Fired My Sleep Doctor

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: Fired My Sleep Doctor

Post by Paper_Nanny » Sat Apr 29, 2017 12:39 pm

Well, that wasn't the direction I wanted this to go. I set the BPM to 6. I don't know if that is responsible for the increase in AHI. I am slightly more groggy this morning. Suggestions?


Image

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
PR System One BiPAP Auto SV Advanced

sickwithapnea17
Posts: 472
Joined: Thu Sep 01, 2011 7:41 pm

Re: Fired My Sleep Doctor

Post by sickwithapnea17 » Sat Apr 29, 2017 1:08 pm

how are you ok with a BPM of 6?
18/14 bipap st

User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: Fired My Sleep Doctor

Post by Paper_Nanny » Sat Apr 29, 2017 10:19 pm

sickwithapnea17 wrote:how are you ok with a BPM of 6?
I don't know. Could be something like gills?

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
PR System One BiPAP Auto SV Advanced

User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: Fired My Sleep Doctor

Post by Paper_Nanny » Sun Apr 30, 2017 10:29 am

Here is the data from last night. I left everything the same. Any suggestions on anything to change to get my AHI down lower?

Image

Here are a couple zoomed in sections where I was having clusters of hypopneas. It doesn't look like the machine is really responding to those. Is this what you were talking about, palerider, when you said a machine with a more aggressive algorithm would be more effective?

Image



Image

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
PR System One BiPAP Auto SV Advanced

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: Fired My Sleep Doctor

Post by raisedfist » Sun Apr 30, 2017 11:25 am

Raise your PS min to 6.0 and see what happens. IPAP is what will knock out obstructive hypopneas.

_________________
Mask
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

User avatar
palerider
Posts: 32304
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Fired My Sleep Doctor

Post by palerider » Sun Apr 30, 2017 12:00 pm

Paper_Nanny wrote: Is this what you were talking about, palerider, when you said a machine with a more aggressive algorithm would be more effective?
yes.

be cautious with raisedfist's advice, because too much minimum PS can exacerbate central apneas.

also, please stop scrolling down the left panel before screenshots.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

sickwithapnea17
Posts: 472
Joined: Thu Sep 01, 2011 7:41 pm

Re: Fired My Sleep Doctor

Post by sickwithapnea17 » Sun Apr 30, 2017 2:48 pm

palerider wrote:
Paper_Nanny wrote: Is this what you were talking about, palerider, when you said a machine with a more aggressive algorithm would be more effective?
yes.

be cautious with raisedfist's advice, because too much minimum PS can exacerbate central apneas.

also, please stop scrolling down the left panel before screenshots.
yes I started going with 0 min ps and I feel a lot better. I do lose clarity and get foggy later on in the day but when I wake up it's much better clarity and maybe energy
18/14 bipap st

User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: Fired My Sleep Doctor

Post by Paper_Nanny » Sun Apr 30, 2017 3:35 pm

palerider wrote:be cautious with raisedfist's advice, because too much minimum PS can exacerbate central apneas.
Thanks for the warning. I will be cautious. It is at 4 right now. I will raise it to 5 tonight to see what happens. Will raise or lower depending on what does happen.
also, please stop scrolling down the left panel before screenshots.
I will pay more attention to that.

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
Last edited by Paper_Nanny on Sun Apr 30, 2017 8:22 pm, edited 2 times in total.
PR System One BiPAP Auto SV Advanced

User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: Fired My Sleep Doctor

Post by Paper_Nanny » Sun Apr 30, 2017 3:42 pm

raisedfist wrote:IPAP is what will knock out obstructive hypopneas.
It is quite probable these are central hypopneas... I was having only central apneas before the fired sleep doctor changed my settings. After the change, I was having no apnea, but many hypopneas.

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
PR System One BiPAP Auto SV Advanced

User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: Fired My Sleep Doctor

Post by Paper_Nanny » Sun Apr 30, 2017 3:56 pm

sickwithapnea17 wrote:how are you ok with a BPM of 6?
raisedfist wrote:For Respironics, their titration guide recommends the following starting point for ASV:

If you experience an inadequate breathing rate on Auto mode: Set fixed rate of 8 - 10bpm, or 2bpm below spontaneous rate. Then set inspiratory time of 1.5 seconds.
So, I double checked this on data from last night. I looked at several section of two to three minutes each when I was breathing fine with no events and counted my breaths. I looked at minutes throughout the night to get a better sample. Sample size was not scientifically wonderful, by any means, but I was only looking for anything close to perfection.

During the minutes I looked at, my spontaneous respiration rate was around 8. I set my rate 2 bpm below my spontaneous rate. Which is how I am okay with a rate of 6.

I know the Respironics titration guide is talking about inadequate breathing, but I would think it would apply to over adequacy, too.

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
PR System One BiPAP Auto SV Advanced

User avatar
palerider
Posts: 32304
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Fired My Sleep Doctor

Post by palerider » Sun Apr 30, 2017 4:36 pm

Paper_Nanny wrote:
raisedfist wrote:IPAP is what will knock out obstructive hypopneas.
It is quite probable these are central hypopneas... I was having only central apneas before the fired sleep doctor changed my settings. After the change, I was having no apnea, but many hypopneas.
I thought that the change was after your medication change in december.

though all I've got is your cpap data, not a good timeline of what external things also happened.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: Fired My Sleep Doctor

Post by Paper_Nanny » Sun Apr 30, 2017 7:03 pm

palerider wrote:though all I've got is your cpap data, not a good timeline of what external things also happened.
12/13/16-- Surgery to replace medication pump. Resulted in dose change. (AHI went up, events were almost exclusively central apnea).
03/10/17-- Increase in medication dose.
03/21/17-- Sleep doctor changed settings on machine. (Events now almost exclusively hypopneas).
04/07/17-- Change in timing of medication, but no change in overall dose.

If additional information about the medication would be helpful, let me know.

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
PR System One BiPAP Auto SV Advanced

User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: Fired My Sleep Doctor

Post by Paper_Nanny » Mon May 01, 2017 1:26 pm

Last night's data. I made no changes to the machine settings

Everything after 8:45 can probably be discounted. My dog woke me up at 8:45, I went back to sleep, but it was pretty restless and I was aware several times of fighting with my machine because I felt like I was having to breathe against what it was doing. There were 24 events during that time period.


Image

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
PR System One BiPAP Auto SV Advanced

HannaJ
Posts: 13
Joined: Thu Apr 20, 2017 12:24 am
Location: Central Coast of California, USA

Re: Fired My Sleep Doctor

Post by HannaJ » Mon May 01, 2017 1:48 pm

Paper_nanny, you don't give your location. Someone might know of a good sleep specialist in your area. Or, if you're near enough to a major metropolitan area to make occasional visits, you might look into a sleep specialist at a major medical center or university hospital. That's what I'm doing--I'll be traveling 4 hours to Los Angeles and staying one night in order to see the fellow who wrote the book on REM behavior disorder, who is at UCLA.

I'm wishing you the best in finding a solution!

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ F10 For Her Full Face Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Pressure 16; heat 2-4 dep. on room temp. I've tried FP Simplus masks and PR nasal pillow mask with chin strap and mouth tape.
HannaJ

sickwithapnea17
Posts: 472
Joined: Thu Sep 01, 2011 7:41 pm

Re: Fired My Sleep Doctor

Post by sickwithapnea17 » Mon May 01, 2017 2:07 pm

so having a 0 ps min is going to help centrals and lower central apnea events?
the epap is supposed to treat obstructives, I don't know if 4cm min is a good idea for me since I have severe obstruction of the soft palate
18/14 bipap st