ASV and (another) High Altitude Question

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Uncle Flapp
Posts: 247
Joined: Thu Nov 14, 2013 2:24 pm
Location: Arizona

ASV and (another) High Altitude Question

Post by Uncle Flapp » Mon Aug 07, 2017 1:22 pm

We're headed to a wedding in Lima and off to the Andes for a few days. Not overly concerned about Machu Picchu as the elevation there is 8,000 feet; however, Cusco is much higher at 11,000. To ward off altitude sickness, I already got my prescription for Diamox and will be drinking cocoa tea. Any concerns about my PR960 ASV? I am hoping that it will actually help me sleep better than most by treating altitude-induced CAs.

In case it has any bearing, I do rather well with my machine settings at the lower end of the spectrum. Machine settings and last night's statistics are below:
  • Settings:
    EPAP Min: 8
    EPAP Max: 15
    PS: 6
    IPAP Min: 14
    IPAP Max: 25

    Last night:
    EPAP 95%: 9
    IPAP 95%: 15
    PTBs: 83.5%
    PB: 0%
    OA: 0
    CA: 0
    HYP: 0.56
    AHI: 0.56
How should I expect my machine to perform? Any concerns?

- Flappy

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dreamstation ASV. UPPP in 2007; Untreated AHI 84

User avatar
Pugsy
Posts: 65196
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ASV and (another) High Altitude Question

Post by Pugsy » Mon Aug 07, 2017 1:35 pm

I don't know that you will see or feel any differences in performance.
The machine will attempt to compensate for the altitude changes as best it can and there's not much we can do about it.
How it actually compensates I don't really understand but again since I can't do much about it I haven't tried real hard to understand it.

I will say that I have used my PR S1 model 760 at about 12,500 altitude one weekend when we camped in the mountains in Utah and I didn't notice a thing difference and actually never even thought about it until after the fact.
Checked my data and all that and nothing really changed in terms of anything. Machine responded pretty much like it always did and the AHI was pretty much what it always was and I slept great and felt great. I did NOT see evidence of the machine going to pressures that I normally don't see. If I didn't know the dates at the higher altitude I wouldn't be able to spot any changes in the reports that would point to a higher altitude. Pretty much same old reports just a different date.

Have a good time and enjoy the trip and take whatever extra precautions might be needed to make sure nothing that is preventable isn't covered and it sounds like you are doing that.

Report back and let us know if you spot anything unusual on the reports or how you feel. Not many people here get a chance to use their equipment at these altitudes so we don't have a lot of experience to draw upon.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

WickedLoki
Posts: 40
Joined: Sat Aug 20, 2016 2:27 pm

Re: ASV and (another) High Altitude Question

Post by WickedLoki » Mon Aug 07, 2017 3:41 pm

Typically they compensate up to 8500 feet per the manufacturers (One more correction of misinformation). I agree with Pugsy concerning higher altitudes. With an ASV in auto mode the machine should provide the necessary pressure for your conditions anyway.

User avatar
Uncle Flapp
Posts: 247
Joined: Thu Nov 14, 2013 2:24 pm
Location: Arizona

Re: ASV and (another) High Altitude Question

Post by Uncle Flapp » Mon Aug 07, 2017 4:07 pm

Thank you for the replies. Sounds like I shouldn't have any issues. I'll post back after the trip.

- Flappy

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dreamstation ASV. UPPP in 2007; Untreated AHI 84

klv329
Posts: 497
Joined: Mon Aug 31, 2015 6:43 pm
Location: Bradenton, Florida

Re: ASV and (another) High Altitude Question

Post by klv329 » Mon Aug 07, 2017 5:25 pm

Maybe a trip to new mexico for a back up resmed aircurve asv for $500, although it is a 2015...

https://albuquerque.craigslist.org/app/ ... 47390.html



Or this one if you are going through atlanta, for $500 obo for a resmed aircurve asv...


https://atlanta.craigslist.org/eat/hab/ ... 50105.html


Funny how they both refence 50 hours of usage...

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Additional Comments: Resmed Aircurve 10 ASVAUTO Min Epap 10.4, Max Epap 11.6, PS 1.6-12.0

ajack
Posts: 977
Joined: Thu Mar 02, 2017 2:54 am
Location: australia

Re: ASV and (another) High Altitude Question

Post by ajack » Mon Aug 07, 2017 6:03 pm

If you are getting UA and H. raise the min epap a cm or 2.

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: S9 ST-A iVAPS and adapt ASV

User avatar
Pugsy
Posts: 65196
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ASV and (another) High Altitude Question

Post by Pugsy » Mon Aug 07, 2017 6:08 pm

If you are getting UA and H.
One needs to note that the Respironics ASV machines don't do UAs...they separate centrals and obstructives on their reporting.
Now the ResMed ASV (and other high end specialty machines) do NOT separate centrals and obstructive apneas so they all go in the UA basket.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Pugsy
Posts: 65196
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ASV and (another) High Altitude Question

Post by Pugsy » Mon Aug 07, 2017 6:17 pm

xxyzx wrote:mine corrects to 10000
maybe yours doesnt
Incorrect... if you have the machine you say you do.
Maybe you don't have the AirCurve 10 ASV you said you had...
Altitude Adjustment Automatic Up To 8500 Feet
From the spec sheet on the AirCurve 10 ASV. Don't believe me...look it up. I just did and found it easily.

I am not aware of any machine that states anything above 8500 ft in terms of auto adjusting for altitude.
Everything I have ever seen has been 8500 or lower (depends on the brand and model).

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
TASmart
Posts: 1071
Joined: Wed Jan 04, 2017 7:23 pm
Location: Eugene, OR

Re: ASV and (another) High Altitude Question

Post by TASmart » Mon Aug 07, 2017 6:55 pm

Pugsy wrote:
xxyzx wrote:mine corrects to 10000
maybe yours doesnt
Incorrect... if you have the machine you say you do.
Maybe you don't have the AirCurve 10 ASV you said you had...
Altitude Adjustment Automatic Up To 8500 Feet
From the spec sheet on the AirCurve 10 ASV. Don't believe me...look it up. I just did and found it easily.

I am not aware of any machine that states anything above 8500 ft in terms of auto adjusting for altitude.
Everything I have ever seen has been 8500 or lower (depends on the brand and model).
Pugsy's information is verified in the aircurve 10 ASV manual, available with a google search.
All posts reflect my own opinion based on my experience and reading.
Your mileage may vary
Past performance is no guarantee of future results
Consult with your own physician as people very

User avatar
Pugsy
Posts: 65196
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ASV and (another) High Altitude Question

Post by Pugsy » Mon Aug 07, 2017 7:31 pm

xxyzx wrote: the OP does not have an aircurve 10
Duh...I know that and his question was already answered.
I was responding to your incorrect statement that
xxyzx wrote: mine corrects to 10000
and since you have already stated you have the AirCurve 10 ASV I was merely correcting your error.

You have said repeatedly that no one ever corrects you (even put it in your signature line) because you are never wrong (cough, cough)...which is simply not true but you just choose to ignore any corrections made to your incorrect or incomplete posts...so I fixed that one for everyone to see. Front and center and easy to find. No abstract interpretation of reading something someone said about something far out in left field.
This one is nice and easy and clear to understand.

Have a nice day.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

klv329
Posts: 497
Joined: Mon Aug 31, 2015 6:43 pm
Location: Bradenton, Florida

Re: ASV and (another) High Altitude Question

Post by klv329 » Mon Aug 07, 2017 7:52 pm

Ha....cpap humor (he tried to get out of bed with FFM on) with a veiled sex joke (spouse says now that you have yours on maybe you can help get mine on) on this CBS sitcom just now.

At least it was a little funny, although another scene had him making a weird noise while sleeping on the couch (snoring or choking or an obstructive apnea) with the kids watching him before they woke him up.

Well, that's how I interpreted it.

Not to change the topic, though, carry on!

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Additional Comments: Resmed Aircurve 10 ASVAUTO Min Epap 10.4, Max Epap 11.6, PS 1.6-12.0

User avatar
WearyOne
Posts: 1798
Joined: Sat Mar 03, 2007 3:30 pm
Location: USA

Re: ASV and (another) High Altitude Question

Post by WearyOne » Mon Aug 07, 2017 7:59 pm

.
ResMed AirCurve 10 ASV Manual
http://www.resmed.com/us/dam/documents/ ... er_eng.pdf

Page 17:

Operating Altitude: Sea level to 8,500' (2,591 m); air pressure range 1013 hPa to
738 hPa

I'm assuming I have the correct manual here. Pugsy or TASmart, please correct me if I'm wrong.

EDIT
: I realize this is not related to the original question. I was just posting a link to a legitimate source to further support correct information due to incorrect information posted by xxyzx, and I wanted to make sure I had the right manual. Pretty sure I do.

_________________
Machine: DreamStation 2 Auto CPAP Advanced with Humidifier
Additional Comments: Oscar Software | APAP: 9-10
Last edited by WearyOne on Tue Aug 08, 2017 11:54 am, edited 3 times in total.

WickedLoki
Posts: 40
Joined: Sat Aug 20, 2016 2:27 pm

Re: ASV and (another) High Altitude Question

Post by WickedLoki » Tue Aug 08, 2017 11:42 am

But you also lied twice in the same thread and were definitively caught in your lie!

User avatar
WearyOne
Posts: 1798
Joined: Sat Mar 03, 2007 3:30 pm
Location: USA

Re: ASV and (another) High Altitude Question

Post by WearyOne » Tue Aug 08, 2017 11:47 am

Just to be clear...I realize the post I made with the link to a manual was not related to the OP's question. I was just correcting a post made by
xxyzx and providing proof within the thread where the error was made.

_________________
Machine: DreamStation 2 Auto CPAP Advanced with Humidifier
Additional Comments: Oscar Software | APAP: 9-10

User avatar
Uncle Flapp
Posts: 247
Joined: Thu Nov 14, 2013 2:24 pm
Location: Arizona

Re: ASV and (another) High Altitude Question

Post by Uncle Flapp » Mon Oct 02, 2017 11:53 am

Just got back from Peru. At higher altitudes (Cusco ~12,000ft), I did see a slight increase in my AHI - all HYPs but the machine kept OAs and CAs at bay. I went from an AHI average of 1.0 to 3.5.

- Flappy

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dreamstation ASV. UPPP in 2007; Untreated AHI 84