just how does the xpap determine hypopnea

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
TASmart
Posts: 1071
Joined: Wed Jan 04, 2017 7:23 pm
Location: Eugene, OR

Re: just how does the xpap determine hypopnea

Post by TASmart » Fri Oct 13, 2017 9:29 pm

How it determines hypopneas is described in my Autoset 10 clinical manual - isn't it in yours?
All posts reflect my own opinion based on my experience and reading.
Your mileage may vary
Past performance is no guarantee of future results
Consult with your own physician as people very

Guest

Re: just how does the xpap determine hypopnea

Post by Guest » Fri Oct 13, 2017 9:38 pm

Page 19 of the Clinical .pdf manual:

Events (AHI) per hour
Apneas and hypopneas measured per hour for one day.
An apnea is when the respiratory flow decreases by more than 75% for at least 10 sec.
A hypopnea is when the respiratory flow decreases to 50% for at least 10 sec.
The Apnea Index (AI) and Apnea-Hypopnea Index (AHI) are calculated by dividing the total number of events that occurred by the total mask-on therapy period in hours.
Note: Under conditions of high leak with EPR enabled, AHI detection may not be optimal.


You're welcome.

User avatar
TASmart
Posts: 1071
Joined: Wed Jan 04, 2017 7:23 pm
Location: Eugene, OR

Re: just how does the xpap determine hypopnea

Post by TASmart » Fri Oct 13, 2017 10:50 pm

Bummer. I guess they figure that by the time you get into an ASV simple definitions are not needed?
All posts reflect my own opinion based on my experience and reading.
Your mileage may vary
Past performance is no guarantee of future results
Consult with your own physician as people very

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: just how does the xpap determine hypopnea

Post by palerider » Sat Oct 14, 2017 12:55 am

xxyzx wrote:
TASmart wrote:How it determines hypopneas is described in my Autoset 10 clinical manual - isn't it in yours?
=======

no
Page 13... get someone to read it to you.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: just how does the xpap determine hypopnea

Post by palerider » Sat Oct 14, 2017 12:56 am

TASmart wrote:Bummer. I guess they figure that by the time you get into an ASV simple definitions are not needed?
No, it's there, he just needs someone to help him with the big words.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Guest

Re: just how does the xpap determine hypopnea

Post by Guest » Sat Oct 14, 2017 1:54 am

I would have said: maybe a picture could help - but he already looked at his charts and couldn't see it - so that seems to be no option at all.

TedVPAP
Posts: 975
Joined: Sat Jun 12, 2010 10:29 am

Re: just how does the xpap determine hypopnea

Post by TedVPAP » Sat Oct 14, 2017 9:08 am

xxyzx wrote:
TASmart wrote:Bummer. I guess they figure that by the time you get into an ASV simple definitions are not needed?
=========

the problem is not the definition

the question is how they are claiming to have met their definition from that data

i dont see any hypops when they claim them
i do see them when they did not

no snore
no flow limitation
no flat topping of the breathing wave
no lack of air flow or reduction in Vt
nothing
and they declared a hypop
You do know that no one can help you understand why a picture does not look the way you think it should, when you don't share the picture don't you? If you want help, we need to see the facts - not your interpretations.
Since you are not interested in a serious answer, I'll vote for your Remed is guessing theory.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/

TedVPAP
Posts: 975
Joined: Sat Jun 12, 2010 10:29 am

Re: just how does the xpap determine hypopnea

Post by TedVPAP » Sat Oct 14, 2017 11:01 am

xxyzx wrote:
TedVPAP wrote:
xxyzx wrote:
TASmart wrote:Bummer. I guess they figure that by the time you get into an ASV simple definitions are not needed?
=========

the problem is not the definition

the question is how they are claiming to have met their definition from that data

i dont see any hypops when they claim them
i do see them when they did not

no snore
no flow limitation
no flat topping of the breathing wave
no lack of air flow or reduction in Vt
nothing
and they declared a hypop
You do know that no one can help you understand why a picture does not look the way you think it should, when you don't share the picture don't you? If you want help, we need to see the facts - not your interpretations.
Since you are not interested in a serious answer, I'll vote for your Remed is guessing theory.
==========

you dont know jack so you dont need anything

my question is clear
HOW do they decide that the flow was limited to declare a hypop

you dont need the pictures
just some knowledge which you have none of
just tell us HOW they decide there is a hypop not the criteria for it
the question is how they know the flow rate was limited by a hypop
I will try to help you - I do love a challenge.
HOW they decide there is a hypopnea is by APPLYING the criteria. That is how criteria is used.
What is the criteria that your machine uses???? Please state it verbatim.
Once you know that, you can then APPLY it to your data.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/

SnoringInOregon

Re: just how does the xpap determine hypopnea

Post by SnoringInOregon » Sat Oct 14, 2017 11:20 am

I've stayed away from Alphabet Boy and his conflicts with everyone here, but this thread is just too much.

It's obvious that xyzzy is an internet troll. Why do people keep responding to him?

en. wikipedia. org/wiki/Internet_troll

If there was ever a discussion that called for a picture, this is the one. And yet he won't supply a picture, even though people have specifically asked him for it.

INTERNET TROLL. AND YOU'RE ALL BEING PLAYED BY HIM.

I'm surprised it's lasted for so many months. And I'm looking forward to being called a "libtard" by him in response. Nobody has ever accused me of being a liberal in my entire life. And there's a first time for everything!

TedVPAP
Posts: 975
Joined: Sat Jun 12, 2010 10:29 am

Re: just how does the xpap determine hypopnea

Post by TedVPAP » Sat Oct 14, 2017 11:25 am

xxyzx wrote:
TedVPAP wrote:
xxyzx wrote:
TedVPAP wrote:
xxyzx wrote: =========

the problem is not the definition

the question is how they are claiming to have met their definition from that data

i dont see any hypops when they claim them
i do see them when they did not

no snore
no flow limitation
no flat topping of the breathing wave
no lack of air flow or reduction in Vt
nothing
and they declared a hypop
You do know that no one can help you understand why a picture does not look the way you think it should, when you don't share the picture don't you? If you want help, we need to see the facts - not your interpretations.
Since you are not interested in a serious answer, I'll vote for your Remed is guessing theory.
==========

you dont know jack so you dont need anything

my question is clear
HOW do they decide that the flow was limited to declare a hypop

you dont need the pictures
just some knowledge which you have none of
just tell us HOW they decide there is a hypop not the criteria for it
the question is how they know the flow rate was limited by a hypop
I will try to help you - I do love a challenge.
HOW they decide there is a hypopnea is by APPLYING the criteria. That is how criteria is used.
What is the criteria that your machine uses???? Please state it verbatim.
Once you know that, you can then APPLY it to your data.
=========


you are truly a jerk

i know the criteria

i am saying the data they show us does not support their claiming hypopnea based on the sleepyhead graphs

so do they have some other magic method that they dont release to the outside world
or do they just guess
When I apply my machine's criteria to my data, it makes perfect sense why the hypopnea event was flagged.
I don't know what the criteria is for your machine since you did not state it. Nor can I comment about how your data looks relative to your hypopnea event flag since you did not show it.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/

User avatar
JimW159
Posts: 804
Joined: Fri Jul 08, 2016 10:26 am
Location: Tarpon Springs, FL

Re: just how does the xpap determine hypopnea

Post by JimW159 » Sat Oct 14, 2017 11:27 am

xxyzx wrote: you dont know jack so you dont need anything

my question is clear
HOW do they decide that the flow was limited to declare a hypop

you dont need the pictures
just some knowledge which you have none of
just tell us HOW they decide there is a hypop not the criteria for it
the question is how they know the flow rate was limited by a hypop
If that is the case, why ask us?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: BU Mach AirSense 10 AutoSet - Mask = F&P Brevida BU ResMed P10 - Pressure = 10-17 On CPAP since 12/05/2008 Prior ID on CPAPTalk.com = JimW203

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: just how does the xpap determine hypopnea

Post by Goofproof » Sat Oct 14, 2017 1:00 pm

Ask a Real Engineer, I've heard they know everything! Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

TedVPAP
Posts: 975
Joined: Sat Jun 12, 2010 10:29 am

Re: just how does the xpap determine hypopnea

Post by TedVPAP » Sat Oct 14, 2017 1:32 pm

xxyzx wrote:you are truly a jerk

i know the criteria

i am saying the data they show us does not support their claiming hypopnea based on the sleepyhead graphs

so do they have some other magic method that they dont release to the outside world
or do they just guess
When I apply my machine's criteria to my data, it makes perfect sense why the hypopnea event was flagged.
I don't know what the criteria is for your machine since you did not state it. Nor can I comment about how your data looks relative to your hypopnea event flag since you did not show it.[/quote]
============

my machine does not tell HOW i t determines the criteria was met

there are no decreases in flow or anything else that would indicate hypops on my data
it has to have something else hidden that it uses but does not end up in sleepyheads printouts[/quote]

Maybe your right.
Maybe your wrong.
Maybe your confused.
Maybe SH didn't put the event flag in the correct location.
Maybe ...

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/

User avatar
ColinP
Posts: 349
Joined: Wed Nov 21, 2007 7:26 am
Location: South Africa

Re: just how does the xpap determine hypopnea

Post by ColinP » Mon Oct 16, 2017 1:43 am

JimW159 wrote:
xxyzx wrote: you dont know jack so you dont need anything

my question is clear
HOW do they decide that the flow was limited to declare a hypop

you dont need the pictures
just some knowledge which you have none of
just tell us HOW they decide there is a hypop not the criteria for it
the question is how they know the flow rate was limited by a hypop
If that is the case, why ask us?
Because, when he’s only an engineer because Mommy gave him a big plastic hammer and screwdriver set for his last birthday, all these graphs and science stuff are a little beyond his intellectual grasp.