How can I tell if my humidifier is working?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
axxell6307
Posts: 62
Joined: Tue Jan 03, 2017 3:41 pm

How can I tell if my humidifier is working?

Post by axxell6307 » Thu Dec 07, 2017 7:45 am

Auto cpap dreamsttation. It's very dry here this time of year. Have humidity up all the way on machine and it's still feeling dry.

User avatar
knothead
Posts: 538
Joined: Fri Jan 24, 2014 8:58 am
Location: Corpus Christi, Tx

Re: How can I tell if my humidifier is working?

Post by knothead » Thu Dec 07, 2017 7:55 am

Should be loosing water if its working. Do you have to fill the tank before going to bed?

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: BP down & in control, Resmed S9 Autoset for backup & travel.

User avatar
Pugsy
Posts: 65137
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: How can I tell if my humidifier is working?

Post by Pugsy » Thu Dec 07, 2017 9:04 am

There's a preheat feature available on your machine you can check it.

If it's using water it's working.
Do you have the heated hose or the non heated hose?
Using the heated hose will allow the machine to crank out more moisture than you can get with the non heated hose.
I think that even with the non heated hose you can change the mode from adaptive to the old classic humidity mode (but I need to check the manual to be sure or you can try it and see what your options are other than adaptive) which will put out more humidity.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Jasba64
Posts: 100
Joined: Sun Oct 30, 2016 4:22 pm
Contact:

Re: How can I tell if my humidifier is working?

Post by Jasba64 » Thu Dec 07, 2017 9:08 am

I recommend heated tubing along with tube and humidity settings each on 3. Has worked very well for me.

_________________
MachineMaskHumidifier
Additional Comments: Heated Tubing, SleepyHead Software, Pressure 9-15

User avatar
LSAT
Posts: 13353
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: How can I tell if my humidifier is working?

Post by LSAT » Thu Dec 07, 2017 12:41 pm

Jasba64 wrote:I recommend heated tubing along with tube and humidity settings each on 3. Has worked very well for me.
These settings may or may not work since you don't know where the OP lives. Humidity needs are different in different areas of the US. The humidity needs in AZ are much different from MI.

axxell6307
Posts: 62
Joined: Tue Jan 03, 2017 3:41 pm

Re: How can I tell if my humidifier is working?

Post by axxell6307 » Sat Dec 09, 2017 3:35 pm

Jasba64 wrote:I recommend heated tubing along with tube and humidity settings each on 3. Has worked very well for me.
So my humidifier is working. Turns out, my heater is running when I thought it should not. I have heat set to zero. How many heat settings are there for it? It's the plate under the tank that is giving heat. I want the heat off.

I'm using the tube that came with it.

axxell6307
Posts: 62
Joined: Tue Jan 03, 2017 3:41 pm

Re: How can I tell if my humidifier is working?

Post by axxell6307 » Sat Dec 09, 2017 3:37 pm

Pugsy wrote:There's a preheat feature available on your machine you can check it.

If it's using water it's working.
Do you have the heated hose or the non heated hose?
Using the heated hose will allow the machine to crank out more moisture than you can get with the non heated hose.
I think that even with the non heated hose you can change the mode from adaptive to the old classic humidity mode (but I need to check the manual to be sure or you can try it and see what your options are other than adaptive) which will put out more humidity.
It's using water so that rules that out. I fill it every day or two. The problem now is I realize the plate under the tank is giving heat and I don't want heat. I've got preheat off and the heat setting at zero. Shoudl it be giving any heat??? Do I need to contact my DME for troubleshooting or exchange?

Does resetting these ever fix issues?

axxell6307
Posts: 62
Joined: Tue Jan 03, 2017 3:41 pm

Re: How can I tell if my humidifier is working?

Post by axxell6307 » Sat Dec 09, 2017 3:44 pm

Pugsy wrote:There's a preheat feature available on your machine you can check it.

If it's using water it's working.
Do you have the heated hose or the non heated hose?
Using the heated hose will allow the machine to crank out more moisture than you can get with the non heated hose.
I think that even with the non heated hose you can change the mode from adaptive to the old classic humidity mode (but I need to check the manual to be sure or you can try it and see what your options are other than adaptive) which will put out more humidity.
Just read manual and looks like heated tube makes it automatically go to heated humidifier. I need to get a non heated hose then. If the heated hose doesn't give you the option of turning heat off, seems they should give you non heated hose too.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: How can I tell if my humidifier is working?

Post by Wulfman... » Sat Dec 09, 2017 4:34 pm

From the Provider manual, Page 4:

Humidification
This setting allows you to select the Humidification Mode being used. You can choose between Fixed or Adaptive (A) Humidification. If a heated tube is attached to the device, then the device will automatically switch to Heated Tube

Humidification Mode.
Fixed mode applies a constant heat on the humidifier heater plate. Under certain conditions and settings, this mode can allow condensation to occur in the tube. Adaptive mode adapts the heater plate temperature to the ambient conditions in the room, and is designed to not allow condensation to occur in the tube.

Humidifier
This setting allows you to choose the desired humidity setting for the humidifier: 0, 1, 2, 3, 4 or 5.

Tube Temperature
This setting allows you to choose the desired temperature for the heated tube: 0, 1, 2, 3, 4 or 5.


Try setting all to 0 (zero).

If you need a copy of the Provider manual, PM me with your email address.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: How can I tell if my humidifier is working?

Post by Wulfman... » Sat Dec 09, 2017 4:42 pm

Or, if you hurry, you can get it here:

http://incenter.medical.philips.com/doc ... Id%3d1.DAD


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

axxell6307
Posts: 62
Joined: Tue Jan 03, 2017 3:41 pm

Re: How can I tell if my humidifier is working?

Post by axxell6307 » Sat Dec 09, 2017 4:56 pm

Wulfman... wrote:From the Provider manual, Page 4:

Humidification
This setting allows you to select the Humidification Mode being used. You can choose between Fixed or Adaptive (A) Humidification. If a heated tube is attached to the device, then the device will automatically switch to Heated Tube

Humidification Mode.
Fixed mode applies a constant heat on the humidifier heater plate. Under certain conditions and settings, this mode can allow condensation to occur in the tube. Adaptive mode adapts the heater plate temperature to the ambient conditions in the room, and is designed to not allow condensation to occur in the tube.

Humidifier
This setting allows you to choose the desired humidity setting for the humidifier: 0, 1, 2, 3, 4 or 5.

Tube Temperature
This setting allows you to choose the desired temperature for the heated tube: 0, 1, 2, 3, 4 or 5.


Try setting all to 0 (zero).

If you need a copy of the Provider manual, PM me with your email address.


Den

.
Thanks. Yes, at all zero, heater plate stayed off. However, doesn't this disable the humidifier and will result in drier air? So am I correct in that if wanting to turn humidity on, the heater plate will be used either fixed or adaptively?

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: How can I tell if my humidifier is working?

Post by Wulfman... » Sat Dec 09, 2017 5:01 pm

axxell6307 wrote:
Wulfman... wrote:From the Provider manual, Page 4:

Humidification
This setting allows you to select the Humidification Mode being used. You can choose between Fixed or Adaptive (A) Humidification. If a heated tube is attached to the device, then the device will automatically switch to Heated Tube

Humidification Mode.
Fixed mode applies a constant heat on the humidifier heater plate. Under certain conditions and settings, this mode can allow condensation to occur in the tube. Adaptive mode adapts the heater plate temperature to the ambient conditions in the room, and is designed to not allow condensation to occur in the tube.

Humidifier
This setting allows you to choose the desired humidity setting for the humidifier: 0, 1, 2, 3, 4 or 5.

Tube Temperature
This setting allows you to choose the desired temperature for the heated tube: 0, 1, 2, 3, 4 or 5.


Try setting all to 0 (zero).

If you need a copy of the Provider manual, PM me with your email address.


Den

.
Thanks. Yes, at all zero, heater plate stayed off. However, doesn't this disable the humidifier and will result in drier air? So am I correct in that if wanting to turn humidity on, the heater plate will be used either fixed or adaptively?
Yes, it gives you "passover" (cool) humidity. Depending on where you live and the relative humidity in your bedroom, the air passing over the water will pick up "some" humidity. You have to turn on the heat to increase the humidity by increasing the water vapor in the air traveling over the water in the tank.
I've been using "passover" humidity in my machine since I started therapy. I tried the heated form for a couple of nights and decided the cooler the air, the better it kept my nasal passages open.......less snoring, lower AHIs.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

axxell6307
Posts: 62
Joined: Tue Jan 03, 2017 3:41 pm

Re: How can I tell if my humidifier is working?

Post by axxell6307 » Sat Dec 09, 2017 5:07 pm

Wulfman... wrote:
axxell6307 wrote:
Wulfman... wrote:From the Provider manual, Page 4:

Humidification
This setting allows you to select the Humidification Mode being used. You can choose between Fixed or Adaptive (A) Humidification. If a heated tube is attached to the device, then the device will automatically switch to Heated Tube

Humidification Mode.
Fixed mode applies a constant heat on the humidifier heater plate. Under certain conditions and settings, this mode can allow condensation to occur in the tube. Adaptive mode adapts the heater plate temperature to the ambient conditions in the room, and is designed to not allow condensation to occur in the tube.

Humidifier
This setting allows you to choose the desired humidity setting for the humidifier: 0, 1, 2, 3, 4 or 5.

Tube Temperature
This setting allows you to choose the desired temperature for the heated tube: 0, 1, 2, 3, 4 or 5.


Try setting all to 0 (zero).

If you need a copy of the Provider manual, PM me with your email address.


Den

.
Thanks. Yes, at all zero, heater plate stayed off. However, doesn't this disable the humidifier and will result in drier air? So am I correct in that if wanting to turn humidity on, the heater plate will be used either fixed or adaptively?
Yes, it gives you "passover" (cool) humidity. Depending on where you live and the relative humidity in your bedroom, the air passing over the water will pick up "some" humidity. You have to turn on the heat to increase the humidity by increasing the water vapor in the air traveling over the water in the tank.
I've been using "passover" humidity in my machine since I started therapy. I tried the heated form for a couple of nights and decided the cooler the air, the better it kept my nasal passages open.......less snoring, lower AHIs.


Den

.
Thanks!