still struggling

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
jchilds582
Posts: 3
Joined: Mon Jun 19, 2017 7:11 pm
Location: Connecticut

still struggling

Post by jchilds582 » Mon Oct 09, 2017 8:48 am

Hi All,
I am struggling with mask fit and CPAP in hopes someone will have some thoughts. I was diagnosed with severe sleep apnea 49 events an hour so I am trying really hard to be compliant.
I am using a full face, as a mouth breather cant seem to handle anything else Resmed F20 both with and without memory foam. The machine ramps up from 4.5-13 depending on what is needed.
I am having problems keeping the mask on longer than 3-4 hours and seem to get a lot of leaking. I struggle to adjust and wind up going back to the lab for multiple adjustments. I don't want anything more evasive like chin straps etc, and sleep on my back. I am not sure if its leaks or something else I do at night but the numbers don't impress me. Over the last two months I am running between 3-4 hours with 2 nights of 5 and 6. Events are reduced to 1.8-4.4. last night was a record 6.55 hours but 19 events each hour. My leaks are always high in the high 30s. I am frustrated don't want to give up. Also my smile muscles hurt every morning . I would welcome any help. I live in CT and there doesn't seem to be any local support groups. I have talked to the provider resmed the vender and the dr and still cant figure it out. The only other mask I used was the amora view which really hurt my facial muscles.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Additional Comments: struggling with mask amora view and airfit F20 having alot of problems with both

User avatar
MaxINTJ
Posts: 632
Joined: Mon Jul 24, 2017 6:56 am
Location: Missouri

Re: still struggling

Post by MaxINTJ » Mon Oct 09, 2017 9:12 am

jchilds582 wrote:Hi All,
I am struggling with mask fit and CPAP in hopes someone will have some thoughts. I was diagnosed with severe sleep apnea 49 events an hour so I am trying really hard to be compliant.
I am using a full face, as a mouth breather cant seem to handle anything else Resmed F20 both with and without memory foam. The machine ramps up from 4.5-13 depending on what is needed.
I am having problems keeping the mask on longer than 3-4 hours and seem to get a lot of leaking. I struggle to adjust and wind up going back to the lab for multiple adjustments. I don't want anything more evasive like chin straps etc, and sleep on my back. I am not sure if its leaks or something else I do at night but the numbers don't impress me. Over the last two months I am running between 3-4 hours with 2 nights of 5 and 6. Events are reduced to 1.8-4.4. last night was a record 6.55 hours but 19 events each hour. My leaks are always high in the high 30s. I am frustrated don't want to give up. Also my smile muscles hurt every morning . I would welcome any help. I live in CT and there doesn't seem to be any local support groups. I have talked to the provider resmed the vender and the dr and still cant figure it out. The only other mask I used was the amora view which really hurt my facial muscles.
At what pressure are you fitting the mask for leaks? I am wrestling with issues similar in which I don't get leaks at the beginning pressure but get leaks when the pressure spikes. Maybe narrow the pressure range to something like 8-13 or 10-13 with no ramp?
Phillips 960 AutoSV Paving Brick, Phillips Dreamwear Mask - Nothing is working.
Diagnosis of crappy sleep, desats under 80, maybe UARS

User avatar
TASmart
Posts: 1071
Joined: Wed Jan 04, 2017 7:23 pm
Location: Eugene, OR

Re: still struggling

Post by TASmart » Mon Oct 09, 2017 9:14 am

The way I understand what you have said is that you are using an APAP of 4.5 to 13. That is a very wide range of pressures and that makes leaks more likely. A ramp is something totally different, a ramp is where the machine starts at a low pressure when you turn it on then slowly increases to a base pressure over time to make it easier to get to sleep. IF you are indeed using an APAP you would need to download sleepyhead for free and install it and get some nightly information posted for people to look at. If you so a search for sleepyhead thee are a number of posters who have given detailed instruction on how to get it and use it.

Fundamentally I would narrow that range, it's not very likely you need that much variation in your therapeutic pressure. Personally, I find anything more than 2 cm h2o causes disruptions to my sleep quality so I generally do best just sticking with CPAP. You will probably differ
All posts reflect my own opinion based on my experience and reading.
Your mileage may vary
Past performance is no guarantee of future results
Consult with your own physician as people very

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: still struggling

Post by Okie bipap » Mon Oct 09, 2017 9:54 am

Use the mask fit option on your machine. It will ramp up to the highest pressure your machine is set to. Check the fit for leaks, then lay down and see how it works. When I first started and had problems with leaks, I would put the machine on fit check, make sure my mask wasn't leaking, and get into bed with the machine still running in the fit check mode. After two to three minutes, the machine would start running normally.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

User avatar
MaxINTJ
Posts: 632
Joined: Mon Jul 24, 2017 6:56 am
Location: Missouri

Re: still struggling

Post by MaxINTJ » Mon Oct 09, 2017 10:01 am

Okie bipap wrote:Use the mask fit option on your machine. It will ramp up to the highest pressure your machine is set to. Check the fit for leaks, then lay down and see how it works. When I first started and had problems with leaks, I would put the machine on fit check, make sure my mask wasn't leaking, and get into bed with the machine still running in the fit check mode. After two to three minutes, the machine would start running normally.
I was under the impression the machine would only go to your average pressure. Do they all go to maximum pressure?
Phillips 960 AutoSV Paving Brick, Phillips Dreamwear Mask - Nothing is working.
Diagnosis of crappy sleep, desats under 80, maybe UARS

Dubbayoo
Posts: 12
Joined: Sun Jul 19, 2015 3:39 pm

Re: still struggling

Post by Dubbayoo » Mon Oct 09, 2017 10:06 am

I would look into the various mask liners. I had a lot of leakage issues till I found those.

User avatar
TASmart
Posts: 1071
Joined: Wed Jan 04, 2017 7:23 pm
Location: Eugene, OR

Re: still struggling

Post by TASmart » Mon Oct 09, 2017 10:17 am

MaxINTJ wrote:
Okie bipap wrote:Use the mask fit option on your machine. It will ramp up to the highest pressure your machine is set to. Check the fit for leaks, then lay down and see how it works. When I first started and had problems with leaks, I would put the machine on fit check, make sure my mask wasn't leaking, and get into bed with the machine still running in the fit check mode. After two to three minutes, the machine would start running normally.
I was under the impression the machine would only go to your average pressure. Do they all go to maximum pressure?
THe machine does not know your average pressure until after you use it. It does not use the previous day's data for anything. So yes it uses the max pressure setting for the fit check. Do you have the clinicians manual? There is a link to get it over on the sleep apnea board. It's got a lot of very detailed information for you.
All posts reflect my own opinion based on my experience and reading.
Your mileage may vary
Past performance is no guarantee of future results
Consult with your own physician as people very

User avatar
MaxINTJ
Posts: 632
Joined: Mon Jul 24, 2017 6:56 am
Location: Missouri

Re: still struggling

Post by MaxINTJ » Mon Oct 09, 2017 10:34 am

TASmart wrote:THe machine does not know your average pressure until after you use it. It does not use the previous day's data for anything. So yes it uses the max pressure setting for the fit check. Do you have the clinicians manual? There is a link to get it over on the sleep apnea board. It's got a lot of very detailed information for you.
I couldn't find where it says what pressure it uses - from the manual:

"Mask Fit is designed to help patients fit the mask properly.
The Mask Fit feature delivers CPAP pressure for a three-minute period,
prior to starting treatment. During this time, the mask can be adjusted to
minimize leaks."
Phillips 960 AutoSV Paving Brick, Phillips Dreamwear Mask - Nothing is working.
Diagnosis of crappy sleep, desats under 80, maybe UARS

User avatar
TASmart
Posts: 1071
Joined: Wed Jan 04, 2017 7:23 pm
Location: Eugene, OR

Re: still struggling

Post by TASmart » Mon Oct 09, 2017 10:37 am

IS that your ASV? I forgot you were using that, I have no clue what that does. but I would bet it's just the max pressure. ASV also does not keep track of your previous data, each day is a new start.
All posts reflect my own opinion based on my experience and reading.
Your mileage may vary
Past performance is no guarantee of future results
Consult with your own physician as people very

a mess

Re: still struggling

Post by a mess » Mon Oct 09, 2017 2:19 pm

My machine was set by Drs office to ramp up from the pressure from 5 to 20. Its pretty stable at about 13-14. Still, no matter what I do still have leaks. AHIs running at about 3.5 down from 68 per hr. If the mask did not leak at all would I get a perfect score of no leaks?

User avatar
ChicagoGranny
Posts: 15386
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: still struggling

Post by ChicagoGranny » Mon Oct 09, 2017 2:46 pm

jchilds582 wrote: jchilds582
To reduce unintentional mask leak -

- Make sure the mask is fitted properly. (See mask manufacturer's video.)

- Make sure the headgear is adjusted properly (See mask manufacturer's video.)

- If you use a low ramp pressure or a low minimum pressure, you should fit and adjust your mask at the higher pressures it is operating at for most of the night. You would temporarily change the machine pressure settings while you do the fitting and adjusting. It's best to do this well before bedtime before you are tired and ready for sleep. Don't forget to change your settings back to the ones you prefer.

- Make sure to use good hose management - the hose should not pull on the mask. Notice how your hose is positioned between machine and mask. It should be positioned to minimize any pull on the mask. Hose management is an individual practice. How it is done depends on the mask, the position of the machine, your sleeping position and other factors specific to the individual.

- If you still have problems, learn from the forum what mask liners are available.

- Start a thread in cpaptalk asking for help. Subject: Need Help with Mask xxxxxx

- If your mask still has excessive leak, try other masks. Many people report trying several masks before they find one that works well for them.

Note: Links to just some of the mask manufacturers' instructional videos can be found on the forum host's website - https://www.cpap.com/cpap-masks.php and https://www.youtube.com/

- CG

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: still struggling

Post by Okie bipap » Mon Oct 09, 2017 4:15 pm

a mess wrote:If the mask did not leak at all would I get a perfect score of no leaks?
It is doubtful if you could get a night without having some leak. When my machine starts, I get air blowing out of the holes in the elbow until the pressure gets high enough to close the flapper valves. This gets reported as a leak. If the mask moves a little during the night, it leaks a little bit. I have never had a night where I did not have some leakage show up.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

User avatar
TASmart
Posts: 1071
Joined: Wed Jan 04, 2017 7:23 pm
Location: Eugene, OR

Re: still struggling

Post by TASmart » Mon Oct 09, 2017 4:55 pm

xxyzx wrote:
TASmart wrote:The way I understand what you have said is that you are using an APAP of 4.5 to 13. That is a very wide range of pressures and that makes leaks more likely. A ramp is something totally different, a ramp is where the machine starts at a low pressure when you turn it on then slowly increases to a base pressure over time to make it easier to get to sleep. IF you are indeed using an APAP you would need to download sleepyhead for free and install it and get some nightly information posted for people to look at. If you so a search for sleepyhead thee are a number of posters who have given detailed instruction on how to get it and use it.

Fundamentally I would narrow that range, it's not very likely you need that much variation in your therapeutic pressure. Personally, I find anything more than 2 cm h2o causes disruptions to my sleep quality so I generally do best just sticking with CPAP. You will probably differ

it is not the difference that causes leaks
it is the maximum pressure that can cause them

many people need a wide range for treatment

Not the difference but the cycling of pressures seems to increasethe leakage.
All posts reflect my own opinion based on my experience and reading.
Your mileage may vary
Past performance is no guarantee of future results
Consult with your own physician as people very

User avatar
ChicagoGranny
Posts: 15386
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: still struggling

Post by ChicagoGranny » Mon Oct 09, 2017 5:25 pm

xxyzx wrote:the mask is designed to leak some
so you dont have co2 buildup in it and keep breathing that air
has to be someplace for the exhale to go so leaks are built in
That's not leak. It's vent.

Learn the terminology.

User avatar
Bill44133
Posts: 1087
Joined: Tue Jul 09, 2013 8:34 pm
Location: North Royalton, OH

Re: still struggling

Post by Bill44133 » Mon Oct 09, 2017 5:51 pm

Dubbayoo wrote:I would look into the various mask liners. I had a lot of leakage issues till I found those.
+1
Liners are great solution to plug up the leaks. Plus it helps to have a material on your face and NOT the mask directly on your skin.

https://www.cpap.com/productpage/remzzz ... iners.html

http://www.padacheek.com/

I wish you well

_________________
Machine: DreamStation BiPAP® Auto Machine
Humidifier: DreamStation Heated Humidifier
Additional Comments: Settings are IPap 23 EPap 19