Overfilling Dreamstation Auto-BiPAP humidifier

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
bernardc
Posts: 11
Joined: Sun Aug 06, 2017 11:01 am

Overfilling Dreamstation Auto-BiPAP humidifier

Post by bernardc » Sat Dec 16, 2017 10:52 pm

FYI:
My Dreamstation Auto-BiPAP quit the other night with the cryptic error message, "Low Leak: Check Mask and Tube." The user manual, of no help, says "Blockage at tube or mask." No, there's no problem with the mask and tube. The provider manual doesn't cover malfunctions. The service manual echos the user manual. So much for the manuals.

Here's the real story: Spilled water from overfilling gets into the adapter that connects the humidifier to the main unit. You'd think the water would just get blown out, but because of the odd fan curve, a weak blockage won't get blown out. The odd fan curve is what makes the pressure-flow relationship funny, making EPAP lower pressure than IPAP. In particular, the pressure is lowest when the flow is blocked. "Low leak" means that there's not enough leakage to allow the controller to turn up the fan power and raise the pressure.

Repair: Disconnect the humidifier from the main unit. Remove the water tray. It's probably adequate to shake out all the water you can, shaking like a madman in every direction. You could also blow it out by mouth. If you like to tinker, you can remove the adapter and silicone gaskets and dry it all out. Run the blower to verify that the error has vanished.

Another method that would probably work: Set the mode to fixed CPAP, and crank the pressure all the way up. That should blow out the water. Then return to Auto BiPAP.

I hope this helps someone.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Overfilling Dreamstation Auto-BiPAP humidifier

Post by chunkyfrog » Sat Dec 16, 2017 11:09 pm

Is that your cat?
Totally gorgeous!

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
bernardc
Posts: 11
Joined: Sun Aug 06, 2017 11:01 am

Re: Overfilling Dreamstation Auto-BiPAP humidifier

Post by bernardc » Sun Dec 17, 2017 2:35 am

Is that your cat?
Totally gorgeous!
Yes, thank you, he's my cat, displaying his ongoing fascination with water from the kitchen faucet.

User avatar
MrsRinPDX
Posts: 246
Joined: Tue Nov 21, 2017 9:53 am
Location: Portland, OR USA

Re: Overfilling Dreamstation Auto-BiPAP humidifier

Post by MrsRinPDX » Sun Dec 17, 2017 3:33 am

Thanks for the tip. I've also noticed that if you slightly overfill, you can hear the water being moved by the air blown over it. As you can attest, the fill line is darn hard to see on this model so this is easy to do. Thankfully, I haven't had any water in the unit, and I simply stop the machine and pour a little water out of the reservoir.

And I agree, beautiful kitty!

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Pressures: IPAP 18/ EPAP 14; Secondary mask - Nuance Pro Nasal Pillow

rob47ert
Posts: 1
Joined: Thu May 27, 2021 3:47 pm

Re: Overfilling Dreamstation Auto-BiPAP humidifier

Post by rob47ert » Thu May 27, 2021 3:51 pm

I have had this same problem, it would give the msg then shutdown.
I disconnected the blower from the water part and turned blower on and shuck it , then let it blow for 15 minutes.
After that i reassembled and had no further issues.

User avatar
CyndiLooWho
Posts: 7
Joined: Mon May 17, 2021 12:47 pm

Re: Overfilling Dreamstation Auto-BiPAP humidifier

Post by CyndiLooWho » Fri May 28, 2021 5:16 am

Agree that fill line is hard to see. I marked mine with a sharpie. It's not very dark on the plastic, but it's easier to see than the original!
CPAP since Jan 2021
Respironics Dream Station w/ humidifier
Dreamwear nasal pillows mask
Central KY

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Overfilling Dreamstation Auto-BiPAP humidifier

Post by chunkyfrog » Fri May 28, 2021 11:00 am

You answered a post that was FOUR YEARS OLD.
I imagine the dreamstation has dried out by now.
:lol: :lol: :lol:

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

hzwlzgd72
Posts: 1
Joined: Tue Aug 23, 2022 9:25 am

Re: Overfilling Dreamstation Auto-BiPAP humidifier

Post by hzwlzgd72 » Tue Aug 23, 2022 9:28 am

Anyone using the new (2022) Dreamstation getting the 'low leak' error message? I have tried the steps suggested in this post with no success. I am going on the 3rd night with little to no sleep, definitely broken sleep :(

User avatar
Pugsy
Posts: 65245
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Overfilling Dreamstation Auto-BiPAP humidifier

Post by Pugsy » Tue Aug 23, 2022 9:33 am

hzwlzgd72 wrote:
Tue Aug 23, 2022 9:28 am
Anyone using the new (2022) Dreamstation getting the 'low leak' error message? I have tried the steps suggested in this post with no success. I am going on the 3rd night with little to no sleep, definitely broken sleep :(
I suggest you start a new topic of your own with your problem.
This is an old thread and pertains to a machine totally different from your machine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.