Sprockets_Zomie wrote:Matt00926 wrote:Your backup rate of 16bpm is rather high for most people. It could be causing asyncrhony where you're trying to take spontaneous breaths but the machine is constantly cutting in because you're breathing spontaneously below 16bpm.
What is your spontaneous breathing rate on the machine (lying down with your mask on your face, machine turned on) while awake?
I would lower your backup rate to 10 or 12bpm and see if it works out better. Your patient triggered breath percentage is low which means the machine is taking a lot of breaths for you while you sleep.
With the Respironics BiPAP, S/T the set Ti only applies to machine initiated timed breaths, not spontaneous ones. Your set Ti of 1.5s is most likely not the culprit. Only in pressure control (PC) mode can you apply the Ti to all breaths.
The Sleepyhead shot shows you running S/T with AVAPS enabled? So not a straight bi-level pressures...you need to list out all your exact therapy mode and settings for us so we can help you.
You also have a high leak percentage -- you need to take some time and carefully practice fitting your Simplus mask.
I dont even see those settings, I dont even know where you turn AVAPS on and off ill post all the settings I can see:
Mode: S/T
IPAP: 18
EPAP: 8
BPM: 16
Ti: 1.5
Rise Time: 1
Tubing Type: 22
SystemOne Resistance: 0
hPa/cmH20: cmH20
According to sleepyhead you are using a philips respironics system one bipap avaps machine. It also shows varying IPAP and EPAP levels in the summary data on the left. It's possible that sleepyhead is misreading your machine, but i can't see why since both are older models and I'm sure support has been improved by now.
On your actual machine, below the therapy screen, does it say BiPAP S/T? If so - then yeah your machine does not have AVAPS function. The BiPAP AVAPS version has that name above the therapy screen instead of below.
It also says PAP therapy mode is AVAPS on sleepyhead. On some machines it's a stand alone therapy mode, and on some it's just an option you can add to therapy modes to target a specific tidal volume. Ex: you can use S/T with AVAPS enabled or not.
Are you positive that you are using straight pressures of 18/8?
Anyway, I wouldn't take any action regarding the AHI until you get the leaks under control.
A Ti of 1.5s is normal and as I said before, only applies to timed breaths anyway on respironics unless in more advanced therapy modes than s/t. I would definitely lower that backup rate though.
Depending on what type of condition you have, maybe raise the rise time as well - 1 is very fast must feel like a wack in the face. Could be causing mask sealing issues as well, especially with a pressure support of 10.