hose is rattling

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Grandma6

hose is rattling

Post by Grandma6 » Mon Jun 12, 2017 11:09 am

I am sleeping with my nasal mask on and the hose is rattling, very disturbing to my sleep!
I unhook the mask and water pours out of the hose...what can I do to stop this??

User avatar
Hopeful50
Posts: 177
Joined: Fri Feb 12, 2016 3:50 pm
Location: Cupertino, CA

Re: hose is rattling

Post by Hopeful50 » Mon Jun 12, 2017 11:19 am

This is happening because humidified air is condensing in the hose, and the solution probably has multiple parts:
- make sure your machine is lower than your head, so any condensation runs back to the humidifier chamber instead of towards you.
- turn the heat on your hose up higher, so the air is warm enough to keep the water from condensing
- turn your humidity level down

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Pressure 13-20, EPR 3

User avatar
Pugsy
Posts: 65411
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: hose is rattling

Post by Pugsy » Mon Jun 12, 2017 12:44 pm

We call it "rain out"...unwanted condensation in the hose or mask. Won't hurt you but highly annoying for sure.

You have several options available to try to fix it. See this thread posts number 2 and 3 for explanation and various options.
viewtopic/t94035/Pugsys-Pointers-3Deali ... -road.html

Easiest I think...get a heated hose if your machine supports a heated hose. I use one year round...warmer in the winter and lower temp in the summer and humidity delivered stays constant (keeps the nasal mucosa happy) year round. It's just easier than trying to fiddle with the humidity levels or hose placement or whatever.

If you don't want to get a heated hose then some sort of hose cozy will often be sufficient. All this explained in the above thread I linked to.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

D.H.
Posts: 3535
Joined: Sun Sep 04, 2016 7:07 pm

Re: hose is rattling

Post by D.H. » Mon Jun 12, 2017 3:13 pm

In addition to what was already suggested, be very careful not to overfill the reservoir.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: hose is rattling

Post by Goofproof » Mon Jun 12, 2017 9:05 pm

D.H. wrote:In addition to what was already suggested, be very careful not to overfill the reservoir.
Especally until it runs out the mask!
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
Petrodon
Posts: 36
Joined: Tue Jun 24, 2014 8:23 am

Re: hose is rattling

Post by Petrodon » Mon Jun 12, 2017 9:26 pm

For Rainout, if the unit you have has no heated hose option, try a hose cover like SnuggleHose.