Airsense 10 for her Pressure Relief Setting

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
triciar
Posts: 2
Joined: Tue Dec 19, 2017 12:14 pm

Airsense 10 for her Pressure Relief Setting

Post by triciar » Tue Dec 19, 2017 12:34 pm

Newbie here....I was originally set up with a machine through insurance but returned it once I found out my insurance wouldn't cover it until my deductible was met ($2000). Prior to returning, I took a photo of the settings knowing I may have to set it up myself once I purchased one on my own. Today I purchase a gently used machine and I was duplicating my original settings however I noticed the original machine set up has Pressure Relief set to On under tube temp. The gently used machine I purchased doesn't have that option. I've tried accessing the clinical settings but I'm unable to find that option. Any suggestions? Being new to this, I don't really know if I need to be concerned about that option not being there.

Thank you,
Tricia

User avatar
Pugsy
Posts: 65417
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Airsense 10 for her Pressure Relief Setting

Post by Pugsy » Tue Dec 19, 2017 12:47 pm

Exhale relief is called EPR and it's on the AirSense 10 for Her machine for sure. I have that machine and I use EPR.

You need to go into the clinical menu setup area.
https://sleep.tnet.com/home/files/resme ... -guide.pdf

I thought that there was a setting to turn it so patient can have access to it without having to go into the clinical menu area but I just looked at my AirSense 10 AutoSet for Her and there wasn't.
I had to go into the clinical setup menu area to turn it on and set the desired choice.
I have no idea how or why it showed up under the hose temp on your other machine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
JimW159
Posts: 804
Joined: Fri Jul 08, 2016 10:26 am
Location: Tarpon Springs, FL

Re: Airsense 10 for her Pressure Relief Setting

Post by JimW159 » Tue Dec 19, 2017 4:33 pm

Pugsy wrote:Exhale relief is called EPR and it's on the AirSense 10 for Her machine for sure. I have that machine and I use EPR.

You need to go into the clinical menu setup area.
https://sleep.tnet.com/home/files/resme ... -guide.pdf

I thought that there was a setting to turn it so patient can have access to it without having to go into the clinical menu area but I just looked at my AirSense 10 AutoSet for Her and there wasn't.
I had to go into the clinical setup menu area to turn it on and set the desired choice.
I have no idea how or why it showed up under the hose temp on your other machine.
If I understand what has been written thus far, I think that "Essentials Plus" must have been selected through the clinical menu in order for users to be able to control the EPR function from the "My Options" screen.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: BU Mach AirSense 10 AutoSet - Mask = F&P Brevida BU ResMed P10 - Pressure = 10-17 On CPAP since 12/05/2008 Prior ID on CPAPTalk.com = JimW203

User avatar
Pugsy
Posts: 65417
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Airsense 10 for her Pressure Relief Setting

Post by Pugsy » Tue Dec 19, 2017 4:49 pm

JimW159 wrote:If I understand what has been written thus far, I think that "Essentials Plus" must have been selected through the clinical menu in order for users to be able to control the EPR function from the "My Options" screen.

I have Essentials set to Plus....and in My Options screen the only thing I see is Pressure Relief (it is under humidity I think) but it's an On/OFF thing and no mention as to EPR setting 1, 2 or 3.
So to actually set EPR I have to go into the clinical menu.

I just now looked...so there is a Pressure Relief setting for turning it on/off but not for adjusting the setting.
Never noticed it was there until now though...I always just go into the setup menu area for any changing of anything anyway.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

triciar
Posts: 2
Joined: Tue Dec 19, 2017 12:14 pm

Re: Airsense 10 for her Pressure Relief Setting

Post by triciar » Tue Dec 19, 2017 7:13 pm

Thank you so much for the help, changing it to the Essentials Plus worked! I'm wasn't sure if the setting would affect me. I was more worried they sold me an really old or somehow different model. Now I just have to hope it helps!

Tricia