Resmed S9 Power restarting randomly

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
gtrrich
Posts: 2
Joined: Mon Sep 15, 2014 12:34 pm
Location: USA

Resmed S9 Power restarting randomly

Post by gtrrich » Sat Dec 23, 2017 11:07 am

My cpap is restarting at night at random. Also, when i wiggle the power cord it turns off & back on, but the power pack seems to be ok when I checked with a voltmeter. I just wondered if this was a common problem and what the fix is, before I attempt disassembly to troubleshoot it. Any info appreciated. Thanks.

_________________
MaskHumidifier
Additional Comments: Need software

User avatar
Pugsy
Posts: 65356
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Resmed S9 Power restarting randomly

Post by Pugsy » Sat Dec 23, 2017 11:13 am

Might be a loose connection where the power cord attaches to the machine on the back side.
There are little pins inside that end of the power supply that sometimes get broken or bent and make the connection intermittently.

I would as whomever you got the machine from if they had a spare power supply you can try and see if that fixes the problem or not before trying to take apart the machine itself.

If that takes care of the problem then just buy a new power supply or maybe your DME will sell you one and insurance could maybe pay for it.


https://www.cpap.com/productpage/resmed ... -cord.html

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

fabioprestes1972
Posts: 1
Joined: Mon Dec 25, 2017 11:50 am

Re: Resmed S9 Power restarting randomly

Post by fabioprestes1972 » Mon Dec 25, 2017 12:09 pm

Same problem with me.

Moreover, when I wiggle the power cord (a new power cord BTW), my S9 turns on, but it stops working -- LCD shows messages like "motor fail" or (more recently) "system error".

In such cases, I have to disconnect the power cord, wait for some minutes, connect it again and slap the S9 plastic bottom some times. Only then it starts.