Sleepy Head Leak Rate Question

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
klm49
Posts: 442
Joined: Sat May 28, 2016 4:24 pm
Location: Texas

Sleepy Head Leak Rate Question

Post by klm49 » Fri Oct 13, 2017 10:46 am

(Don't Answer xxyzx, you are on ignore)

Does anyone know how Sleepyhead Leak Rate is arrived at? Is their a calculation done.
Reason I ask is I was looking at someone's leak rate and the the S/H graph showed a rate that was less than what the manufacturer shows as the vent rate for the mask at a given pressure.. Curious how this can happen?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Amara View Full Face CPAP Mask with Headgear
Additional Comments: Setting: PS 4.0 over 14.0-25.0; Humidifer 4 & Heated Hose
Resmed AirCurve10 Vauto w/humidifier
Amara View mask
O2 - 2-4 lpm

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Sleepy Head Leak Rate Question

Post by Goofproof » Fri Oct 13, 2017 11:48 am

Sleephead can only read the data your machine gives it, all machines aren't created equal.

The two main brands of machines measure leak rate totally differently, one shows mask vent+leak (for people that can do basic math), the other measures just the leak not including not the mask vent, (for those that can't do basic math, and have to guess that the machine they have is set for their mask). Jim

I prefer to do my own math. A paper comes with the mask has a chart of pressure at vent leak numbers.

I think Resmed units measure just other leaks not including vent leaks. (only if you use their masks and are set correctly. I consider this a POOR method.

Remstar or Phillips, combine leaks for a total so they are higher, you just have to check the specs of your mask at your pressure, anything over that pressure is a unwanted leak. Jim (Preferred) by me.
Last edited by Goofproof on Fri Oct 13, 2017 12:06 pm, edited 1 time in total.
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
TASmart
Posts: 1071
Joined: Wed Jan 04, 2017 7:23 pm
Location: Eugene, OR

Re: Sleepy Head Leak Rate Question

Post by TASmart » Fri Oct 13, 2017 12:08 pm

As noted above, Resmed subtracts a mask type typical leak rate from the total leak rate, and that's what is reported. Other machines do things differently.
All posts reflect my own opinion based on my experience and reading.
Your mileage may vary
Past performance is no guarantee of future results
Consult with your own physician as people very

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Sleepy Head Leak Rate Question

Post by Goofproof » Fri Oct 13, 2017 12:15 pm

Incorrect as usual, Leak rate is the output of your XPAP that vented thru the mask plus any leaks! Resmed just thinks you are to Dumb to compute the extra leak on your own, so they guess your mask leak rate, which many users have set incorrectly. So the result is wrong. BAD RESMED!!! Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
TASmart
Posts: 1071
Joined: Wed Jan 04, 2017 7:23 pm
Location: Eugene, OR

Re: Sleepy Head Leak Rate Question

Post by TASmart » Fri Oct 13, 2017 12:23 pm

I was agreeing with you Jim.
All posts reflect my own opinion based on my experience and reading.
Your mileage may vary
Past performance is no guarantee of future results
Consult with your own physician as people very

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Sleepy Head Leak Rate Question

Post by Goofproof » Fri Oct 13, 2017 1:23 pm

TASmart wrote:I was agreeing with you Jim.
Was directed at xxxzx. I should have noted that.
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
Pugsy
Posts: 65327
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Sleepy Head Leak Rate Question

Post by Pugsy » Fri Oct 13, 2017 1:32 pm

klm49 wrote:Does anyone know how Sleepyhead Leak Rate is arrived at? Is their a calculation done.
Reason I ask is I was looking at someone's leak rate and the the S/H graph showed a rate that was less than what the manufacturer shows as the vent rate for the mask at a given pressure.. Curious how this can happen?
What machine?

It depends on the machine and how the leak gets reported.
ResMed machines report only excess leak. Vent rates are removed prior to reporting the leak so those will always be less than vent rate unless the excess leak is quite high.

Now with Respironics machines which report total leak sometimes the total leak is less than expected vent rate by a little bit just because of the normal plus or minus variance of the vent rate. When that happens the person simply isn't having any excess leaks to speak of.
I used to see it occasionally with my Respironics reports on Encore ....leak line for total leak was actually slightly below the expected vent rate for my mask at my pressure.

SleepyHead only reports what the machine records. Doesn't really do any leak calculations except the one calculation for % of time above the red line leak threshold.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
klm49
Posts: 442
Joined: Sat May 28, 2016 4:24 pm
Location: Texas

Re: Sleepy Head Leak Rate Question

Post by klm49 » Fri Oct 13, 2017 4:07 pm

Pugsy wrote:
klm49 wrote:Does anyone know how Sleepyhead Leak Rate is arrived at? Is their a calculation done.
Reason I ask is I was looking at someone's leak rate and the the S/H graph showed a rate that was less than what the manufacturer shows as the vent rate for the mask at a given pressure.. Curious how this can happen?
What machine?

It depends on the machine and how the leak gets reported.
ResMed machines report only excess leak. Vent rates are removed prior to reporting the leak so those will always be less than vent rate unless the excess leak is quite high.

Now with Respironics machines which report total leak sometimes the total leak is less than expected vent rate by a little bit just because of the normal plus or minus variance of the vent rate. When that happens the person simply isn't having any excess leaks to speak of.
I used to see it occasionally with my Respironics reports on Encore ....leak line for total leak was actually slightly below the expected vent rate for my mask at my pressure.

SleepyHead only reports what the machine records. Doesn't really do any leak calculations except the one calculation for % of time above the red line leak threshold.
Thanks Pugsy & all (excluding the alphabet poster)
Curious,
So when looking at someone else's leak rate I guess it might be a good idea to see if their "mask type" has been set.
I think I will experiment with my "mask type" setting and see what it does to my leak rate.
I'll report back after a few nights.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Amara View Full Face CPAP Mask with Headgear
Additional Comments: Setting: PS 4.0 over 14.0-25.0; Humidifer 4 & Heated Hose
Resmed AirCurve10 Vauto w/humidifier
Amara View mask
O2 - 2-4 lpm

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Sleepy Head Leak Rate Question

Post by Goofproof » Fri Oct 13, 2017 4:19 pm

klm49 wrote:
Pugsy wrote:
klm49 wrote:Does anyone know how Sleepyhead Leak Rate is arrived at? Is their a calculation done.
Reason I ask is I was looking at someone's leak rate and the the S/H graph showed a rate that was less than what the manufacturer shows as the vent rate for the mask at a given pressure.. Curious how this can happen?
What machine?

It depends on the machine and how the leak gets reported.
ResMed machines report only excess leak. Vent rates are removed prior to reporting the leak so those will always be less than vent rate unless the excess leak is quite high.

Now with Respironics machines which report total leak sometimes the total leak is less than expected vent rate by a little bit just because of the normal plus or minus variance of the vent rate. When that happens the person simply isn't having any excess leaks to speak of.
I used to see it occasionally with my Respironics reports on Encore ....leak line for total leak was actually slightly below the expected vent rate for my mask at my pressure.

SleepyHead only reports what the machine records. Doesn't really do any leak calculations except the one calculation for % of time above the red line leak threshold.
Thanks Pugsy & all (excluding the alphabet poster)
Curious,
So when looking at someone else's leak rate I guess it might be a good idea to see if their "mask type" has been set.
I think I will experiment with my "mask type" setting and see what it does to my leak rate.
I'll report back after a few nights.
Just because the mask has been set doesn't mean it was set correctly, hence errors. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
Pugsy
Posts: 65327
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Sleepy Head Leak Rate Question

Post by Pugsy » Fri Oct 13, 2017 4:34 pm

Respironics machines don't have a mask type selection like ResMed machines do.
Respironics machines do have the mask resistance setting thing but that is NOT used as part of any leak calculations.
When I had a PR S1 machine I played with all the mask resistance settings available including off when using my ResMed nasal pillow mask and there wasn't anything anywhere that I could find that changed with any of the settings including off.
I slept all night too at each setting choice....not just a quick 15 minute test while awake.
Didn't feel any difference and nothing different in the data from the machine above the normal very minor nightly variations.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.