CPAP to APAP

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Guest

Post by Guest » Mon Jan 08, 2007 5:28 pm

Do you mean "original" as in "not the M-Series"?
Yes, that's what he means... and he's correct. I just read the manual and found that fact in writing.

AltPwrPuffer
Posts: 40
Joined: Sun Sep 17, 2006 1:42 pm
Location: Sallisaw, Oklahoma

Post by AltPwrPuffer » Mon Jan 08, 2007 5:35 pm

If you suffer more from hypopneas like I do, I would set the lower at your titrated pressure of 13 and the upper pressure at like 16. Thought is that the hypopnea is low flow that needs to be maintained and anything else would need the higher pressure.

If that's not the case then I would opt for the lower of 10 to 15 for the comfort of pressure relief. Just a different approach, have fun. Charlie


snorestats
Posts: 46
Joined: Tue Dec 19, 2006 8:48 pm

Post by snorestats » Tue Jan 09, 2007 9:21 pm

Well, my first night going to APAP (or I should say AFLEX), was not too bad.

As mentioned, I went 10-15. The machine started up at 10, but it seemed like it was too low of pressure... almost not comfortable.

I also noticed in my graphs (for the first time) I showed some FLI (Flow Limitations). What exactly is that? Could it be because I was using pressure below what I usually use? I had ramp of 5 for 20 minutes before changing to apap, and I never had any values for Flow limitations before, but now I showed 5 for yesterday. My AHI was 1.2 (best value so far).

I'll give this a try for an entire week then re-evaluate.


_________________
Mask

User avatar
Wulfman
Posts: 12316
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Post by Wulfman » Tue Jan 09, 2007 9:30 pm

A Flow Limitation is a "wannabe hypopnea"......but didn't quite make it.
They only show up in the "Auto" mode reports. I wouldn't worry about them.

Glad to hear you had a good night at those settings.

Den

(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
jrfoster
Posts: 408
Joined: Sat Jun 10, 2006 1:48 pm
Location: Tennessee

Post by jrfoster » Tue Jan 09, 2007 9:36 pm

Let me ask a question since we are discussing going from CPAP to APAP.
When I had my sleep study with mask, they discovered that at a pressure of 10cm my apneas ceased, so they prescribed 10 as my setting. I had a ramp feature that went from 4-10 in 45 minutes with CPAP.

I have since at my doctors recommendation gone to an APAP, and have set the parameters between 5-10cm. Most of the night I am at 9.8 or 10.
However, the last time I ran the software for a 15 night stretch, I discovered that I was having an average AHI of 6.9, ( avg for the 15 days) and anywhere from 2.5 up to 6 Apnea events when at a pressure of 10!

I was tested using a nasal mask, but have switched to a full face mask, would that make any difference?

What is causing apneas to occur at 10cm of pressure when during the sleep study there were 0?

I have 890 hours on XPAP right now and am still not sleeping like I think I should.

Jeff


_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Pressure of 12cm Respironics Battery Pack

User avatar
Wulfman
Posts: 12316
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Post by Wulfman » Tue Jan 09, 2007 9:56 pm

Jeff,

Without having any reports to look at, I would suppose that your events are probably happening at the lower pressures, including the ramp time.
Also, the events may be happening before your pressure rises enough to stop them.
OR, you may have gotten used to the pressure of 10 and need to set your top pressure up a bit.
OR, a combination of various aspects of the above.

Since I've never used a nasal mask, I can't comment on any pressure differences on the different types of masks.

Den

(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05