Im back and trying again! Grade my logs! Issues...

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
SouthSeaPirate
Posts: 45
Joined: Mon Jun 20, 2016 8:45 am
Location: Florida, USA

Im back and trying again! Grade my logs! Issues...

Post by SouthSeaPirate » Sat Sep 02, 2017 9:51 am

Hi all
Im back and trying this again
Gave up on my useless 'insurance' and just bought what I needed myself

Already have Resmed Airsense 10
Bought a heated hose (which I still have rainout), Nuance Pro, and filters.

Issues as of first use; getting rainout, nose a lil' red, and when I blew my nose I had a small bit of blood.

Screenshot of logs and settings;

Image

Image

Looking for thoughts on this and how to fix issues.
No idea how well or bad I did.

TIA!

User avatar
Pugsy
Posts: 65075
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Im back and trying again! Grade my logs! Issues...

Post by Pugsy » Sat Sep 02, 2017 10:20 am

For rain out issues and options check this thread posts #2 and 3.
viewtopic/t94035/Pugsys-Pointers-3Deali ... -road.html

Is your nasal mucosa a bit dry? That might explain the slightly bloody nose....might try increasing the humidity level..right now set to 3 which is on the low side.

I can't enlarge your screen shot and can't see all the numbers but what I do see like AHI and leak rate are well within acceptable ranges.
So in that area....you did very well last night.
No large leaks and AHI well below 5.0.

Go get some Lansinoh lanolin ointment/cream and apply it to your nostril area and a little up inside where it might be dried out.
It can be found in the baby aisle of most stores. It's used by breast feeding mom's to help with dried/cracked/painful nipple issues.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

SouthSeaPirate
Posts: 45
Joined: Mon Jun 20, 2016 8:45 am
Location: Florida, USA

Re: Im back and trying again! Grade my logs! Issues...

Post by SouthSeaPirate » Sat Sep 02, 2017 1:33 pm

xxyzx wrote:your AHI is good

if you still feel tired then look for other causes than apnea
Thank you, good to know lol I had no idea
Feel fine, maybe not as good as i would have expected if I did really good, but deff fine

SouthSeaPirate
Posts: 45
Joined: Mon Jun 20, 2016 8:45 am
Location: Florida, USA

Re: Im back and trying again! Grade my logs! Issues...

Post by SouthSeaPirate » Sat Sep 02, 2017 1:39 pm

Pugsy wrote:For rain out issues and options check this thread posts #2 and 3.
viewtopic/t94035/Pugsys-Pointers-3Deali ... -road.html

Is your nasal mucosa a bit dry? That might explain the slightly bloody nose....might try increasing the humidity level..right now set to 3 which is on the low side.

I can't enlarge your screen shot and can't see all the numbers but what I do see like AHI and leak rate are well within acceptable ranges.
So in that area....you did very well last night.
No large leaks and AHI well below 5.0.

Go get some Lansinoh lanolin ointment/cream and apply it to your nostril area and a little up inside where it might be dried out.
It can be found in the baby aisle of most stores. It's used by breast feeding mom's to help with dried/cracked/painful nipple issues.
Yeah it did feel a bit dry, kinda like a wind burn when I would breathe in. At what level do you think I should set it?
Is AHI when the machine failed or when the machine had to jump in? Sorry I still don't 100% understand this stuff.
Here are the links to the pics, should be much larger;

http://www.use.com/images/s_3/2017_09_0 ... 5baa35.jpg

http://www.use.com/images/s_3/2017_09_0 ... aa35_2.jpg

Ill have a look at the ointment. Im thinking maybe I just had it too tight. At least thats what the documentation says about redness.

Another note, I just realized I need to check those settings. They do not look right at all. I had it set to nasal, not full mask... Climate control was auto...
Im going to start readin up on your threads here in a bit. Thanks!!!

User avatar
Pugsy
Posts: 65075
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Im back and trying again! Grade my logs! Issues...

Post by Pugsy » Sat Sep 02, 2017 1:46 pm

Humidity setting depends a lot on what your nose/skin wants and what you are starting out with in your bedroom ambient humidity.

Since you are already having rain out issues....increase the hose air temp also...I would make small changes in the humidity setting. Like 1 at a time.

The settings stuff reported by SH...some of that is known buggy and reports weird stuff. Confirm on your machine that you have what you think you have and don't worry about what SH reports in that settings area.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
RogerSC
Posts: 1910
Joined: Sat Jul 21, 2012 1:11 pm

Re: Im back and trying again! Grade my logs! Issues...

Post by RogerSC » Sat Sep 02, 2017 2:16 pm

SouthSeaPirate wrote: Yeah it did feel a bit dry, kinda like a wind burn when I would breathe in. At what level do you think I should set it?
Is AHI when the machine failed or when the machine had to jump in? Sorry I still don't 100% understand this stuff.
Here are the links to the pics, should be much larger;

http://www.use.com/images/s_3/2017_09_0 ... 5baa35.jpg

http://www.use.com/images/s_3/2017_09_0 ... aa35_2.jpg

Ill have a look at the ointment. Im thinking maybe I just had it too tight. At least thats what the documentation says about redness.

Another note, I just realized I need to check those settings. They do not look right at all. I had it set to nasal, not full mask... Climate control was auto...
Im going to start readin up on your threads here in a bit. Thanks!!!
Your AHI as shown by Sleepyhead represents the apneas, etc. that were not prevented by the CPAP machine. Apneas that were prevented by your CPAP aren't shown in your sleep data.

SouthSeaPirate
Posts: 45
Joined: Mon Jun 20, 2016 8:45 am
Location: Florida, USA

Re: Im back and trying again! Grade my logs! Issues...

Post by SouthSeaPirate » Sat Sep 02, 2017 2:19 pm

Pugsy wrote:Humidity setting depends a lot on what your nose/skin wants and what you are starting out with in your bedroom ambient humidity.

Since you are already having rain out issues....increase the hose air temp also...I would make small changes in the humidity setting. Like 1 at a time.

The settings stuff reported by SH...some of that is known buggy and reports weird stuff. Confirm on your machine that you have what you think you have and don't worry about what SH reports in that settings area.
Living in Florida it's prob pretty high lol
Ill increase to 4 and change hose temp
Good to know. Ill deff chack my settings again just to be sure
Thanks so much!!!

SouthSeaPirate
Posts: 45
Joined: Mon Jun 20, 2016 8:45 am
Location: Florida, USA

Re: Im back and trying again! Grade my logs! Issues...

Post by SouthSeaPirate » Sat Sep 02, 2017 2:21 pm

Oh AHI was the events my Dr was talking about then?
So I went from 45 per hour to 1.34 per night?!

User avatar
Pugsy
Posts: 65075
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Im back and trying again! Grade my logs! Issues...

Post by Pugsy » Sat Sep 02, 2017 2:49 pm

SouthSeaPirate wrote:Oh AHI was the events my Dr was talking about then?
So I went from 45 per hour to 1.34 per night?!
Yep the 45 per hour was pre cpap therapy.
the 1.34 is with the therapy and a huge improvement.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

TedVPAP
Posts: 975
Joined: Sat Jun 12, 2010 10:29 am

Re: Im back and trying again! Grade my logs! Issues...

Post by TedVPAP » Sat Sep 02, 2017 2:59 pm

SouthSeaPirate wrote:Oh AHI was the events my Dr was talking about then?
So I went from 45 per hour to 1.34 per night?!
Your reading is 1.34 events per hour.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/

SouthSeaPirate
Posts: 45
Joined: Mon Jun 20, 2016 8:45 am
Location: Florida, USA

Re: Im back and trying again! Grade my logs! Issues...

Post by SouthSeaPirate » Sat Sep 02, 2017 3:01 pm

TedVPAP wrote:
SouthSeaPirate wrote:Oh AHI was the events my Dr was talking about then?
So I went from 45 per hour to 1.34 per night?!
Your reading is 1.34 events per hour.
Oh lol
Silly me

Well, at least way better

User avatar
Julie
Posts: 20056
Joined: Tue Feb 28, 2006 12:58 pm

Re: Im back and trying again! Grade my logs! Issues...

Post by Julie » Sat Sep 02, 2017 3:49 pm

Set your machine on something a bit lower than your bed... you shouldn't get rainout then.

SouthSeaPirate
Posts: 45
Joined: Mon Jun 20, 2016 8:45 am
Location: Florida, USA

Re: Im back and trying again! Grade my logs! Issues...

Post by SouthSeaPirate » Mon Sep 04, 2017 8:10 pm

So changed settings and rainout got worse, way worse

Machine is already lower than my bed

User avatar
Pugsy
Posts: 65075
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Im back and trying again! Grade my logs! Issues...

Post by Pugsy » Mon Sep 04, 2017 8:18 pm

Hose air temp has to be warmer some how some fashion.
See this thread posts #2 and 3

viewtopic/t94035/Pugsys-Pointers-3Deali ... -road.html

Putting the machine lower doesn't work well with ResMed machines anyway. I have the AirSense and with the way the hose attaches at the back any water in the hose can't drain well anyway.
People who aren't using this machine don't have a clue how that hose is attached to the humidifier.
Water has to make a 90 degree turn and it will pool in the hose first.
Besides...water in the mask usually has to go uphill some how anyway...and it can't go uphill.
Exception might be if you slept with face totally face down and not many people do that.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

TedVPAP
Posts: 975
Joined: Sat Jun 12, 2010 10:29 am

Re: Im back and trying again! Grade my logs! Issues...

Post by TedVPAP » Mon Sep 04, 2017 8:22 pm

SouthSeaPirate wrote:So changed settings and rainout got worse, way worse

Machine is already lower than my bed
I only get rainout when it is cold in my room. What temperature is your bedroom? Are you sure that you are not overfilling your humidifier?

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/