Remstar Auto 1 Hour Delay Encore Pro

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Malibu
Posts: 136
Joined: Mon Feb 27, 2006 7:15 pm
Location: Illinois

Remstar Auto 1 Hour Delay Encore Pro

Post by Malibu » Sat Oct 21, 2006 12:49 pm

Hello all...

Question...My wife just picked up her Remstar Auto (NOT the M series) because she wanted a proven product. I noticed it had version 2.5 software on it versus the 2.6 and 2.7 on my main unit and my backup unit.

We downloaded the data from last night into Encore Pro 1.6i and the first hour was blank and then it started recording.

I know in version 2.5 and 2.6 Remstar Auto software if you shut it off or just starting out the evening you lose 10 minutes of data and then it starts recording again.

Does anyone out there have the 1 hour delay in recording with version 2.5??

Thanks,

Marc


User avatar
Snoredog
Posts: 6399
Joined: Sun Mar 19, 2006 5:09 pm

Post by Snoredog » Sat Oct 21, 2006 1:25 pm

does your Auto have a blue CFlex sticker on top? I purchased my Remstar Auto w/cflex back in June 2006, it came with firmware version 2.6. I got it from cpap.com and they move a lot a machines so I doubt very much it sat on any shelf for very long.

The SmartCard is like a floppy, you format it so you can write data to it.

When you get a new machine with the SmartCard and you have the software, use the EncorePro to erase and setup the card, then you won't have a problem with missing data.

While it is true if you stop/start the machines it will leave a blank in the report, but I just went through numerous reports of past where I stopped and restarted the machine, I am guesstimating the actual time because there is no reference but 1 hour, it shows my "breaks" if any are the time it was off and that is it, if it was 40 minutes I get slightly over the 1/2hr mark.

So I don't think there is ANY set "timeframe" if you stop/start or at least I'm not seeing it in any of the dozen PDF reports I've gone through.


CollegeGirl
Posts: 1038
Joined: Thu Oct 20, 2005 6:49 pm
Location: VA

Post by CollegeGirl » Sat Oct 21, 2006 1:35 pm

Check the end of the data - the "clock" on my EncorePro is an hour slow when compared to my time - so if I go to bed at 8 pm it says 7 pm, etc. That would account for your missing hour, as long as there's an "extra" hour of data at the end.
Machine: M-Series Auto
Mask: Headrest
No humidifier
On the hose since 2005.

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Post by Wulfman » Sat Oct 21, 2006 2:00 pm

Malibu wrote:I know in version 2.5 and 2.6 Remstar Auto software if you shut it off or just starting out the evening you lose 10 minutes of data and then it starts recording again.
Marc,

I'm curious as to where you have confirmed this information.....
I don't recall this information being posted on this forum.

I purchased my original Pro 2 in April/May of '05. Other than losing somewhere between one to two minutes per month on the internal time clock, it recorded data from the very moment of it being started.

In the early part of this year, I purchased a "previously owned" Auto which was of the same vintage and same characteristics.....internal clock losing time, but recorded data immediately after starting.

In August, I purchased another brand new Auto. This one exhibited the characteristics you described......starts for a minute or two and then stops recording data for EXACTLY 10 minutes.
I reported this and got RMAs for the Auto and Pro 2 (to get a new battery and time reset and to correct a slight noise as the C-Flex changed from inhale to exhale).
When I got both of these machines back, the Pro 2 ALSO exhibited the "missing 10 minutes" syndrome as the Auto previously did and STILL did.

So, now the only machine that begins recording data from the beginning of the night is my "previously owned" Auto.

If anybody knows anything about this "feature" (we're going to drop the first 10 minutes out of every night) please let me know.
If others have experienced this characteristic, please speak up.

Thanks,

Den

(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

CollegeGirl
Posts: 1038
Joined: Thu Oct 20, 2005 6:49 pm
Location: VA

Post by CollegeGirl » Sat Oct 21, 2006 2:05 pm

It's just settling time, Den. Do a search for "settling." The machine gives you ten minutes to fall asleep before it starts recording data, because recording data while you're awake can cause false hypops/apneas to be recorded and skew your data. HTH.

--CG.
Machine: M-Series Auto
Mask: Headrest
No humidifier
On the hose since 2005.

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Post by Wulfman » Sat Oct 21, 2006 2:11 pm

CG,

ResMad has a setting to turn on and off the "settling" on their machines, but I've never seen this mentioned with regard to the Respironics machines. AND, when would they have started doing this? The machines from 2005 didn't do this.....

Den

(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

CollegeGirl
Posts: 1038
Joined: Thu Oct 20, 2005 6:49 pm
Location: VA

Post by CollegeGirl » Sat Oct 21, 2006 2:16 pm

Den, Respironics autos have settling, you just can't adjust it (afaik). My machine is one of the pre-M Series autos, and it does indeed do this. I never get data until ten minutes past when I've gone to bed.

--CG

Machine: M-Series Auto
Mask: Headrest
No humidifier
On the hose since 2005.

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Post by Wulfman » Sat Oct 21, 2006 2:29 pm

And it actually "drops" the 10 minutes out of the total time?
In your Encore Pro reports, does it start recording for a minute or two at the very beginning and then leaves a blank space (10 minutes) and then starts again?

Do the "M Series" machines do this too.......Snoredog?

Why would not the Respironics techies have relayed this message to me (through CPAP.COM)?

Den

(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

CollegeGirl
Posts: 1038
Joined: Thu Oct 20, 2005 6:49 pm
Location: VA

Post by CollegeGirl » Sat Oct 21, 2006 2:33 pm

No, there's no minute or two at the beginning. There's just nothing for ten minutes.

I went to bed at 12:15 last night. There's no data at all until exactly 12:25. It's that way every night.

--CG
Machine: M-Series Auto
Mask: Headrest
No humidifier
On the hose since 2005.

Malibu
Posts: 136
Joined: Mon Feb 27, 2006 7:15 pm
Location: Illinois

Post by Malibu » Sat Oct 21, 2006 2:37 pm

Den,

I just went back and looked at all my reports from Encore Pro, both 1.5i and 1.6i and the first 10 minutes is not recorded. If I took a bathroom break in the middle of the night, I get 10 minutes of no data then it starts recording again. I have 2 Remstar Autos, from March of this year, software version 2.6 and from August of this year, software version 2.7.

I will try to remember where I read this but at the moment it escapes me.

Do you have a version 2.5 software machine by chance? I was wondering if the older Remstars had an hour of settling and they changed it to 10 minutes in the newer software...

Later,

Marc


CollegeGirl
Posts: 1038
Joined: Thu Oct 20, 2005 6:49 pm
Location: VA

Post by CollegeGirl » Sat Oct 21, 2006 2:38 pm

Marc - I'm curious - did you check for that extra hour at the end? I've definitely never heard of an hour of settling. That wouldn't make sense.
Machine: M-Series Auto
Mask: Headrest
No humidifier
On the hose since 2005.

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Post by Wulfman » Sat Oct 21, 2006 2:40 pm

Well, on mine, the "start time" is recorded at the top of the Encore Pro report for each night (which is correct) and there is a little colored "blip" at the beginning of each chart for a minute or two and then the blank spot and then the solid line when it starts recording again, but when I do the math, there's 10 minutes missing out of the total time. If I have to get up during the night, it does it again.
On a couple of occasions, it DIDN'T drop the 10 minutes (once at the beginning and once during the night.....these were on different nights).

Den

(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

Malibu
Posts: 136
Joined: Mon Feb 27, 2006 7:15 pm
Location: Illinois

Post by Malibu » Sat Oct 21, 2006 2:47 pm

College Girl...

The interesting part is the Encore Pro graph says the machine was on from 10 pm until 5 am which is 7 hours...

The machine started at 0 hours and now has 6.1 hours on it...

It not only didn't record the first hour, the hour meter on the machine is missing an hour...hummm

Marc


User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Post by Wulfman » Sat Oct 21, 2006 2:52 pm

I'll be darned if I can find where it displays the firmware level on ANY of my machines......

Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

Malibu
Posts: 136
Joined: Mon Feb 27, 2006 7:15 pm
Location: Illinois

Post by Malibu » Sat Oct 21, 2006 3:05 pm

Den...

Unplug and then plug it back in...

You will see a number like 2.5 flash on the screen for about 3 seconds then go to the regular screen you're used to seeing...
Marc