CPAP just isn't working. Need advice

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Longwalksonthe
Posts: 16
Joined: Sun Apr 30, 2017 9:09 am

CPAP just isn't working. Need advice

Post by Longwalksonthe » Mon Sep 18, 2017 7:49 am

Hello everyone,

I've posted here before asking for advice and most everyone said i should raise my min pressure, so i did. If i raise it higher than 6.2ish i get really bad Aerophagia, even sometimes at 6.2 or lower i will get it as well, but not nearly as much.

So i've left it at 6.2 min and 20 max for a few weeks and i also have been using max liners because i drool really bad and will always wake up because of the drool and have to wipe off my mask. The mask liners have helped with that, but i think they are also causing large leaks, as you will be able to see in my data.

I've been using CPAP for almost a year, and it's been a nightmare since the beginning. I can remember maybe a handful of times where i've really noticed it helped me but 85% of the time i'm still feeling drowsy all day, especially while driving which is why i got this in the first place.

I did an at home study in the beginning and was determined to have about 15 events per hour. I'm 23, not overweight at all, go to the gym every week and have a good diet. I cannot deal with this my whole life and i've been thinking about trying to get an in lab sleep study to really pinpoint what my issues are and then try to get surgery to fix this. CPAP obviously isn't working and my sleep doctor doesn't seem to know crap so this is why i'm coming here again.

Here is some of my data:
https://imgur.com/a/4dtDp

You will notice i usually use it 3-5 times a night, sometimes longer than other times because i literally can't stand the machine, sometimes i'll wake up with really bad aerophagia and just take the mask off because my stomach is absolutely killing me. Whats the point in using it if it hasn't helped me for a year.

So any advice you have would be greatly appreciated. Thanks

_________________
MachineMask

User avatar
Pugsy
Posts: 65413
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: CPAP just isn't working. Need advice

Post by Pugsy » Mon Sep 18, 2017 8:08 am

Talk to your doctor about using a bilevel machine because of the aerophagia issues.

6 hours of sleep using the machine isn't enough to feel better and add in wake ups due to belly issues...a recipe for failure.

There are times where your pressure needs are significantly higher than your minimum pressure. You are already using maximum exhale relief...time to consider using a different machine to hopefully better help with the aerophagia.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Longwalksonthe
Posts: 16
Joined: Sun Apr 30, 2017 9:09 am

Re: CPAP just isn't working. Need advice

Post by Longwalksonthe » Mon Sep 18, 2017 8:12 am

Pugsy wrote:Talk to your doctor about using a bilevel machine because of the aerophagia issues.

6 hours of sleep using the machine isn't enough to feel better and add in wake ups due to belly issues...a recipe for failure.

There are times where your pressure needs are significantly higher than your minimum pressure. You are already using maximum exhale relief...time to consider using a different machine to hopefully better help with the aerophagia.
Thank you for all the replies you've given me since i first came to this site! I will talk to my doctor about that then. Do you know how that would work with insurance coverage? I just graduated college so i don't have the cash to shell out for a new machine.. But insurance should cover the new machine just like they did the old machine right?

Also, do you think i should try and get an in lab sleep study as well since they give so much more information than an at home study?

_________________
MachineMask

User avatar
LSAT
Posts: 13412
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: CPAP just isn't working. Need advice

Post by LSAT » Mon Sep 18, 2017 8:28 am

It's NOT the min pressure that is causing you problems. You are averaging 11-12 for most of the night. When you start the machine it starts at the min and then takes you to whatever pressure your auto machine feels you need to stop events. I'll bet that if you set the machine at about 8-13, you would feel better but your AHI may increase. Another option would be to set the machine to CPAP rather than AUTO and set pressure at 12.

Longwalksonthe
Posts: 16
Joined: Sun Apr 30, 2017 9:09 am

Re: CPAP just isn't working. Need advice

Post by Longwalksonthe » Mon Sep 18, 2017 8:33 am

LSAT wrote:It's NOT the min pressure that is causing you problems. You are averaging 11-12 for most of the night. When you start the machine it starts at the min and then takes you to whatever pressure your auto machine feels you need to stop events. I'll bet that if you set the machine at about 8-13, you would feel better but your AHI may increase. Another option would be to set the machine to CPAP rather than AUTO and set pressure at 12.
But wouldn't that give me bad aerophagia since i get it when my min pressure is higher than 6.2?

_________________
MachineMask

User avatar
Pugsy
Posts: 65413
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: CPAP just isn't working. Need advice

Post by Pugsy » Mon Sep 18, 2017 8:58 am

If your doctor will get on board and document the need for bilevel because of inability to tolerate the pressures you need because of the aerophagia your insurance should cover the expense.

And yes...going to fixed pressure is likely to cause the aerophagia to be worse but you are using a lot more than 6.2 most of the time anyway. It's going quite high and staying up for a considerable bit of time.

You might be able to find a workable compromise with your current machine by limiting the maximum and letting a few apnea events slide by and that could be tried while waiting on bilevel trial and/or approval.

I wonder what would happen if you did 7 minimum and 12 maximum???? I think that most likely the AHI will be too high but you never know...maybe those high pressures are because of the Flow limitations and maybe allow some FLs to happen would be an acceptable compromise.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Longwalksonthe
Posts: 16
Joined: Sun Apr 30, 2017 9:09 am

Re: CPAP just isn't working. Need advice

Post by Longwalksonthe » Mon Sep 18, 2017 9:01 am

Pugsy wrote:If your doctor will get on board and document the need for bilevel because of inability to tolerate the pressures you need because of the aerophagia your insurance should cover the expense.

And yes...going to fixed pressure is likely to cause the aerophagia to be worse but you are using a lot more than 6.2 most of the time anyway. It's going quite high and staying up for a considerable bit of time.

You might be able to find a workable compromise with your current machine by limiting the maximum and letting a few apnea events slide by and that could be tried while waiting on bilevel trial and/or approval.

I wonder what would happen if you did 7 minimum and 12 maximum???? I think that most likely the AHI will be too high but you never know...maybe those high pressures are because of the Flow limitations and maybe allow some FLs to happen would be an acceptable compromise.
Okay, thank you very much for all the advice. I will contact my doctor this week and try those suggestions as well, Thank you.!

_________________
MachineMask

User avatar
fog.apnea
Posts: 38
Joined: Wed Aug 23, 2017 10:18 am
Location: Maryland, USA

Re: CPAP just isn't working. Need advice

Post by fog.apnea » Tue Sep 19, 2017 6:21 am

Is your prescribed max setting really 20? Since you mentioned using the hose therapy for an year am assuming you might have tried to fiddle with the max pressure setting earlier on ... do you recall noticing any difference (if and) when you lowered the max pressure setting?. Couple more responses also indicating you need to take a look closer at your max pressure setting.

In my personal experience I have noticed that the apap machines are more reactive to settings on the machine than what you need, so the settings should be configured as best as possible to what you need. A single titration study is not going to be accurate for few reasons as I understand. My prescribed setting was 10-13 and during a 2 mo visit to the doc I tried to convince him that the min is a bit high for me based on the charts .. he didn't agree (limited time to go through the charts in detail) so I took it down 1cm bit a time myself and currently at 8-13 and my therapy response is no different. Not saying it might be that case for your but take it FWIW, do your own due diligence and ask your doc the right questions in response to what you noticed.

_________________
MachineMask
Additional Comments: Software - SleepyHead
APAP pressure: 9.2-10.6
EPR 2cm
Humidity Level: Auto
Diagnosis: OSA 35 AHI. Diagnosed May '17. Therapy started Jun '17.

mangos21
Posts: 57
Joined: Tue Jul 19, 2016 3:40 pm

Re: CPAP just isn't working. Need advice

Post by mangos21 » Thu Sep 21, 2017 8:34 am

have you tried sleeping in other positions.. a poster named roster solved his aerophagia and cpap issues by sleeping in the "falcon position"

Longwalksonthe
Posts: 16
Joined: Sun Apr 30, 2017 9:09 am

Re: CPAP just isn't working. Need advice

Post by Longwalksonthe » Thu Sep 21, 2017 12:07 pm

fog.apnea wrote:Is your prescribed max setting really 20? Since you mentioned using the hose therapy for an year am assuming you might have tried to fiddle with the max pressure setting earlier on ... do you recall noticing any difference (if and) when you lowered the max pressure setting?. Couple more responses also indicating you need to take a look closer at your max pressure setting.

In my personal experience I have noticed that the apap machines are more reactive to settings on the machine than what you need, so the settings should be configured as best as possible to what you need. A single titration study is not going to be accurate for few reasons as I understand. My prescribed setting was 10-13 and during a 2 mo visit to the doc I tried to convince him that the min is a bit high for me based on the charts .. he didn't agree (limited time to go through the charts in detail) so I took it down 1cm bit a time myself and currently at 8-13 and my therapy response is no different. Not saying it might be that case for your but take it FWIW, do your own due diligence and ask your doc the right questions in response to what you noticed.

Yea, my prescribed settings were 5-20. I've tried messing with the settings to no avail. I've lowered max to as low as like 12, i've lowered and raised min settings, i've tried everything and my doctor doesn't really seem to know much because when i told her i was changing settings, she said "you're not even supposed to have access to those settings" and that was it. I've slowly lowered settings over time and my results are all scattered.

_________________
MachineMask

Guest

Re: CPAP just isn't working. Need advice

Post by Guest » Thu Sep 21, 2017 12:39 pm

LSAT wrote:It's NOT the min pressure that is causing you problems. You are averaging 11-12 for most of the night. When you start the machine it starts at the min and then takes you to whatever pressure your auto machine feels you need to stop events. I'll bet that if you set the machine at about 8-13, you would feel better but your AHI may increase. Another option would be to set the machine to CPAP rather than AUTO and set pressure at 12.
+1

It's obvious that ranges of pressures ain't workin' for ya. In essence, you're doing the same thing over and over and expecting different results. Time to "K I S S" (Keep It Simple) and see how that works for awhile. Look at your reports where the pressure is when you DON'T have any events and take some clues from that. When the pressure goes up, you have more.

User avatar
Cpapian
Posts: 417
Joined: Sat Jun 24, 2017 2:00 pm
Location: Toronto, Ontario

Re: CPAP just isn't working. Need advice

Post by Cpapian » Thu Sep 21, 2017 12:59 pm

Have you always used a full face mask?

The reason I ask is I just recently went from a nasal mask to nasal pillows. It felt more refreshing with the pillows, stronger force from the air. So I turned the pressure down from 10.5 to 9.5 and results are basically the same so far.

What I am thinking is the larger the mask, the more undirected air floating around. The mask being kind of like the plemun of a furnace. So what blows into the mask and doesn't go into your lungs is a lot and you swallow it.

If you tried a smaller mask type and lowered your maximum pressure ( a lot) you might get the same amount of amount of air in your lungs but less in your stomach.

Just something to think about.

_________________
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack
Additional Comments: SleepyHead , P10 and Dreamwear FFM, Airfit N20

Rhrh
Posts: 7
Joined: Fri Aug 18, 2017 3:55 pm
Location: nyc

Re: CPAP just isn't working. Need advice

Post by Rhrh » Fri Sep 22, 2017 8:12 pm

6 hours per night is not enough to feel better? I'm averaging like 4 hours, maybe 4.5, and I was told I was doing great!

I do feel like crap still, and well, I've got a lot of new issues that may be coincidental.

I am not sleeping more than 6 hours per night since I first starting having sleep issues 6 months ago, CPAP or not. I have a lot of aerophagia too, among other things.

With insomnia before CPAP, and still present, I could not get 6 hours of sleep regularly.
-----------------------------------------------------------
Newbie with supportive spouse :)
Started CPAP August 1, 2017
ResMed AirSense 10 Autoset machine.
ResMed F20 Ultrasoft Memory Foam mask.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: CPAP just isn't working. Need advice

Post by palerider » Sat Sep 23, 2017 9:41 am

Rhrh wrote:6 hours per night is not enough to feel better? I'm averaging like 4 hours, maybe 4.5, and I was told I was doing great!
You were told wrong, you need to use the cpap all night, every night. Not using it is not sleeping.

Few, if any, people are going to feel good only sleeping four hours a night.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
LSAT
Posts: 13412
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: CPAP just isn't working. Need advice

Post by LSAT » Sat Sep 23, 2017 12:27 pm

xxyzx aka idiot