AirSense 10 Elite: Nasal vs Pillows??

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
EdNerd
Posts: 127
Joined: Sat Feb 04, 2017 6:33 pm

AirSense 10 Elite: Nasal vs Pillows??

Post by EdNerd » Mon Jun 26, 2017 12:34 pm

The ResMed AirSense 10 Elite has Mask settings for Full Face / Nasal / Pillows.
Just curious - what's the difference in these settings??

The only thing I could find was in the clinical guide:
If you're using pillows and have a pressure less than 7, the AutoStart might not work.

Is that all??
Ed

_________________
Machine: AirSense™ 10 Elite CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: 11cmH2O; Pad-A-Cheek cover
Very helpful stuff:
"Taming the Mirage Quattro" http://tinyurl.com/2ft3lh8

User avatar
Pugsy
Posts: 65402
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AirSense 10 Elite: Nasal vs Pillows??

Post by Pugsy » Mon Jun 26, 2017 1:35 pm

Call them up and ask them.
I would do it but I am going out the door right now and don't have time.


For Customer Assistance including Technical and Patient Support:

1 (800) 424-0737 (toll-free) in United States and Canada

Monday to Friday, 5:30am - 5pm PST.

Years ago the mask setting choice was used in leak calculation to determine excess leak but I can't find anything like that now so I don't know if it has changed or not.
It may have something to do with the resistance in each mask but I can't find anything that addresses the need for mask type selection on anything remotely current.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: AirSense 10 Elite: Nasal vs Pillows??

Post by chunkyfrog » Mon Jun 26, 2017 2:08 pm

I usually set the machine to match the mask.
Outside of the autostart, I cannot tell the difference.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AirSense 10 Elite: Nasal vs Pillows??

Post by palerider » Mon Jun 26, 2017 7:15 pm

EdNerd wrote:The ResMed AirSense 10 Elite has Mask settings for Full Face / Nasal / Pillows.
Just curious - what's the difference in these settings??

The only thing I could find was in the clinical guide:
If you're using pillows and have a pressure less than 7, the AutoStart might not work.

Is that all??
Ed
Because the sensor is within the device and not in the mask, the device must calculate the predicted pressure at the mask based on the flow measurements at a different point in the system. The pressure at two points varies with the flow. For turbulent flow, this can be calculated by Flow (q) = constant√(p1 − p2).3 The length and diameter of the tubing and mask characteristics affect the constant value, and so the technician must input the mask and tube type for the microprocessor to make the correct calculations.
https://www.ncbi.nlm.nih.gov/pmc/articles/PMC4629962/ (I haven't gotten around to reading the referenced patents to see if there's more info)

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
RogerSC
Posts: 1919
Joined: Sat Jul 21, 2012 1:11 pm

Re: AirSense 10 Elite: Nasal vs Pillows??

Post by RogerSC » Mon Jun 26, 2017 8:21 pm

Since ResMed cpaps are designed to normalize leak data for mask venting, the vent rate of various style masks needs to be taken into account in your sleep data. I don't have the reference at hand, but I've seen a table of the characteristic vent rates of various kinds of ResMed masks, and they are different. So that's at least part of the story. Don't know what else they're doing based on the mask type, though.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Additional Comments: ResMed AirFit N30 mask

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AirSense 10 Elite: Nasal vs Pillows??

Post by palerider » Tue Jun 27, 2017 12:39 am

RogerSC wrote:Since ResMed cpaps are designed to normalize leak data for mask venting, the vent rate of various style masks needs to be taken into account in your sleep data. I don't have the reference at hand, but I've seen a table of the characteristic vent rates of various kinds of ResMed masks, and they are different. So that's at least part of the story. Don't know what else they're doing based on the mask type, though.
they're more similar than they are different, and there are larger variations in single classes of masks than there are between classes.

it's not the vent rate, it's the resistance, which better matches between full face, nasal, and pillow.