddit.com/r/respiratorytherapy. A lot of knowledgeable people on there when it comes to COPD - I know some of the RT's there work for the dme companies that visit people in homes.

_________________
Mask
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

User avatar
Oltremare
Posts: 187
Joined: Tue Dec 29, 2015 7:01 pm
Location: On the other side of the sea - Italy

Re: High flow oxygen therapy

Post by Oltremare » Wed Apr 11, 2018 6:32 pm

I thank all of you for your comments.
The situation is very difficult and gets worse as the days go by.
Now, physiological solution is no longer a big problem. My mother-in-law is very tired and depressed.
She lives attached to the oxygen tank for many years.
In 2017, she ended up in the emergency room eleven times and she can't take it anymore and she's letting go. She's thinking about suicide ...
Sometimes life is really hard and me, I hope everything will soon pass.
Thanks again for your precious help and excuse me for my outburst ...

Oltremare

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Quattro FX FF mask > CMS50D+ pulse oximeter > Climate line > Settings: auto-CPAP pressure Min 11.00 Max 18.0 EPR 3
I hope you will forgive my language errors. To write in English I use a translator online. Thank you

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: High flow oxygen therapy

Post by prodigyplace » Wed Apr 11, 2018 6:38 pm

We all need to express our frustrations sometimes. We can be good sympathetic listeners for you too.

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.

User avatar
Pugsy
Posts: 65093
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: High flow oxygen therapy

Post by Pugsy » Wed Apr 11, 2018 7:17 pm

Sending you 2 big cyber hugs...one for you and one for you to take to your mother in law from me.