Use Any Mask with your AirMini

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
C0d3mnky
Posts: 2
Joined: Wed Sep 05, 2018 10:00 am

Re: Use Any Mask with your AirMini

Post by C0d3mnky » Wed Sep 05, 2018 10:25 am

I started on this project because I didn't like the noise coming from the humidifier, and I don't use a humidifier. I wanted to connect my mask directly to the hose. While post #1 does work, I didn't like the extra hose in the middle. I ordered the same pieces, removed one of the rubber connectors from the 18" humidifier hose, cut the end off of my resmed hose, and used some epoxy to connect the connector to the resmed hose.
20180905_095402.jpg
18" Humidifier Coupler Hose
20180905_095402.jpg (132.3 KiB) Viewed 24031 times
20180905_095521.jpg
Removing the rubber end from the coupler hose
20180905_095521.jpg (160.8 KiB) Viewed 24031 times
20180905_095552.jpg
Attached coupler to resmed hose with epoxy
20180905_095552.jpg (112.63 KiB) Viewed 24031 times

Willy4401
Posts: 2
Joined: Wed Aug 29, 2018 12:19 pm

Re: Use Any Mask with your AirMini

Post by Willy4401 » Wed Sep 12, 2018 3:58 pm

Thank you for the suggestion and information. Fortunately, our local library has 3D printing capability and they created the adapter for me at no cost. I can now connect the hose and mask I use with my AirSense 10 without having the humidifier in the line. It is noisier from wind noise than the AirSense 10 which is virtually silent. The noise seems to be both at the mask and at the connection to the AirMini. I will order the Qtube to see if that helps as well.

Thank you again for taking the time to reply.

eville
Posts: 11
Joined: Fri Apr 20, 2018 2:47 pm
Location: NorCal

Re: Use Any Mask with your AirMini

Post by eville » Tue Oct 02, 2018 5:03 pm

[/quote]

My Qtube muffler arrived today and I'm happy to report that it solved the noise problem with the mask, at least for me. I'm not sure how to give an objective number in terms of the sound pressure levels as I don't know how to measure it. However, from a personal observation, I will say that the Airmini with the Qtube setup is now as quiet as my AirSense 10 and way better than my laminar airflow adapter. Further, I don't notice any decrease in the pressure level as I'm not starving for air.

I'll start a new thread later today on my setup.
[/quote]

Do you have any pictures of how you connected the qtube to your airmini?
TIA.
Steve

cpapjustin
Posts: 2
Joined: Mon Oct 08, 2018 4:29 pm

Re: Use Any Mask with your AirMini

Post by cpapjustin » Mon Oct 08, 2018 4:34 pm

digitaldog888 wrote:
Tue Sep 04, 2018 3:32 pm
Willy4401 wrote:
Tue Sep 04, 2018 2:40 pm
I recently started using CPAP a few months ago. I currently use a Resmed Air Sense 10 at home with a Resmed N20 nasal mask. When I started, I was using a Resmed AirFit F20 full face mask but didn’t like the fit. I much prefer the nasal mask.

I purchased the AirMini for travel and like the machine. Initially I purchased the F20 setup pack which allows me to connect the F20 full face mask, but as I said I don’t like the fit. I then purchased the setup pack for the N20 mask and while it works, the noise from the inline humidifier is too much. I don’t get any noise when using the F20 full face mask.

I have been trying to find a way to connect either the Resmed N20 nasal mask, or Resmed Mirage FX nasal mask to the airmini without an inline humidifier. As I have learned, Resmed does not make a hose for the Airmini that will work. I am a novice at this, but is anyone able to tell me how I might do this?

Any help will be appreciated.
I had exactly the same problem. The OP in post #1 of this thread had a solution that hacked together a tubing that would work with any mask including the N20. My original solution was to by-pass the humidX housing using a custom 3D printed connector as noted on page 2 of this thread. Unfortunately, it did not solve the noise problem. The only solution that worked for me was by using a Qtube muffler. More information could be found here:

viewtopic.php?f=1&t=172660&p=1264551

My solution requires access to a 3D printer. If you don't have access to a 3D printer, there are a number of 3D printing shops that could print the parts for you at a price.

I hope this helps.
hi, thanks for this 3d model. however, if you bypass the humidx housing, are you also bypassing the vents to expel cO2? which is deadly!

Stom
Posts: 383
Joined: Sat Jul 21, 2018 9:44 am

Re: Use Any Mask with your AirMini

Post by Stom » Mon Oct 08, 2018 4:55 pm

cpapjustin wrote:
Mon Oct 08, 2018 4:34 pm

hi, thanks for this 3d model. however, if you bypass the humidx housing, are you also bypassing the vents to expel cO2? which is deadly!
All regular CPAP masks have passive vents built into them. The AirMini humidity retention system needs the vent to be on the supply side of the humidity disc so humidity doesn't vent and defeat the purpose, thus the AirMini uses special venting on modified P10 and N20 masks. Attaching a normal, vented CPAP mask without a humidity retention system with this hose adapter should be fine. Or, put another way, you cannot use the humidity retention system if you use a regular mask with this adapter.
Official Foe™ on @Palerider's public list of foes. ;-)

Rules to live by: Be wary of anyone boasting their behavior is like that of a fictional sociopath.

cpapjustin
Posts: 2
Joined: Mon Oct 08, 2018 4:29 pm

Re: Use Any Mask with your AirMini

Post by cpapjustin » Mon Oct 08, 2018 6:09 pm

Stom wrote:
Mon Oct 08, 2018 4:55 pm
cpapjustin wrote:
Mon Oct 08, 2018 4:34 pm

hi, thanks for this 3d model. however, if you bypass the humidx housing, are you also bypassing the vents to expel cO2? which is deadly!
All regular CPAP masks have passive vents built into them. The AirMini humidity retention system needs the vent to be on the supply side of the humidity disc so humidity doesn't vent and defeat the purpose of the humidity retention system, thus the AirMini uses special venting on modified P10 and N20 masks as part of the humidification. Attaching a normal, vented CPAP mask without a humidity retention system with this hose adapter should be fine. That is, you cannot use the humidity retention system if you use a regular mask with this adapter.
hi storm, thanks for your reply. i failed to mention that i'm using n20 nasal mask. which the vent is on the lower end of the tube. https://www.thecpapclinic.com.au/files/ ... ac_eng.pdf , Image C. Item 8 and 9.
i misunderstood the comment. i thought he's still using the connector kit that came with the airmini. where in fact he's using the elbow and tube that came with the n20 mask with the vent at the top (that blows air like mad).

austinyue
Posts: 2
Joined: Wed Oct 10, 2018 8:22 pm
Location: CHINA

Re: Use Any Mask with your AirMini

Post by austinyue » Wed Oct 10, 2018 8:38 pm

Hello,i get one AirMini recently,but I cannot use N20 M size ok,I also want to use other brand nasal mask.I get one airmini adaptor for common tube,however there have large noise so that i cannot use well.

How about the noise since use this commen tube and own mask ?

Thanks!
2.jpg
2.jpg (92.05 KiB) Viewed 23791 times
TangledHose wrote:
Wed Jun 21, 2017 10:12 am
I have a special need for a small ultra packable travel machine primarily for my extensive motorcycle touring trips where space is extremely limited. I started several years ago with a Transcend which frankly worked just fine, however, I got intrigued with the HDM Z1 and began to use it instead because it was somewhat smaller and a better packing shape than the Transcend. On a recent trip I managed to kill my Z1 (through my own negligence) so when I went to replace my busted Z1 I decided to give the AirMini a try instead.
The AirMini is a great little machine although all these ultra-small travel size machines do suffer from the high turbine speed Darth Vadar annoying breathing noise to varying degrees........I would rate the AirMini as suffering the effect the least, the Transcend almost tied with the AirMini maybe just slightly noisier, and the Z1 a very, very distant third (the Z1 is by far the noisiest of the three). I was really rather pleasantly surprised with the AirMini both in how smoothly it breathed, and how low it scored on my own personal breathing noise test/scale.

One major concern with the AirMini is ResMed's decision to attempt to make the AirMini proprietary in terms of the mask that can be used with the machine (huge, huge marketing mistake on ResMed's part!!!). However, where there is a will, there is a way -- I ordered the machine with the F20 full face mask option/bundle and tried the F20 which was acceptable, but not my personal first pick in full face masks in terms of comfort (personally I much prefer the F&P Simplus mask -- much more comfortable for me personally). With the F20 bundle you don't get the HME humidifier option (I personally don't need it for a travel machine anyway) but that leaves the door wide open to simply modify the F20 supplied hose so that you can then use any CPAP mask that you want. This also gives you the ability to use a third party HME cartridge if you feel the need such as the HDM HME cartridge, or a third party one that I think cpap.com carries if you search for it. But again, I really don't personally have a need for the not so effective HME option.

So the mask end of the F20 AirMini hose has a small swivel connector on it -- I have several hoses from other machines and one short hose piece that I think came with an F&P add-on external humidifier, but at any rate all you need is a short hose piece or coupler with the rubber/silicone attachment flange on the end. The silicone flange fits right over the swivel connector on the AirMini F20 hose end nice and snug, and then just use a liberal amount of clear silicone seal to glue and seal the connection. The connection is nice and snug and will not come loose as you toss and turn at night, I have used my modified hose successfully several nights with no problem whatsoever.........pictures below:

My modification adds a little more than an additional foot the ResMed hose with the short addition I used, but you may be able to find a coupler that is shorter than this one.......I will look when I get a chance and see if I can find a shorter option, but this short F&P hose splice works just fine with the pliable silicone end flanges.

Edit: I found a page on cpap.com with short hose coupler options that would work, just pick one with the rubber ends that can fit over the AirMini swivel coupler: https://www.cpap.com/cpap-hose/cpap-hose-short.php

Image

Image

Image
I am a OSA patients,also i have a store for cpap .

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: Use Any Mask with your AirMini

Post by prodigyplace » Thu Oct 11, 2018 9:56 am


_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.

shlinder
Posts: 14
Joined: Wed Feb 20, 2019 5:46 am

Re: Use Any Mask with your AirMini

Post by shlinder » Wed Feb 20, 2019 5:51 am

The question is, though, once you jimmy a new adapter to work with any hose, you lose the ability to have that HumidX, correct? I am looking for an adapter that INCLUDES the HumidX, meaning an adapter at the mask end. Is there anything like that?

I want to use it with my Dreamwear Nasal mask.

User avatar
billbolton
Posts: 2264
Joined: Wed Jun 07, 2006 7:46 pm
Location: Sydney, Australia

Re: Use Any Mask with your AirMini

Post by billbolton » Thu Feb 21, 2019 1:07 am

shlinder wrote:
Wed Feb 20, 2019 5:51 am
I am looking for an adapter that INCLUDES the HumidX, meaning an adapter at the mask end. Is there anything like that?

I want to use it with my Dreamwear Nasal mask.
The HumidX approach relies on none of the water vapour in the exhaled being lost. If you connect a standard mask of any type to the output end of a HumidX (by any method) it will have C02 scavenging exhaust vents, so spme of the water vapour in the exhaled breath will be lost.

The HumidX humidification approach relies on having specific masks which do not have C02 scavenging exhaust vents, with C02 exhaust scavenging being done at the HumidX capsule holder, after water vapour was been captured.

Basically, if you want to connect any standard mask to a AirMini, you wont be able to use HumidX humidifcation. :idea:

_________________
MachineMask
Additional Comments: Airmini, Medistrom Pilot 24, CMS 60C Pulse Oximeter, ResScan 6

digitaldog888
Posts: 22
Joined: Sun Jul 22, 2018 3:56 pm

Re: Use Any Mask with your AirMini

Post by digitaldog888 » Thu Feb 21, 2019 11:31 am

billbolton wrote:
Thu Feb 21, 2019 1:07 am
shlinder wrote:
Wed Feb 20, 2019 5:51 am
I am looking for an adapter that INCLUDES the HumidX, meaning an adapter at the mask end. Is there anything like that?

I want to use it with my Dreamwear Nasal mask.
The HumidX approach relies on none of the water vapour in the exhaled being lost. If you connect a standard mask of any type to the output end of a HumidX (by any method) it will have C02 scavenging exhaust vents, so spme of the water vapour in the exhaled breath will be lost.

The HumidX humidification approach relies on having specific masks which do not have C02 scavenging exhaust vents, with C02 exhaust scavenging being done at the HumidX capsule holder, after water vapour was been captured.

Basically, if you want to connect any standard mask to a AirMini, you wont be able to use HumidX humidifcation. :idea:
The approach I took was to use the Z1 HME for humidification:

viewtopic/t172660/Reduce-noise-on-Airmi ... 5#p1290520

shlinder
Posts: 14
Joined: Wed Feb 20, 2019 5:46 am

Re: Use Any Mask with your AirMini

Post by shlinder » Thu Feb 21, 2019 5:08 pm

digitaldog888 wrote:
Thu Feb 21, 2019 11:31 am
billbolton wrote:
Thu Feb 21, 2019 1:07 am
shlinder wrote:
Wed Feb 20, 2019 5:51 am
I am looking for an adapter that INCLUDES the HumidX, meaning an adapter at the mask end. Is there anything like that?

I want to use it with my Dreamwear Nasal mask.
Basically, if you want to connect any standard mask to a AirMini, you wont be able to use HumidX humidifcation. :idea:
The approach I took was to use the Z1 HME for humidification:

viewtopic/t172660/Reduce-noise-on-Airmi ... 5#p1290520
Wow! Amazing responses that completely clarify the situation. Thanks.

As a counter, is it true that the big complaint about the AirMini being loud is BECAUSE of the HumidX component?

What would you all think of the AirMini connected to one of those small adaptors from Amazon into the Q muffler attached to a TravelHose into the Z1 HME and a DreamWear nasal mask? For daily use in addition to travel!

I was just about to buy the AirSense 10 to use with my DreamWear nasal mask, but if that setup works... well...

digitaldog888
Posts: 22
Joined: Sun Jul 22, 2018 3:56 pm

Re: Use Any Mask with your AirMini

Post by digitaldog888 » Thu Feb 21, 2019 5:29 pm

shlinder wrote:
Thu Feb 21, 2019 5:08 pm
digitaldog888 wrote:
Thu Feb 21, 2019 11:31 am
billbolton wrote:
Thu Feb 21, 2019 1:07 am
shlinder wrote:
Wed Feb 20, 2019 5:51 am
I am looking for an adapter that INCLUDES the HumidX, meaning an adapter at the mask end. Is there anything like that?

I want to use it with my Dreamwear Nasal mask.
Basically, if you want to connect any standard mask to a AirMini, you wont be able to use HumidX humidifcation. :idea:
The approach I took was to use the Z1 HME for humidification:

viewtopic/t172660/Reduce-noise-on-Airmi ... 5#p1290520
Wow! Amazing responses that completely clarify the situation. Thanks.

As a counter, is it true that the big complaint about the AirMini being loud is BECAUSE of the HumidX component?

What would you all think of the ResMini connected to one of those small adaptors from Amazon attached to a TravelHose into the Z1 HME and a DreamWear nasal mask?

I was just about to buy the AirSense 10 to use with my DreamWear nasal mask, but if that setup works... well...
I can't say for sure if the HumidX is the cause of the noise but what I can say is that it makes no difference in the noise level if it was removed. I designed an adaptor that bypasses the HumidX and it made no difference. Others on this forum have suggested that the nature of how the Airmini is designed is what is causing the unacceptable noise level. I have to agree as it's difficult to design such a small platform with the same ability as the larger units to "smooth" out the air flow. I believe it's the turbulent airflow of the Airmini that is causing the noise. Hence, the QTube muffler helps to dampen the airflow somewhat.

The setup I've been using with the custom adapter, standard tubing, Z1 HME, and N20 mask has worked reasonably well. I've used it a number of times for travel and have had no issues so far. However, I will say that it's still not as quiet as my AirSense 10 that I used permanently at home.

shlinder
Posts: 14
Joined: Wed Feb 20, 2019 5:46 am

Re: Use Any Mask with your AirMini

Post by shlinder » Sun Feb 24, 2019 7:22 am

With this kind of setup, does anyone use the AirMini for normal, daily home use? If it's acceptable for travel, why not for every day?

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: Use Any Mask with your AirMini

Post by prodigyplace » Sun Feb 24, 2019 9:46 am

shlinder wrote:
Sun Feb 24, 2019 7:22 am
With this kind of setup, does anyone use the AirMini for normal, daily home use? If it's acceptable for travel, why not for every day?
A couple of ideas come to mind.

If somebody needs to prove CPAP compliance, the AirMini does not provide that data, AFAIK.

The AirMini may not be tested / rated for continuous use and therefore may fail sooner than the expected 5 year life of a home CPAP machine.
This would me much like most all electrical seasonal decorations. Most of them are only rated for use 3 months maximum out of the year,