I've been trying to improve the results of my therapy, and feel I would like to try bumping my pressure up from 10 to 11 for a couple of nights this weekend to see if I might feel better. I have the Remstar Pro 2 w/C-flex. I am running straight cpap mode with no ramp. I am going page by page and am wondering if anyone would please be able to tell me how to change this setting. Thanks in advance for ANY help!

Dawn


User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Post by Wulfman » Fri Sep 08, 2006 7:19 pm

Dawn,

You might want to copy and paste these instructions into a document on your computer.

Den


REMstar Pro 2 Setup

When in the Setup Menu, the humidifier ^ and ramp v buttons operate as up and down keys to change the settings, the left/right user buttons < > allow you to go to the previous/next question or setting, and the pressure start/stop button is used to exit the Setup Menu. Holding the humidifier or ramp buttons down will cause the values to change more quickly.

To enter the Therapy Setup Menu, hold the two top user buttons < > down while plugging in the power cord.
Continue holding the buttons down until the REMstar Pro 2 with C-Flex beeps twice.

Note: The word "setup" will appear on all of the screens indicating that you are in the Therapy Setup Menu.
If you press the Pressure start/stop button, you will exit the Setup Menu.

a. Compliance hours/nights: (recommend leaving alone, but CAN be cleared at this point)
Select next setting with >

b. Therapy mode: (CPAP or CFLE) select with ^ or v
Select next setting with >

c. CPAP or CFLE pressure setting: Select with ^ or v
Select next setting with >

d. C-Flex mode (if you chose CFLE mode): Select setting 1, 2 or 3 with ^
Select next setting with >

e: Ramp time setting: Select with ^ or v
(ramp time will be turned off with a setting of 0)
Select next setting with >

f. Ramp pressure setting: Select with ^ or v
Select next setting with >

g. Patient disconnect setting: 1 = on 0 = off Select with ^ or v
Select next setting with >

h. Buttons lights setting: 1 = on 0 = off Select with ^ or v

(Last setting. Use On/Off button to exit Setup)

(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.