Resmed Aircurve 10 Auto-Start Feature

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
pdx1953
Posts: 6
Joined: Sat Jul 21, 2012 5:03 am
Location: Connecticut

Resmed Aircurve 10 Auto-Start Feature

Post by pdx1953 » Sat Oct 21, 2017 4:15 pm

Greetings -

I just received a new Aircurve Bi-Pap yesterday & supposedly it'll start after the mask is on & I start breathing. My previous Bi-Pap (Respironics) would start after three or four breaths & I found it quite handy. I know how to access the clinician settings screen but can't find anything there or in the user settings. Maybe the wording is changed?

Any suggestions?

User avatar
Pugsy
Posts: 65297
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Resmed Aircurve 10 Auto-Start Feature

Post by Pugsy » Sat Oct 21, 2017 4:28 pm

On ResMed machines it is called SmartStart and it's in the clinical settings and there is only one choice.
SmartStart being on does the auto on and also will shut off when you take the mask off after just a few seconds.

Respironics machines have 2 settings to get this done.
Auto On...which means when on the machine will start after a few breaths
Auto Off...when it is on it means the machine will turn itself off after you remove the mask but it takes a lot longer than the ResMed does to turn it off. Respironics takes about a minute and ResMed takes just a few seconds.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

pdx1953
Posts: 6
Joined: Sat Jul 21, 2012 5:03 am
Location: Connecticut

Re: Resmed Aircurve 10 Auto-Start Feature

Post by pdx1953 » Sat Oct 21, 2017 5:12 pm

Thanks for the info but I just went through every line item 3 times & there's no Smart Start in either the user or clinician settings.

What sub-menu would it be under? I'm pretty sure I hit everyone of them as well.

pdx1953
Posts: 6
Joined: Sat Jul 21, 2012 5:03 am
Location: Connecticut

Re: Resmed Aircurve 10 Auto-Start Feature

Post by pdx1953 » Sat Oct 21, 2017 5:32 pm

Well - I did find Smart Start under Options but it's greyed out & can't be set. Rather strange I think unless there's something else that needs to be turned on/off to allow it to be selected.

User avatar
Pugsy
Posts: 65297
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Resmed Aircurve 10 Auto-Start Feature

Post by Pugsy » Sat Oct 21, 2017 5:39 pm

I have it available in 2 places...one from the patient screen (My Options) and the other in the clinical settings menu right under Essentials found in Settings.

If you have leak alert turned on you may not have it available. Leak alert needs to be off.
Edit...if it is grayed out it is because of Leak Alert being on...turn leak alert off and it will become available and not grayed out.

Do you have the provider manual for your machine?
If not request it here.
http://www.apneaboard.com/adjust-cpap-p ... tup-manual


Image


Image

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

pdx1953
Posts: 6
Joined: Sat Jul 21, 2012 5:03 am
Location: Connecticut

Re: Resmed Aircurve 10 Auto-Start Feature

Post by pdx1953 » Sat Oct 21, 2017 5:48 pm

"Manual? We don't need no steeking manual....." (Loosely related to a famous movie quote.)

Auto Leak was on, I turned it off & viola`!! I now have Smart Start.

Thanks a million, which isn't worth as much as it was 50 years ago.

MikeNM
Posts: 1
Joined: Sat Dec 01, 2018 10:11 am

Re: Resmed Aircurve 10 SmartStart grayed out

Post by MikeNM » Sat Dec 01, 2018 10:29 am

Pugsy, thank you so much. I just acquired a second Aircurve 10 ASV for travel, and I couldn't figure out why the SmartStart was grayed out. I incorrectly thought that perhaps the older 2015 machine didn't have that feature.

But sure enough, once I followed your instructions and turned off Leak Detection, I was able to turn on SmartStart.

One other feature that appears (??) to be related is that previously the Sleep Report was not showing my AHI score. That line did not appear on the screen. But once I turned on SmartStart, the line with AHI score showed up on the Sleep Report.

I'm so pleased to have these two features working, so now my two machines work identically.

Thanks again, Pugsy.

q49q49
Posts: 12
Joined: Tue Feb 09, 2016 9:37 am

Re: Resmed Aircurve 10 Auto-Start Feature

Post by q49q49 » Sat Dec 01, 2018 11:09 am

Thanks, Pugsy!

We picked up DH's AirCurve 10 ASV yesterday and were told that there isn't an auto-start option. I've just requested the manuals from the Apnea Board for his ASV and my AirSense 10 AutoSet. Anything to help DH adapt to the ResMed after his previous PR System One ASV

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: DreamStation autoSV EPAP(4,15) PS(2,15) Mx-30 BPM AUTO
8/21 - Resmed AirCurve 10 ASV EPAP(4,10) PS(2,7) Mx-25
2/19 DreamStation autoSV EPAP(4,15) PS(2,20) Mx-30 BPM AUTO 3/20 PSmax 15
11/18 Resmed AirCurve 10 ASV
7/13 PR AutoSV Adv System One EPAP(4,15) PS(2,20) Mx-25 Rate-Auto BPM Biflex-3

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Resmed Aircurve 10 Auto-Start Feature

Post by palerider » Sat Dec 01, 2018 2:18 pm

q49q49 wrote:
Sat Dec 01, 2018 11:09 am
Thanks, Pugsy!

We picked up DH's AirCurve 10 ASV yesterday and were told that there isn't an auto-start option. I've just requested the manuals from the Apnea Board for his ASV and my AirSense 10 AutoSet. Anything to help DH adapt to the ResMed after his previous PR System One ASV
Whoever told you that... was wrong. It's under the 'options' menu in the clinical settings. Though, Leak Alert has to be off.

Probably a good idea to discount, or at least, triple check anything that person told you.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.