CMS60F Pulse Ox Meter Help

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Madalot
Posts: 4287
Joined: Sat Jan 30, 2010 7:47 am

CMS60F Pulse Ox Meter Help

Post by Madalot » Wed Aug 02, 2017 1:54 pm

My old 50F (wrist model) died and I bought this a few months ago. It comes with a different software from the 50F, SP02 Assistant, which I installed. I admit that I had a little trouble getting it to work right initially and upload the overnight data, but I DID get it to work, doing a full overnight after I got everything working.

Then I put it away for a few months.

Got it out last week, charged it fully and did an overnight. Now I'm having an issue that I just can't seem to resolve.

When I plug the monitor into with the cord into a USB port (and I've tried several ports) - the computer DONGS meaning it recognizes something is there, but it's not able to access any data (doesn't list it as a viable "F" drive as it should). I go through the entire thing through SP02 Assistant and select "Connect Device" and it finds it. Then I select "Device Stored Data" and it takes a few seconds, then comes back with a "Device Not Responding" error.

I know there IS data because you can review it on the device itself and I can see it for the entire night.

I admit I've tried everything I know, including resetting the monitor to factory defaults and uninstalling and reinstalling the software.

I'm stumped and any suggestions would be greatly appreciated. Thanks in advance.

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy EVO. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7

User avatar
Pugsy
Posts: 65373
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: CMS60F Pulse Ox Meter Help

Post by Pugsy » Wed Aug 02, 2017 2:00 pm

Does that model require a proprietary cable? Some of those models do and some don't. I don't remember which ones though.
Palerider should know though.
Are you for sure using the cable that came with it?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
WearyOne
Posts: 1799
Joined: Sat Mar 03, 2007 3:30 pm
Location: USA

Re: CMS60F Pulse Ox Meter Help

Post by WearyOne » Wed Aug 02, 2017 2:22 pm

It sounds like if you're getting a "dong" that you're plugging it in after the computer is up and running. Two things you could try if you haven't already: 1) unplug it and plug back in while the computer is on. 2) Turn the computer off completely, plug it in, and then turn the computer back on. I've had issues with peripheral equipment not being recognized or seemed to be recognized but wouldn't work where one of these two things has worked.

_________________
Machine: DreamStation 2 Auto CPAP Advanced with Humidifier
Mask: AirFit™ F40 System - L/STD
Additional Comments: Oscar Software | APAP: 11-13 | 2nd Mask: Simplus Medium

User avatar
Madalot
Posts: 4287
Joined: Sat Jan 30, 2010 7:47 am

Re: CMS60F Pulse Ox Meter Help

Post by Madalot » Wed Aug 02, 2017 2:24 pm

Pugsy wrote:Does that model require a proprietary cable? Some of those models do and some don't. I don't remember which ones though.
Palerider should know though.
Are you for sure using the cable that came with it?
It came with a cable and even though it looked the same as the 50F, I switched it to the new one to be sure.
WearyOne wrote:It sounds like if you're getting a "dong" that you're plugging it in after the computer is up and running. Two things you could try if you haven't already: 1) unplug it and plug back in while the computer is on. 2) Turn the computer off completely, plug it in, and then turn the computer back on. I've had issues with peripheral equipment not being recognized or seemed to be recognized but wouldn't work where one of these two things has worked.
I'll try some of that as soon as I can. I had just rebooted the computer that morning (shut it down for nasty storms the night before) but I'll try your suggestions.

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy EVO. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7

User avatar
Madalot
Posts: 4287
Joined: Sat Jan 30, 2010 7:47 am

Re: CMS60F Pulse Ox Meter Help

Post by Madalot » Wed Aug 02, 2017 2:26 pm

Just for more info - I am running Windows 7.

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy EVO. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7

klv329
Posts: 500
Joined: Mon Aug 31, 2015 6:43 pm
Location: Bradenton, Florida

Re: CMS60F Pulse Ox Meter Help

Post by klv329 » Wed Aug 02, 2017 2:59 pm

Bluetooth wireless turned off?

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Additional Comments: Resmed Aircurve 10 ASVAUTO Min Epap 10.4, Max Epap 11.6, PS 1.6-12.0

User avatar
palerider
Posts: 32306
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: CMS60F Pulse Ox Meter Help

Post by palerider » Wed Aug 02, 2017 3:36 pm

Pugsy wrote:Does that model require a proprietary cable? Some of those models do and some don't. I don't remember which ones though.
Palerider should know though.
Are you for sure using the cable that came with it?
The 50D+ and E models do, the F and I models don't.
I don't have any experience with the 60F models yet, but it can't hurt to use the cable that came with the device.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
wm_hess
Posts: 334
Joined: Sat Oct 16, 2010 2:39 am

Re: CMS60F Pulse Ox Meter Help

Post by wm_hess » Wed Aug 02, 2017 4:28 pm

Madalot wrote:My old 50F (wrist model) died and I bought this a few months ago. It comes with a different software from the 50F, SP02 Assistant, which I installed. I admit that I had a little trouble getting it to work right initially and upload the overnight data, but I DID get it to work, doing a full overnight after I got everything working.

Then I put it away for a few months.

Got it out last week, charged it fully and did an overnight. Now I'm having an issue that I just can't seem to resolve.

When I plug the monitor into with the cord into a USB port (and I've tried several ports) - the computer DONGS meaning it recognizes something is there, but it's not able to access any data (doesn't list it as a viable "F" drive as it should). I go through the entire thing through SP02 Assistant and select "Connect Device" and it finds it. Then I select "Device Stored Data" and it takes a few seconds, then comes back with a "Device Not Responding" error.

I know there IS data because you can review it on the device itself and I can see it for the entire night.

I admit I've tried everything I know, including resetting the monitor to factory defaults and uninstalling and reinstalling the software.

I'm stumped and any suggestions would be greatly appreciated. Thanks in advance.
Madalot

No guarantees, but look at this site. https://innovo-medical.com/products/innovo-cms-50f-plus. Under the tab "Download to PC Instructions" It has a video regarding downloading to SPC Assistant from a 50f. The left hand video is for non blue tooth transfer of data. Since you're getting the 'ding' it's registering something.

Start up the Assistant, then plug in the meter. Select the rs232 plug by clicking on it. Select your monitor. Go to /File/Device Stored Data.

Good luck!

Edited to add that the right hand video is for blue tooth transfer. I know this is not your monitor, but I'm guessing the software is the same. Could be wrong, could be right, no guaranty like I mentioned.

-Bill

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: CMS-50I Wrist Pulse Oximeter, SP02 Review, Sleepyhead
Last edited by wm_hess on Wed Aug 02, 2017 11:47 pm, edited 1 time in total.

User avatar
jsielke
Posts: 162
Joined: Fri Feb 24, 2017 11:21 am

Re: CMS60F Pulse Ox Meter Help

Post by jsielke » Wed Aug 02, 2017 7:49 pm

xxyzx wrote:
palerider wrote:
Pugsy wrote:Does that model require a proprietary cable? Some of those models do and some don't. I don't remember which ones though.
Palerider should know though.
Are you for sure using the cable that came with it?
The 50D+ and E models do, the F and I models don't.
I don't have any experience with the 60F models yet, but it can't hurt to use the cable that came with the device.
====

wrong again

the 50f+ has bluetooth
As usual, xxyzx has posted an erroneous answer. The 50F+ was not even mentioned in the question. The 60F is the hand held version. I had one but the rechargeable battery died after less than a month. That may be the problem. My 50f is still working fine. The 60f can use a generic micro USB cable. It worked fine until the battery died. Incidentally the 50f+ is made by Innovo, and differs in that it connects via Bluetooth. It does not work with Sleepyhead. I realize that was not the question but wanted to avoid confusion.

_________________
Mask
Additional Comments: Backup: Transcend Auto Travel
Old man, new machine. Better sleep![b]

User avatar
Madalot
Posts: 4287
Joined: Sat Jan 30, 2010 7:47 am

Re: CMS60F Pulse Ox Meter Help

Post by Madalot » Thu Aug 03, 2017 3:42 am

wm_hess wrote:No guarantees, but look at this site. https://innovo-medical.com/products/innovo-cms-50f-plus. Under the tab "Download to PC Instructions" It has a video regarding downloading to SPC Assistant from a 50f. The left hand video is for non blue tooth transfer of data. Since you're getting the 'ding' it's registering something.

Start up the Assistant, then plug in the meter. Select the rs232 plug by clicking on it. Select your monitor. Go to /File/Device Stored Data.

Good luck!

Edited to add that the right hand video is for blue tooth transfer. I know this is not your monitor, but I'm guessing the software is the same. Could be wrong, could be right, no guaranty like I mentioned.
Thanks for sharing this information. I'm going to try again, trying to follow the instructions exactly, but I suspect it won't matter. I get to the point where it's time to access the data and keep getting "Device Not Responding."

Also, this specific device does NOT have Bluetooth. The software lists it, but the device itself does not have that option.

And as I indicated, I am using the USB cable that came with it. I'm 100% positive because my 50F had a gray cable and this one is black. The computer "bongs" when I plug it in, the software finds the device, but it can't access the data. It's like it's not transmitting.

But I am not admitting defeat. Once I get a few personal things completed (quitting my job, documenting for new person, etc) I'll have time to really work on this.

I appreciate all the input and any more suggestions will be greatly appreciated.

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy EVO. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7

User avatar
jsielke
Posts: 162
Joined: Fri Feb 24, 2017 11:21 am

Re: CMS60F Pulse Ox Meter Help

Post by jsielke » Thu Aug 03, 2017 6:47 am

One last suggestion. Make sure you have the correct drivers. CP210x_VCP_Windows32, or CP210x_VCP_Windows64.

_________________
Mask
Additional Comments: Backup: Transcend Auto Travel
Old man, new machine. Better sleep![b]

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: CMS60F Pulse Ox Meter Help

Post by raisedfist » Thu Aug 03, 2017 9:06 am

Did you try to uninstall everything and start from scratch?

_________________
Mask
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

User avatar
Madalot
Posts: 4287
Joined: Sat Jan 30, 2010 7:47 am

Re: CMS60F Pulse Ox Meter Help

Post by Madalot » Thu Aug 03, 2017 9:12 am

raisedfist wrote:Did you try to uninstall everything and start from scratch?
Yes and No. I did it, but not in the most logical order. I have a tendency to just keep trying things as I think of them, which sometimes works, sometimes does not. LOL

That is another step to take for sure. Thanks!

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy EVO. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7

User avatar
jsielke
Posts: 162
Joined: Fri Feb 24, 2017 11:21 am

Re: CMS60F Pulse Ox Meter Help

Post by jsielke » Thu Aug 03, 2017 9:25 am

I had a problem at first with mine. I used the software on the CD they provide, and thought the drivers it installed were OK, but was not the case. Someone on another forum recommended I get the drivers directly from the site. Try https://www.silabs.com/products/develop ... cp-drivers

_________________
Mask
Additional Comments: Backup: Transcend Auto Travel
Old man, new machine. Better sleep![b]

User avatar
Madalot
Posts: 4287
Joined: Sat Jan 30, 2010 7:47 am

Re: CMS60F Pulse Ox Meter Help

Post by Madalot » Thu Aug 03, 2017 9:29 am

jsielke wrote:I had a problem at first with mine. I used the software on the CD they provide, and thought the drivers it installed were OK, but was not the case. Someone on another forum recommended I get the drivers directly from the site. Try https://www.silabs.com/products/develop ... cp-drivers
Excellent! Thank you!

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy EVO. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7