How does an ASV work?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Pugsy
Posts: 65400
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: How does an ASV work?

Post by Pugsy » Fri Nov 17, 2017 11:24 pm

Oh...PR thinks that the ResMed algorithm is "better" because it is more responsive...or more quick to react.
I have owned both the Respironics 960 and the ResMed S9 Adapt....so both ASV machines and the ResMed is quicker to react and I was able to use less pressure starting out the night (and all night) with the ResMed because of its faster/quicker response.
It's simply more aggressive....but that's not to say that the Respironics algorithm is crap...it's not.
And I could see where some people might do better with taking things a bit slower.

PR just likes the more aggressive approach...and it suits his needs better so he likes ResMeds the best.
So do I for that matter but I can actually use both brands....I just like the more aggressive algorithm a little better myself but the end result or goal for both brands...each meets their goals. They just go about it in a little different manner.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: How does an ASV work?

Post by palerider » Sat Nov 18, 2017 3:37 am

Pugsy wrote:PR just likes the more aggressive approach...and it suits his needs better so he likes ResMeds the best.
Exactly, I've seen SO many data reports from people who're having problems sleeping, not getting rest, clearly disturbed breathing, and the respironics machines are just sitting there, sucking their thumbs, not helping. All in the name of "being less disturbing"? I call it "not being as advanced."

Also: viewtopic.php?f=1&t=99737&p=927976#p927976

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Kiralynx
Posts: 2432
Joined: Wed Oct 08, 2008 5:42 am

Re: How does an ASV work?

Post by Kiralynx » Sun Nov 19, 2017 1:02 am

xxyzx wrote: the ASV is just a bipap with a back up rate in case you stop breathing due to centrals (snip)

ASV self titrates
the machine is not that complicated
No. It is not. It does not self-titrate. And yes, it is a complex machine.

Mine was set incorrectly at first. It did not correct itself. There are a number of combinations which can be utilized depending on what is being treated.

The simplest setting is from EPAPMin (Exhalation pressure minimum) to EPAPMax (Exhalation Pressure Maximum) which gives a range of exhalation pressures, depending on what is required. Sitting on top of that range is PSMin and PSMax, (Pressure Support Minimum and Maximum) which creates a range of inhalation pressures.

There is also auto-backup, timed backup, breaths per minute, tidal, minimum vent -- this is taken from the set-up menu on my 960.

It is not "just" a bipap with a back-up rate.
-- Kiralynx
Beastie, 2008 Beastie2, PRS1 960, 2014-05 Beastie3, Dream Station ASV, 2017-10-17. Resmed Aircurve 11, 2026-06 PadaCheek Hosecover. Homemade Brandy Keg Chin Support. TapPap Mask.
Min PS = 4, Max PS = 8
Epap Range = 6 - 8

User avatar
Kiralynx
Posts: 2432
Joined: Wed Oct 08, 2008 5:42 am

Re: How does an ASV work?

Post by Kiralynx » Sun Nov 19, 2017 1:12 am

palerider wrote:
mangos21 wrote:also, any difference between respironics ASVs and resmed ASVs? Krakow seems to exclusively rx resmed's
because they're better.
I think I would have to disagree with you on that, Palerider.

I've used three different Respironics ASVs. The original, which had a setting just for EPAP and then a Min and MaxIPAP. My 960, which allows for a range of EPAP, plus pressure support on top of that range to give a range of IPAP, and the Dreamstation ASV.

I will note that there were changes in the algorithms between my original Beastie and the 960 -- I needed a slightly higher EPAP on it, so set a range. I also noted a difference between the 960 and the Dreamstation -- my 960 had the 2.xx software. When I got it back from Accurate Biomed, they'd upgraded it to the latest software, and there was no significant difference between it and the Dreamstation.

I will acknowledge that I have not had the opportunity to try a Resmed ASV. After all, if what you have works, and works well, why try to fix what ain't broken?

So I don't know if it is, indeed, "better" than a Respironics. But my Respironics machines have done well by me for nine years.
-- Kiralynx
Beastie, 2008 Beastie2, PRS1 960, 2014-05 Beastie3, Dream Station ASV, 2017-10-17. Resmed Aircurve 11, 2026-06 PadaCheek Hosecover. Homemade Brandy Keg Chin Support. TapPap Mask.
Min PS = 4, Max PS = 8
Epap Range = 6 - 8

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: How does an ASV work?

Post by palerider » Sun Nov 19, 2017 2:24 pm

Kiralynx wrote:
palerider wrote:
mangos21 wrote:also, any difference between respironics ASVs and resmed ASVs? Krakow seems to exclusively rx resmed's
because they're better.
I think I would have to disagree with you on that, Palerider. ....I will acknowledge that I have not had the opportunity to try a Resmed ASV. After all, if what you have works, and works well, why try to fix what ain't broken?

So I don't know if it is, indeed, "better" than a Respironics. But my Respironics machines have done well by me for nine years.
The resmeds that I've taken apart are built better than the respironics I've taken apart, there are many reports of build quality failures in the respironics system one line. (burning humidifier connections, etc).

Resmed invented the ASV, and, (I believe) has a better algorithm that works better for more people.

Does that mean that you can't be very well served by your respironics? of course not. And, like you say, if it's working fine for you, why change?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Thewino
Posts: 231
Joined: Sat May 04, 2013 5:13 pm

Re: How does an ASV work?

Post by Thewino » Mon Nov 20, 2017 4:54 am

Kiralynx wrote:
palerider wrote:
mangos21 wrote:also, any difference between respironics ASVs and resmed ASVs? Krakow seems to exclusively rx resmed's
because they're better.
I think I would have to disagree with you on that, Palerider.

I've used three different Respironics ASVs. The original, which had a setting just for EPAP and then a Min and MaxIPAP. My 960, which allows for a range of EPAP, plus pressure support on top of that range to give a range of IPAP, and the Dreamstation ASV.

I will note that there were changes in the algorithms between my original Beastie and the 960 -- I needed a slightly higher EPAP on it, so set a range. I also noted a difference between the 960 and the Dreamstation -- my 960 had the 2.xx software. When I got it back from Accurate Biomed, they'd upgraded it to the latest software, and there was no significant difference between it and the Dreamstation.

I will acknowledge that I have not had the opportunity to try a Resmed ASV. After all, if what you have works, and works well, why try to fix what ain't broken?

So I don't know if it is, indeed, "better" than a Respironics. But my Respironics machines have done well by me for nine years.
Kiralynx and Pugsy,

I’ve been using the Respironics BiPAP autoSV Advanced for quite a while.

I have been wondering if I am missing out on any BiPAP ASV technological advances with with the Dreamstation machine. It looks like you’ll have had the opportunity to try them both out. Do you prefer the Dreamstation BiPAP ASV over the older Respironics autoSV Advanced machine?

Thank you,
TheWino
The wino and I know the joy of the ocean... J. Buffett ***Please do not take any information from my post/s as medical advice. I am simply providing personal experiences. Thank you***

User avatar
Pugsy
Posts: 65400
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: How does an ASV work?

Post by Pugsy » Mon Nov 20, 2017 5:12 am

Thewino wrote: I have been wondering if I am missing out on any BiPAP ASV technological advances with with the Dreamstation machine. It looks like you’ll have had the opportunity to try them both out. Do you prefer the Dreamstation BiPAP ASV over the older Respironics autoSV Advanced machine?
I probably won't be having the opportunity to try the new DreamStation ASV to compare it to the System One 960 ASV. Sorry.
From what I have read the way the Respironics ASV works didn't change though. Just some minor cosmetic changes along with a couple more minor settings changes with humidifier humidity selection choices. I don't know that these minor changes alone would be enough to impact the therapy itself.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
bdz
Posts: 102
Joined: Fri Dec 20, 2013 7:35 pm
Location: PSL, Florida

Re: How does an ASV work?

Post by bdz » Mon Nov 20, 2017 4:45 pm

I have centrals and do I ever love it when I feel my ASV kick in if my breathing hesitates. Usually I don't feel it at all, only if I have a major hesitation in my breathing

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Luckily the Oracle is not flimsy. Bought several. Thanks to JDS74, I have lots of inner flaps.
Peace begins with each of us as individuals

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: How does an ASV work?

Post by palerider » Tue Nov 21, 2017 1:53 am

Thewino wrote:I’ve been using the Respironics BiPAP autoSV Advanced for quite a while.

I have been wondering if I am missing out on any BiPAP ASV technological advances with with the Dreamstation machine. It looks like you’ll have had the opportunity to try them both out. Do you prefer the Dreamstation BiPAP ASV over the older Respironics autoSV Advanced machine?
To the best of my knowledge, neither Respironics or Resmed changed the operating parameters of their ASV machines between the current and previous models... just the shape.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 65400
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: How does an ASV work?

Post by Pugsy » Tue Nov 21, 2017 8:20 am

palerider wrote:To the best of my knowledge, neither Respironics or Resmed changed the operating parameters of their ASV machines between the current and previous models... just the shape.
That's my thoughts also.
The basic guts of how the machines go about their business (the working algorithms) didn't change...just the packaging and a few very minor comfort type of options.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Kiralynx
Posts: 2432
Joined: Wed Oct 08, 2008 5:42 am

Re: How does an ASV work?

Post by Kiralynx » Wed Nov 22, 2017 5:54 pm

Thewino wrote:
Kiralynx and Pugsy,

I’ve been using the Respironics BiPAP autoSV Advanced for quite a while.

I have been wondering if I am missing out on any BiPAP ASV technological advances with with the Dreamstation machine. It looks like you’ll have had the opportunity to try them both out. Do you prefer the Dreamstation BiPAP ASV over the older Respironics autoSV Advanced machine?

Thank you,
TheWino
Given that my 960 ended on the floor in a motel, I had occasion to go back to my old Beastie. Which was just an ASV, not an ASV Advanced. I did notice a difference in the depth of breath I could take -- the 960 was smoother than the old one. When I got the Dreamstation, I noticed that the algorithm was much smoother than I remembered from my 960. However, my 960's software version was the 2.xx, and when I got it back with the 3.xx software update, I couldn't really tell any difference between the two.

So I don't know whether to suggest you go to the Dreamstation or not. I can tell you that Sleepyhead does not work with the Dreamstation ASV, but does work with the 960, if that's a factor in your consideration.
-- Kiralynx
Beastie, 2008 Beastie2, PRS1 960, 2014-05 Beastie3, Dream Station ASV, 2017-10-17. Resmed Aircurve 11, 2026-06 PadaCheek Hosecover. Homemade Brandy Keg Chin Support. TapPap Mask.
Min PS = 4, Max PS = 8
Epap Range = 6 - 8