Air pressure falls below minimum

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
languageusa
Posts: 13
Joined: Thu Oct 25, 2012 5:10 pm
Location: Naples, Florida, USA

Air pressure falls below minimum

Post by languageusa » Sat Jan 13, 2018 10:26 am

Just started to use an APAP machine -- ResMed AirSense 10 -- after 5 years of being on a CPAP machine. Noticed that the air pressure falls (rather consistently throughout the night) below the preset minimum of 10 cm occasionally showing 6, 7 up to 9... cm. In all other respects, the machine seems to work fine. To me this is abnormal: the minimum is expected to be just that -- the minimum. I would greatly appreciate your thoughts on that.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: PR System One REMstar Pro CPAP Machine with C-Flex Plus

User avatar
Pugsy
Posts: 65377
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Air pressure falls below minimum

Post by Pugsy » Sat Jan 13, 2018 10:33 am

Are you using exhale relief (EPR) and if so at what setting?
What mode of operation are you using...exactly what are ALL the settings?
Do you know how to go into the clinical setup menu are to verify all the settings?
https://sleep.tnet.com/home/files/resme ... -guide.pdf
What software are you using that is showing you the drop in the minimum?
Can you share an image of what you are seeing so that we can see what you are seeing?
https://sleep.tnet.com/reference/tips/imgur

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
LSAT
Posts: 13408
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Air pressure falls below minimum

Post by LSAT » Sat Jan 13, 2018 10:39 am

If you have EPE set at 3, the pressure is reduced by that amount during exhale. If your setting for Minimum is at 10 and you are inhaling at 10, with a EPR of 3 you would be exhaling at 7 regardless of your minimum setting.

User avatar
languageusa
Posts: 13
Joined: Thu Oct 25, 2012 5:10 pm
Location: Naples, Florida, USA

Re: Air pressure falls below minimum

Post by languageusa » Sat Jan 13, 2018 10:57 am

Exhale relief (EPR) = 3
What mode of operation = APAP
RAMP = 5 cm for 10 minutes (am going to set it to OFF, just don't need it)
Software used = SleepyHead
Yes, I have the clinical setup menu
Will prepare the image a little later, though the pressure change is detectable when you move the vertical line reflecting the pressure changes along the graph.

Pugsy, I want to use this opportunity to THANK YOU immensely for the invaluable help you offered me over 5 years ago. It was like a lifeline for me at the time of despair -- no sleep, huge AHIs, totally ignorant Dr., anxiety and, finally, depression. All this, no doubt, attributed to my CPAP experience. You don't even know what support it was! I wanted to send you a note before, so I am doing it know. Thank you!..

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: PR System One REMstar Pro CPAP Machine with C-Flex Plus

User avatar
languageusa
Posts: 13
Joined: Thu Oct 25, 2012 5:10 pm
Location: Naples, Florida, USA

Re: Air pressure falls below minimum

Post by languageusa » Sat Jan 13, 2018 11:01 am

Pugsy wrote:Are you using exhale relief (EPR) and if so at what setting?
What mode of operation are you using...exactly what are ALL the settings?
Do you know how to go into the clinical setup menu are to verify all the settings?
https://sleep.tnet.com/home/files/resme ... -guide.pdf
What software are you using that is showing you the drop in the minimum?
Can you share an image of what you are seeing so that we can see what you are seeing?
https://sleep.tnet.com/reference/tips/imgur
Exhale relief (EPR) = 3
What mode of operation = APAP
RAMP = 5 cm for 10 minutes (am going to set it to OFF, just don't need it)
Software used = SleepyHead
Yes, I have the clinical setup menu
Will prepare the image a little later, though the pressure change is detectable when you move the vertical line reflecting the pressure changes along the graph.

Pugsy, I want to use this opportunity to THANK YOU immensely for the invaluable help you offered me over 5 years ago. It was like a lifeline for me at the time of despair -- no sleep, huge AHIs, totally ignorant Dr., anxiety and, finally, depression. All this, no doubt, attributed to my CPAP experience. You don't even know what support it was! I wanted to send you a note before, so I am doing it know. Thank you!..

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: PR System One REMstar Pro CPAP Machine with C-Flex Plus

User avatar
Pugsy
Posts: 65377
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Air pressure falls below minimum

Post by Pugsy » Sat Jan 13, 2018 11:05 am

Using the ramp and EPR is what is causing SleepyHead to record the minimum being so low.
Ignore it or turn ramp off if you don't need it.

With the minimum set to 10 and EPR to 3 then during exhale the pressure will drop to 7 and that will be the "new" minimum pressure per SH's minimum.

Mainly it is the ramp thing making things whacky. When you start at 5 cm then even with EPR at 3 it can only drop to 4 and while it increases up to 10 fairly quickly the ramp numbers are throwing off your minimum averages.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.