Dreamstation clinician changes not saving

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
grumpycamel
Posts: 2
Joined: Tue Feb 20, 2018 7:23 am

Dreamstation clinician changes not saving

Post by grumpycamel » Tue Feb 20, 2018 8:21 am

Hi all,
Im a relatively new CPAP user and am starting to feel the benefits of CPAP therapy. I have a Philips Respironics Dreamstation CPAP Pro (fixed pressure) with a humidifier and have been wanting to change some of the settings on the machine.
I have read the clinicians guide obtained from the Apnea Board as previously posted on this forum and have followed the instructions for changing the settings as directed in the manual.
The first time I use the CPAP machine after the changes, the CPAP machine has all the changed settings that I entered. If I stop the CPAP machine and restart it, the settings revert back to the values before the changes.
So currently, each night, I'm having to go into the clinicians menu and entering the changes which is a huge pain in the neck.
Can anyone help me and point out what I'm doing wrong please?
Thank you.

TedVPAP
Posts: 975
Joined: Sat Jun 12, 2010 10:29 am

Re: Dreamstation clinician changes not saving

Post by TedVPAP » Tue Feb 20, 2018 8:41 am

Since you are new to CPAP, perhaps your machine is in contact with the mother ship and they are resetting your treatment parameters.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/

User avatar
greatunclebill
Posts: 1503
Joined: Mon Feb 20, 2012 7:48 pm
Location: L.A. (lower alabama)

Re: Dreamstation clinician changes not saving

Post by greatunclebill » Tue Feb 20, 2018 9:10 am

the settings are probably embedded in the sd card. if you don't care what they think, use a new sd card.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: myAir, OSCAR. cms-50D+. airsense 10 auto & (2009) remstar plus m series backups
First diagnosed 1990
please don't ask me to try nasal. i'm a full face person.
the avatar is Rocco, my Lhasa Apso. Number one "Bama fan. 18 championships and counting.
Life member VFW Post 4328 Alabama
MSgt USAF (E-7) medic Retired 1968-1990

D.H.
Posts: 3533
Joined: Sun Sep 04, 2016 7:07 pm

Re: Dreamstation clinician changes not saving

Post by D.H. » Tue Feb 20, 2018 4:25 pm

greatunclebill wrote:
Tue Feb 20, 2018 9:10 am
the settings are probably embedded in the sd card. if you don't care what they think, use a new sd card.
Per Respironics, the machine stores up to six months of data internally, and that would transfer to a new card. I suggest taking out the card tonight before you make the changes, leave it out all day, and then see if it changes back anyhow.

If than doesn't work, remove the modem the next night, leave it out, and see if that fixes the problem. You will find the modem in a compartment similar to the card/filter compartment on the opposite side of the machine.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

User avatar
Pugsy
Posts: 65207
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Dreamstation clinician changes not saving

Post by Pugsy » Tue Feb 20, 2018 5:02 pm

Try a new/blank SD card if you have one.
If you don't have one erase this one and then make the changes to your machine and put the clean card back in the machine and see if the settings hold or not.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
greatunclebill
Posts: 1503
Joined: Mon Feb 20, 2012 7:48 pm
Location: L.A. (lower alabama)

Re: Dreamstation clinician changes not saving

Post by greatunclebill » Wed Feb 21, 2018 9:50 pm

Pugsy wrote:
Tue Feb 20, 2018 5:02 pm
Try a new/blank SD card if you have one.
If you don't have one erase this one and then make the changes to your machine and put the clean card back in the machine and see if the settings hold or not.
my guess is the setting are in a hidden system file that wouldn't show up and the average person wouldn't think to look for them to delete them. easiest to use a new card or format the present card.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: myAir, OSCAR. cms-50D+. airsense 10 auto & (2009) remstar plus m series backups
First diagnosed 1990
please don't ask me to try nasal. i'm a full face person.
the avatar is Rocco, my Lhasa Apso. Number one "Bama fan. 18 championships and counting.
Life member VFW Post 4328 Alabama
MSgt USAF (E-7) medic Retired 1968-1990

D.H.
Posts: 3533
Joined: Sun Sep 04, 2016 7:07 pm

Re: Dreamstation clinician changes not saving

Post by D.H. » Thu Feb 22, 2018 9:05 am

greatunclebill wrote:
Wed Feb 21, 2018 9:50 pm
Pugsy wrote:
Tue Feb 20, 2018 5:02 pm
Try a new/blank SD card if you have one.
If you don't have one erase this one and then make the changes to your machine and put the clean card back in the machine and see if the settings hold or not.
my guess is the setting are in a hidden system file that wouldn't show up and the average person wouldn't think to look for them to delete them. Easiest to use a new card or format the present card.
Since six months worth of data are stored in the machine internally (according to Respironics), it's possible that even this file (if it exists) is stored. However, my strongest suspicion is that your provider is resetting it remotely.

I would ask the following:
(1) Does it reset if you take a potty break and turn the machine off?
(2) Is your preferred pressure still active in the AM? If so, turn the machine off in the AM for about 60 seconds, then turn it back on and see if the reset takes place at that point.
(3) On a day when you're home, do periodic checks and see at exactly what time the reset occurs.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

grumpycamel
Posts: 2
Joined: Tue Feb 20, 2018 7:23 am

Re: Dreamstation clinician changes not saving

Post by grumpycamel » Thu Feb 22, 2018 5:30 pm

Thank you all for your replies and suggestion.
I decided to take the card out and make some changes to the clinicians menu. Lo and behold, the changes stuck. So there must be a file in the SD card which override any changes someone makes manually.
I found the offending prescription file on the card, removed it and used Encore Pro software to create a new prescription file on the card. All previous sleep data was retained.
Seems to work fine so far.

User avatar
klm49
Posts: 442
Joined: Sat May 28, 2016 4:24 pm
Location: Texas

Re: Dreamstation clinician changes not saving

Post by klm49 » Thu Feb 22, 2018 6:41 pm

grumpycamel wrote:
Thu Feb 22, 2018 5:30 pm
Thank you all for your replies and suggestion.
I decided to take the card out and make some changes to the clinicians menu. Lo and behold, the changes stuck. So there must be a file in the SD card which override any changes someone makes manually.
I found the offending prescription file on the card, removed it and used Encore Pro software to create a new prescription file on the card. All previous sleep data was retained.
Seems to work fine so far.
Interesting! I have a Dream Station and have never had this happen. Maybe your DME caused the file to be there initially. I don't think my DME is smart enough ro do that, last time Dr. changed my pressures they couldn't even get the settings right and I had to change it myself. I guess I need to look at Encore Pro and find that setting in there.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Amara View Full Face CPAP Mask with Headgear
Additional Comments: Setting: PS 4.0 over 14.0-25.0; Humidifer 4 & Heated Hose
Resmed AirCurve10 Vauto w/humidifier
Amara View mask
O2 - 2-4 lpm

User avatar
Pugsy
Posts: 65207
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Dreamstation clinician changes not saving

Post by Pugsy » Thu Feb 22, 2018 6:59 pm

klm49 wrote:
Thu Feb 22, 2018 6:41 pm
Interesting! I have a Dream Station and have never had this happen. Maybe your DME caused the file to be there initially. I don't think my DME is smart enough ro do that, last time Dr. changed my pressures they couldn't even get the settings right and I had to change it myself. I guess I need to look at Encore Pro and find that setting in there.
It's done with the Encore software but not all DMEs or doctors or whomever supplies the equipment will bother using Encore and the SD card for settings changing. It's a bit of extra work to use the software to change things especially now that we have so much stuff that can be done over the cellular system. So you might be correct about your DME not being smart enough. I can do it but it's a PITA to use the software. I can do it manually in the time it takes Encore to load for me. :lol:
It used to be a lot more common than it is now so it doesn't surprise me the your DME didn't do it.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
klm49
Posts: 442
Joined: Sat May 28, 2016 4:24 pm
Location: Texas

Re: Dreamstation clinician changes not saving

Post by klm49 » Thu Feb 22, 2018 7:04 pm

Pugsy wrote:
Thu Feb 22, 2018 6:59 pm
klm49 wrote:
Thu Feb 22, 2018 6:41 pm
Interesting! I have a Dream Station and have never had this happen. Maybe your DME caused the file to be there initially. I don't think my DME is smart enough ro do that, last time Dr. changed my pressures they couldn't even get the settings right and I had to change it myself. I guess I need to look at Encore Pro and find that setting in there.
It's done with the Encore software but not all DMEs or doctors or whomever supplies the equipment will bother using Encore and the SD card for settings changing. It's a bit of extra work to use the software to change things especially now that we have so much stuff that can be done over the cellular system. So you might be correct about your DME not being smart enough. I can do it but it's a PITA to use the software. I can do it manually in the time it takes Encore to load for me. :lol:
It used to be a lot more common than it is now so it doesn't surprise me the your DME didn't do it.
Last time I was at the Dr. I told her the DME wasn't so hot and would she recommend someone else? She told me "Their aren't any good ones".......

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Amara View Full Face CPAP Mask with Headgear
Additional Comments: Setting: PS 4.0 over 14.0-25.0; Humidifer 4 & Heated Hose
Resmed AirCurve10 Vauto w/humidifier
Amara View mask
O2 - 2-4 lpm

User avatar
greatunclebill
Posts: 1503
Joined: Mon Feb 20, 2012 7:48 pm
Location: L.A. (lower alabama)

Re: Dreamstation clinician changes not saving

Post by greatunclebill » Tue Feb 27, 2018 8:56 pm

klm49 wrote:
Thu Feb 22, 2018 7:04 pm
Pugsy wrote:
Thu Feb 22, 2018 6:59 pm
klm49 wrote:
Thu Feb 22, 2018 6:41 pm
Last time I was at the Dr. I told her the DME wasn't so hot and would she recommend someone else? She told me "Their aren't any good ones".......
True that.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: myAir, OSCAR. cms-50D+. airsense 10 auto & (2009) remstar plus m series backups
First diagnosed 1990
please don't ask me to try nasal. i'm a full face person.
the avatar is Rocco, my Lhasa Apso. Number one "Bama fan. 18 championships and counting.
Life member VFW Post 4328 Alabama
MSgt USAF (E-7) medic Retired 1968-1990