How does one setup the REMstar Pro 2 with C-Flex? I want to be able to see the settings to make sure they are correct.
Setup REMstar Pro 2 with C-Flex
Please copy and paste the following into a document on your computer.
REMstar Pro 2 Setup
When in the Setup Menu, the humidifier ^ and ramp v buttons operate as up and down keys to change the settings, the left/right user buttons < > allow you to go to the previous/next question or setting, and the pressure start/stop button is used to exit the Setup Menu. Holding the humidifier or ramp buttons down will cause the values to change more quickly.
To enter the Therapy Setup Menu, hold the two top user buttons < > down while plugging in the power cord.
Continue holding the buttons down until the REMstar Pro 2 with C-Flex beeps twice.
Note: The word "setup" will appear on all of the screens indicating that you are in the Therapy Setup Menu.
If you press the Pressure start/stop button, you will exit the Setup Menu.
a. Compliance hours/nights: (recommend leaving alone, but CAN be cleared at this point)
Select next setting with >
b. Therapy mode: (CPAP or CFLE) select with ^ or v
Select next setting with >
c. CPAP or CFLE pressure setting: Select with ^ or v
Select next setting with >
d. C-Flex mode (if you chose CFLE mode): Select setting 1, 2 or 3 with ^
Select next setting with >
e: Ramp time setting: Select with ^ or v
(ramp time will be turned off with a setting of 0)
Select next setting with >
f. Ramp pressure setting: Select with ^ or v
Select next setting with >
g. Patient disconnect setting: 1 = on 0 = off Select with ^ or v
Select next setting with >
h. Buttons lights setting: 1 = on 0 = off Select with ^ or v
(Last setting. Use On/Off button to exit Setup)
REMstar Pro 2 Setup
When in the Setup Menu, the humidifier ^ and ramp v buttons operate as up and down keys to change the settings, the left/right user buttons < > allow you to go to the previous/next question or setting, and the pressure start/stop button is used to exit the Setup Menu. Holding the humidifier or ramp buttons down will cause the values to change more quickly.
To enter the Therapy Setup Menu, hold the two top user buttons < > down while plugging in the power cord.
Continue holding the buttons down until the REMstar Pro 2 with C-Flex beeps twice.
Note: The word "setup" will appear on all of the screens indicating that you are in the Therapy Setup Menu.
If you press the Pressure start/stop button, you will exit the Setup Menu.
a. Compliance hours/nights: (recommend leaving alone, but CAN be cleared at this point)
Select next setting with >
b. Therapy mode: (CPAP or CFLE) select with ^ or v
Select next setting with >
c. CPAP or CFLE pressure setting: Select with ^ or v
Select next setting with >
d. C-Flex mode (if you chose CFLE mode): Select setting 1, 2 or 3 with ^
Select next setting with >
e: Ramp time setting: Select with ^ or v
(ramp time will be turned off with a setting of 0)
Select next setting with >
f. Ramp pressure setting: Select with ^ or v
Select next setting with >
g. Patient disconnect setting: 1 = on 0 = off Select with ^ or v
Select next setting with >
h. Buttons lights setting: 1 = on 0 = off Select with ^ or v
(Last setting. Use On/Off button to exit Setup)
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Thanks for info
I noticed you have Encore Pro Smart Card listed for you. What is it for and how does one use it?
You have to install the program Encore Pro from cd, and install a special smard card reader. Both are for sale at cpap.com for $170 Free shipping.
The program reads the data in the smartcard, and lets you know how your treatment is going, very good thing to know. Jim
The program reads the data in the smartcard, and lets you know how your treatment is going, very good thing to know. Jim
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
Your machine should have come with a Smart Card. The card captures nightly sleep statistics from your machine. Then they can be downloaded with a card reader and viewed with the Encore Pro software to determine how your therapy is working.
Here's the link where you can buy it and read more about the software:
https://www.cpap.com/productpage-bundle ... undle.html
The reason I show the Smart Card in my profile is because the only available software link is for a newer version (1.6i) than what I actually use (1.4.99).
Best wishes,
Den
Here's the link where you can buy it and read more about the software:
https://www.cpap.com/productpage-bundle ... undle.html
The reason I show the Smart Card in my profile is because the only available software link is for a newer version (1.6i) than what I actually use (1.4.99).
Best wishes,
Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05