Trouble breathing when I first lie down

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
InnerGlow
Posts: 49
Joined: Sat Oct 07, 2017 10:44 pm
Location: NJ

Trouble breathing when I first lie down

Post by InnerGlow » Sun Dec 03, 2017 3:30 pm

I've been gradually solving the issues that have arisen with me using the CPAP, but there's one issue I can't seem to fix. When I first lie down at night with the CPAP on, I feel like I can't breathe/am short of breath/can't get enough air (to the point where I'm afraid I'll lose control of my bladder). This lasts for a couple minutes as I try to relax and slow down my breathing. Sometimes I'm able to work through it and then the rest of the night is fine. Other times I end up taking the mask off, catch my breath, and then I try again. Occasionally, I don't have this issue when I lie down, but I'm unable to figure out why those times are different.

I don't think the problem is that the starting pressure (6.5) is too low. When I first put the mask on at night, I'm sitting on the bed. I sit with it on for about 5 minutes while I get used to breathing with it before I lie down. During that time I don't have any trouble breathing, and after the initial lying down, I don't have trouble breathing.

Anyone have any suggestions?

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack
Humidifier: DreamStation Heated Humidifier
Additional Comments: Min 7.5 cm / Max 14 cm, no ramp

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Trouble breathing when I first lie down

Post by Goofproof » Sun Dec 03, 2017 4:07 pm

Do you use the "Ramp" (Feature?)? If so turn it off. 6.5 is a low pressure, maybe it too low for you. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Trouble breathing when I first lie down

Post by Wulfman... » Sun Dec 03, 2017 4:51 pm

Suggestion:
Increase minimum pressure.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
InnerGlow
Posts: 49
Joined: Sat Oct 07, 2017 10:44 pm
Location: NJ

Re: Trouble breathing when I first lie down

Post by InnerGlow » Sun Dec 03, 2017 4:56 pm

Do you use the "Ramp" (Feature?)?
No, haven't used the ramp feature since the first few days when I found it useless (it was set to 4!).

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack
Humidifier: DreamStation Heated Humidifier
Additional Comments: Min 7.5 cm / Max 14 cm, no ramp

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Trouble breathing when I first lie down

Post by Goofproof » Sun Dec 03, 2017 5:06 pm

InnerGlow wrote:
Do you use the "Ramp" (Feature?)?
No, haven't used the ramp feature since the first few days when I found it useless (it was set to 4!).
I never used it, I need air, in that case I'd up the air pressure, I don't know your mask, maybe if it uses has other air Prongs or pillows with it the move more air. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

TedVPAP
Posts: 975
Joined: Sat Jun 12, 2010 10:29 am

Re: Trouble breathing when I first lie down

Post by TedVPAP » Sun Dec 03, 2017 5:35 pm

I feel this way some times. What I do is to increase my breathing rate but make it shallow for about 10 seconds, then I relax and start saying my prayers. Among other benefits, praying occupies my mind so my breathing can naturally stabilize.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/

lilly747
Posts: 450
Joined: Thu Sep 27, 2012 9:39 am
Location: Northern Calif, USA

Re: Trouble breathing when I first lie down

Post by lilly747 » Sun Dec 03, 2017 5:43 pm

TedVPAP wrote:I feel this way some times. What I do is to increase my breathing rate but make it shallow for about 10 seconds, then I relax and start saying my prayers. Among other benefits, praying occupies my mind so my breathing can naturally stabilize.

I had this problem when I first started using cpap and over time I learned to relax and it just went away.

Janknitz
Posts: 8518
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Trouble breathing when I first lie down

Post by Janknitz » Sun Dec 03, 2017 6:42 pm

Is it just for the first 10 - 20 seconds or so? My machine is set with no ramp, but it doesn't come on full blast like my older PRS1. It takes 10 or 20 seconds to get to full pressure and I feel like I'm suffocating for just that little while.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
Neicy
Posts: 43
Joined: Sun Feb 26, 2012 2:20 pm
Location: Massachusetts USA

Re: Trouble breathing when I first lie down

Post by Neicy » Sun Dec 03, 2017 7:42 pm

I was suffocating when My doc started me 6-12. Try 8 minimum.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Amara View Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: 21/25 PS3 A work in progress.
Neicy
Non-treated AHI = 29 non-REM, 55 REM
Diagnosed Nov. 2011
Started CPAP Jan. 25, 2012

TedVPAP
Posts: 975
Joined: Sat Jun 12, 2010 10:29 am

Re: Trouble breathing when I first lie down

Post by TedVPAP » Sun Dec 03, 2017 8:07 pm

InnerGlow wrote:I've been gradually solving the issues that have arisen with me using the CPAP, but there's one issue I can't seem to fix. When I first lie down at night with the CPAP on, I feel like I can't breathe/am short of breath/can't get enough air (to the point where I'm afraid I'll lose control of my bladder). This lasts for a couple minutes as I try to relax and slow down my breathing. Sometimes I'm able to work through it and then the rest of the night is fine. Other times I end up taking the mask off, catch my breath, and then I try again. Occasionally, I don't have this issue when I lie down, but I'm unable to figure out why those times are different.

I don't think the problem is that the starting pressure (6.5) is too low. When I first put the mask on at night, I'm sitting on the bed. I sit with it on for about 5 minutes while I get used to breathing with it before I lie down. During that time I don't have any trouble breathing, and after the initial lying down, I don't have trouble breathing.

Anyone have any suggestions?
I am interpreting this to mean that you have the machine on so you are at pressure - correct?
When sitting there your breathing rate is slower and deeper than when you are normally awake. When falling asleep, the breathing rate needs to increase (and become more shallow). My data shows that it takes me about 4 minutes to transition from a respiratory rate from 10 to 16. I suggest laying down and being ready to sleep and then turn the machine on.

There is confusion as to whether you are uncomfortable when you first lay down, or later after you started to sleep. If the problem occurs when you are sleeping, then your pressure is likely off. Post data using the three links below would help us understand if your pressure settings are correct.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/
Last edited by TedVPAP on Sun Dec 03, 2017 8:47 pm, edited 1 time in total.

User avatar
chronic
Posts: 105
Joined: Sat Jan 16, 2010 5:14 pm

Re: Trouble breathing when I first lie down

Post by chronic » Sun Dec 03, 2017 8:23 pm

Your low is too low and the high is too high. Someone will come on here soon an get you lined about, but, my guess is you should set the low to about 11 and the high at 15. You have way, way too much swing for the machine to handle appropriately. That is NOT the way to have it set up.

_________________
MachineMask

User avatar
JimW159
Posts: 804
Joined: Fri Jul 08, 2016 10:26 am
Location: Tarpon Springs, FL

Re: Trouble breathing when I first lie down

Post by JimW159 » Mon Dec 04, 2017 2:01 pm

How frequently do you look at your actual usage pattern with SleepyHead or Encore (if that is still useful with DreamStation)? While the machine may be set at 6.5cm min to 20cm max, that doesn't show you what mask pressures are actually being produced in response to any events that call on added pressure to clear/eliminate/prevent. Until you know what actual pressures are useful in your therapy, you have no idea of what is the 'right way' to set up your pressure range.

Among the many influences contributing to discomfort and breathing restrictions when lying down is body type. Persons who have a large upper body and heavy chest but a relatively weak diaphragm muscle may well feel like they are suffocating when lying down. I know I felt that way virtually every day for more than a month while undergoing guided x ray therapy for prostate cancer where I had to lie perfectly still for a few minutes on a stiff platform - the pressure on my chest definitely made the experience uncomfortable at best.

Regarding xPAP, in my case I have set the ramp to auto and it ramps up to the starting pressure (9cm) very quickly - a matter of a few seconds or so. When I first lie down I take several deep breaths and try to clear my mind and relax. I have discovered the worst time to think about projects, tasks, problems, or anything else that requires detailed thought or visualization is when I am trying to go to sleep. This is where my wife and I are in disagreement: she likes to fall asleep listening to stories on TV, I prefer innocuous elevator music that triggers no emotion or thought. Somehow we have worked it out so we both get the sleep we need.

Good luck finding your own sleep sweet spot.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: BU Mach AirSense 10 AutoSet - Mask = F&P Brevida BU ResMed P10 - Pressure = 10-17 On CPAP since 12/05/2008 Prior ID on CPAPTalk.com = JimW203

User avatar
Pugsy
Posts: 65373
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Trouble breathing when I first lie down

Post by Pugsy » Mon Dec 04, 2017 6:23 pm

I would NOT advise raising the minimum pressure to 11 as someone suggested without proof of need meaning an AHI that was too high and pointing to more pressure being needed.
Just because a machine can go to 20 doesn't necessarily mean that it will go to 20 ...for all we know your machine may never even go over 10.

Have you tried a starting minimum of 7 cm just in case it is just the lack of air movement that is sometimes causing the difficulty?
Just to try it to see???

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
InnerGlow
Posts: 49
Joined: Sat Oct 07, 2017 10:44 pm
Location: NJ

Re: Trouble breathing when I first lie down

Post by InnerGlow » Tue Dec 05, 2017 4:42 pm

I tried raising the minimum pressure to 7 for the last two nights and it helped. Thanks all for the advice about that.

As far as lowering the maximum pressure, it's my (newfound) understanding that the reason for this is sleep quality, so that you don't wake up from the pressure jumping up. Is that correct?

What is the best way to determine the correct maximum pressure? Do I just gradually lower it as long as my AHI is good? Do I set it based on my history (i.e. Sleepyhead data)?

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack
Humidifier: DreamStation Heated Humidifier
Additional Comments: Min 7.5 cm / Max 14 cm, no ramp

User avatar
Pugsy
Posts: 65373
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Trouble breathing when I first lie down

Post by Pugsy » Tue Dec 05, 2017 5:18 pm

If you are sleeping decently and don't seem to be waking because of the variations in pressure and you aren't having in air in the belly (aerophagia) issues then there isn't any urgent need to do anything with the maximum pressure.

You can use available software to see how high it is going if you want to but having a maximum higher than it ever goes hurts nothing.
Doesn't affect response or anything like that. Years ago the older machines used to chase leaks with more pressure and sometimes cause some runaway pressures but nowadays with these new machines they don't do that with large leaks and in fact will most often back off on the pressure.
Just because it can go higher doesn't mean it will.

Don't know about the software??
https://sleep.tnet.com/equipment

Sometimes I need much higher pressures in REM ...up around 16 or 18 but this is only in REM. The changing pressures don't wake me up and don't cause aerophagia issues so I just set the max at 20 and let the machine sort it out. Some nights it might hit 18...some nights maybe 12.
There's no urgent need to set the machine to 18 because it changes nothing in terms of how it responds except where it can go to.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.